Coincidentally, I posted yesterday a Venn Diagram illustration in an unconnected
calibration thread to help explain what ISO/AS says in 7.1.5. I conclude that out of the entire universe of measuring and monitoring devices existing in an organization, the requirements of 7.1.5.1 only apply to those devices related to a product conformity decision, and only a subset of those, where traceability is required
or deemed important by the organization, does the 7.1.5.2 requirement to calibrate, verify, or both, apply. That is how I parse 7.1.5. To
@Bev D's point, the organization can choose to do more than what the quality standard requires, and that may be smart, but it is also wise to know what actions are required and which are at your election.
To
@Ron Rompen's point, there is a requirement to document fitness of purpose for a M/M device involved in a product conformity decision (7.1.5.1.3) but no requirement to document why the organization thinks calibration and/or verification are not necessary. I chose to add a separate tab to my Calibration Register to document those decisions, and I apply a Calibration Not Required label to many of such devices, although labeling ("identification to determine status") is only required for M/M devices that fall inside 7.1.5.2.