Micrometer Calibration Using a CMM
Snaggy
Please can someone confirm that you cannot calibrate a micrometer using a CMM and be within the requirements of ISO 9001 / AS9100.
We have a guy here who thinks that it is acceptable and he has managed to convince the management. However we have our annual third party audit coming up and I'm certain that if picked up on it we will fail.
Thanks in advance
Richard
Michael_M
As long as the device used for calibration is traceable to NIST, you can use the CMM for calibration, however I do not think it would be the best choice for the following reasons:
Speed: Gage blocks are a faster method to be able to check the range of the micrometer.
Anvil flatness: Setting the orient, alignment and origin would be cumbersome every time you moved the position of the micrometer.
I have used our CMM to calibrate tools, mainly because of size issue (our CMM can go to 18" and our gauge blocks cannot). Our CMM is calibrated by an outside calibration company so it is traceable to NIST.
I would also think the CMM may not be accurate enough to test .00005" accurately.
Snaggy
Many thanks both of you for taking the time to answer.
The CMM is calibrated by an outside company and that calibration is traceable, the resolution of the micrometers are standard, 0.01mm. The CMM has a repeatability of 0.002mm (so not accurate enough?). However my main concern was that the CMM wasn't an acceptable method of calibrating a micrometer, and as both of you have said that there doesn't appear to be an issue with it, then that will do for me. I will double check it status as per Clause 7.6.
Again many thanks
Thanks in advance
Richard
Richard, it's not likely to be a "failure" issue at the audit, frankly. It might raise an eyebrow if the auditor is familiar with the limitations of the cmm and other factors of the micrometer, but it is highly unlikely to be anything much.
The fact is, you can "calibrate" a micrometer with a cmm! Is it as quick as using a length bar or a set of gauge blocks? No. Is it as accurate, not very likely, for all the previously stated reasons. But that won't stop you at an AS audit!
Kales Veggie
People: The Vital Few
Thanks in advance
Richard
Sure it is possible, but at what cost, how much time, fixturing, alignment, interpreting data, repeatability.
So many questions.
Gauge blocks are so much easier, faster.
shamhaider
Using right tool for right job is the key word...Gauge block are best choices when calibrating Mics/Verniers and widely adopted/acceped by Auditors...Bottom line..Keep it simple..
Snaggy
I completely agree I would always use gauge blocks. However it is not me that does the internal calibration. I will get the grief should we come up short on our third party audit, even though I have told management that I believe that the method currently used is incorrect.
I'm not sure you will "come up short" during an audit, frankly. No-one can tell the organization a calibration has to be done conventionally... It really isn't likely to be a big deal, IMHO.
Big Jim
Keep in mind that the standard does not prohibit companies from doing things in an unconventional manner. They can choose to be stupid and find ways to meet the requirements in a manner that is more time consuming and less cost effective.
I would think that some auditors would be more amused than annoyed. I would be amused to run into such a plan.
C Pin Gauge Calibration using a Micrometer - Accuracy & Calibration Level Question General Measurement Device and Calibration Topics 10
N Confuse: Calibration using master equipment - Resolution - Micrometer General Measurement Device and Calibration Topics 7
C Micrometer and Vernier Caliper simple calibration process General Measurement Device and Calibration Topics 8
G From an ISO 17025 auditor perspective must micrometer calibration check anvil flatness? General Measurement Device and Calibration Topics 4
T A question on Calibration on In-line Laser Micrometer General Measurement Device and Calibration Topics 1
C One Inch Micrometer Calibration Method General Measurement Device and Calibration Topics 2
M Calibration of a Micrometer used as a Standard General Measurement Device and Calibration Topics 5
W First Draft Micrometer Calibration Procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
C Micrometer Parallelism and Flatness Calibration Requirements and Specs General Measurement Device and Calibration Topics 10
J Calibration of Micrometer Reading up to Three Decimals General Measurement Device and Calibration Topics 3
K Calibration Procedure for ID Micrometer Calibration & Extension Rods General Measurement Device and Calibration Topics 5
P Micrometer Calibration Methods Calibration Frequency (Interval) 11
K Micrometer Calibration - Resolution, Accuracy and Range General Measurement Device and Calibration Topics 16
P How to Calibrate Laser Micrometer when no Calibration Company in the Country? General Measurement Device and Calibration Topics 3
V Calibration Procedure for Thread Pitch Micrometer General Measurement Device and Calibration Topics 3
T In-House Calibration - Digital Micrometer found Out of Calibration General Measurement Device and Calibration Topics 10
J Calibration of Flute Micrometer General Measurement Device and Calibration Topics 3
apestate Thread micrometer questions, discussion regarding MU and calibration. General Measurement Device and Calibration Topics 3
apestate Micrometer Calibration Procedure - Please critique General Measurement Device and Calibration Topics 20
N Calibration of micrometer by gauge block General Measurement Device and Calibration Topics 5
D Micrometer Calibration - Seeking actual Calibration Procedure for 0-1" Micrometer General Measurement Device and Calibration Topics 19
S How often for micrometer calibration? General Measurement Device and Calibration Topics 10
A Outside micrometer anvils parallelism General Measurement Device and Calibration Topics 3
G 0.00005" tolerance on micrometer with 0.0005" resolution? General Measurement Device and Calibration Topics 4
A Depth micrometer rod adjustment? General Measurement Device and Calibration Topics 8
A Micrometer anvil chipped General Measurement Device and Calibration Topics 19
G How to compute or estimate a micrometer or caliper flatness using optical flat General Measurement Device and Calibration Topics 2
G Procedure Creation specifically for dimensional (caliper, micrometer, dti) Document Control Systems, Procedures, Forms and Templates 8
W Standard for Micrometer Anvil Flatness Oil and Gas Industry Standards and Regulations 5
G Uncertainty Budget Examples for Caliper, Micrometer and Dial Gauge Measurement Uncertainty (MU) 3
G Caliper or Micrometer's Dimensional Flatness General Measurement Device and Calibration Topics 5
G Manual Procedure with Uncertainty Budgeting for either a Caliper or Micrometer General Measurement Device and Calibration Topics 6
M Uncertainty Budget for a Micrometer Measurement Uncertainty (MU) 5
M Laser Micrometer Suggestion (if any) for diameters that are +/-.0001 inch General Measurement Device and Calibration Topics 1
M Micrometer Anvil Placement to measure the diameter of Steel Wires Measurement Uncertainty (MU) 6
W Micrometer Storage while not in use General Measurement Device and Calibration Topics 4
D Can anyone share a completed Gage R & R on a Micrometer? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
W Thread Micrometer Standard Tolerances General Measurement Device and Calibration Topics 4
5 Reasonable GRR result for a 0-1" micrometer with a total tolerance of .002" Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 16
S Laboratory to Calibrate Objective Micrometer needed General Measurement Device and Calibration Topics 4
D Procedure on How to Calibrate a External Micrometer General Measurement Device and Calibration Topics 1
J Air Micrometer Basics - What are the internal parts and how do they work Manufacturing and Related Processes 7
M Trigonometric Calculation (V-Anvil Micrometer). I'm not getting it. General Measurement Device and Calibration Topics 6
4 Federal Specification GGG-C-105C - Caliper, Micrometer, Gage, Depth Micrometer General Measurement Device and Calibration Topics 7
D Determining Instrument Purchase based on Accuracy (Micrometer) General Measurement Device and Calibration Topics 3
A Generic or Specific Gauge R&R on Standard Equipment (ie.: Micrometer) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
D What is the correct Working Standard for Digital Micrometer? General Measurement Device and Calibration Topics 21
D Laboratory where we can have parts checked for Thickness with a Laser Micrometer General Measurement Device and Calibration Topics 6
D Laser Micrometer for checking Thickness of Steel - Any Recommendations or Comments General Measurement Device and Calibration Topics 3
M How to Fix the Standard Bar of a Screw Thread Micrometer General Measurement Device and Calibration Topics 9