Hello I have some questions concerning the IEC 60601-1 standard, I would be very grateful if you could clarify me these points:
- In our device 3 wires go from the power ac inlet to the power supply, how 2 MOOP are achieved between the two phases?, and between each phase and protective earth?. It is enough with the normal insulation of the wires or something more is required?, creepage and clearance distance apply in this case?. Does the fastening of the wires need special protection?
- Are creepage and clearance distances required in secondary circuits with voltages below 15 V?
-Does plastic enclosure counts as 1 MOOP or as 2 MOOP?
- Can/Should secondary circuits and SIP/SOP circuits be earthed to protective earth?
Thanks in advance!
Peter Selvey
Re: MOP (means of protection), Creepage and Clearance questions
The wires from the wall to the equipment are covered by separate standards and regulations, usually the designer of the equipment just buys approved cords/plugs etc.
If the wire is non-detachable (wired into the equipment to terminals), there are a bunch of requirements in the standard to make it safe. Clause 7.3 covers marking, 8.6.2 for the earthing terminal, 8.11.3.5 for cord anchorage, 8.11.4 for general requirements for terminals. There may be more. It is an important area so better to involve someone with some reasonable electrical background in the design, and best is to then get it independently checked out by a qualified test lab to catch any residual mistakes.
Generally 15Vdc circuits do not need any special isolation unless there is a potential for direct connection to the patient.
SIP/SOPs circuits can be earthed or floating. For SOP/SOPs, there is some complicated stuff in the standard but in the end there is usually nothing special in the end and it would take a lot to explain.
F
fonk140
Thank you very much for your answer Peter.
The questions about the wires were about the internal wires of the equipment that connect internally the mains voltage to the internal power supply. Do they need any special insulation or protection?
Does creepage and clearance apply only to traces in PCB or also to "normal" wires and cables?
Regarding the plastic enclosure, how many level of protection it provides, 1 or 2 MOP?
Peter Selvey
The best way to think about it is to consider the mains voltages as being "hot" and the "means of protection" is then the barrier around them to prevent people burning themselves.
For electric shock, the standard does not care about design or spacings inside the mains parts, only the quality of the barrier around them.
For simplicity, most designers segregate circuits so the barrier is nice and clear and easy to verify for compliance. On PCBs, sometimes designers will put a dotted line on the silkscreen to show where the barrier is. Usually only a few component bridge the barrier, like optocouplers, capacitors, transformers, and these are specially designed to provide a safety barrier. Outside of PCBs, wiring insulation or physical separation is the most common method to form a barrier.
A common mistake is to use an certified open frame power supply but then run secondary wiring touching or close to bare mains parts such as a primary heatsink. The only barrier then is insulation on the secondary wiring, which is usually not enough for 2 MOP. Common solutions are for example routing the secondary wiring away from the power supply, or putting sleeving the secondary wiring.
Generally, the external enclosure is well removed from the mains circuits, with the air gap to the enclosure is >> 2 MOP (typically 5mm is enough for 2 MOP). If so there are no electrical tests. But it still may form a mechanical protection which is subject to non-electrical tests, such as force tests, impact, mold stress etc.
F
fonk140
And how is calculated the protection level of an insulated wire?
For example if we have a 18 AWG wire with a nominal tension of 300V can we determine that it has 2 MOOP at 220V or it necessary to do a an HIPOT test?
Peter Selvey
If the wiring is rated and certified for 300V this can be treated as 1 MOP without further test.
Strictly the 300V rating should be using IEC standards, but in practice most agencies (and manufacturers) will also accept US based tests (e.g. UL rating for 300V).
F
fonk140
And what voltage should be certified to have 2 MOP?
F
fonk140
And what about the fastening of the wires, how can we determine if it provides 1 or 2 MOPs?
Peter Selvey
You need to check the rating (spec sheet) of the wiring. It should show if the wiring is designed to provide double insulation (2 MOP). It is fairly common so should not be hard to find. The rating may be 600V, but check also that it is certified as being double insulation.
If a detached wire can cause double insulation to be bridged (e.g. mains wire disconnects and can touch secondary, or vice versa), it should have two means of securement. Examples of common methods are double crimp connectors (connector with wire and insulation separately crimped), or solder + cable tie.
F
fonk140
Two more questions:
1) if a secondary cable could go near or even touch a primary (mains) wire, it is necessary that the secondary cable has a nominal working voltage equal or superior to mains voltage to comply with 2 MOP between primary and secondary?
2) you said in your first post that secondary circuits doesn't need isolation, so in what situations are creepage and clearance of table 15 (distances for secondary circuits) used?
F Evaluation of MOP (Means of Protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
T Battery Powered Device - MOP and Patient Auxiliary Currents IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Do I need to audit the supplier supplying mop heads for our Clean Room ? ISO 13485:2016 - Medical Device Quality Management Systems 8
B QM, POL, SOP, WI -- COP, MOP - Clarification needed. Document Control Systems, Procedures, Forms and Templates 3
S Process Description Worksheet (COP, MOP, SOP) example needed Manufacturing and Related Processes 3
A COP, MOP, SOP - ISO 9001:2000 requirements? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
B Objectives for each process, COPs, MOP's and SOP's - How many is enough? IATF 16949 - Automotive Quality Systems Standard 5
H COP (Customer Oriented Process), MOP and SOP Examples and Explanation IATF 16949 - Automotive Quality Systems Standard 21
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
M Informational How to perform a clinical evaluation of medical devices – Part 2 – Level of clinical evidence and what sufficient clinical evidence means Medical Device and FDA Regulations and Standards News 9
M Applicability of Means of Protection, working voltage in an Automated External Defibrillator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J President does not know what the Quality Policy means ISO 13485:2016 - Medical Device Quality Management Systems 10
G What degrees of freedom means in uncertainty budget and how do I set the value Measurement Uncertainty (MU) 2
P Uncertainty and Significantly Different Means - Buffer Solutions Measurement Uncertainty (MU) 4
G Definition Parallel Loop - Does anyone know what the term Parallel loop means? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 2
N Explanation - Quality means doing it right when no one is looking ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
P Question on Table 6 "Test voltages for solid insulation forming a MEANS OF PROTECTION IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Uncertainty Measurement - What this accuracy specification "1% + 10^5": means? General Measurement Device and Calibration Topics 2
F How to provide 2 MOPP (Means of Patient Protection) - Heating Rod IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
L What does "conformance matrix" means in terms of ISO 9001:2008? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
B Use of Y1 capacitors for MOPP (means of patient protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Correlated data or not? Determining the significance of differences in means Statistical Analysis Tools, Techniques and SPC 3
Q Definition C - What the "C" in C=0 Sample Plan means Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 10
J Minitab Sample Size Calculation: 1 Sample t vs Minimum Sample Size for Means Using Minitab Software 3
L What exactly does Configuration Management in ISO 9001, Clause 7.5.3 means? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
somashekar Friday the 13th - What it means to people around here. Coffee Break and Water Cooler Discussions 9
M Can I continue using IEC60601-1 Ed2 PSU for MOOP (Means Of Operator Protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
B Test of Means or Medians for Non Normal Populations? Using Minitab Software 6
F What Compliance to ISO 9001, 6.2.2 Competence, Awareness and Training Means ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 35
A Customer Demands (MSA=Bias + GRR/2). Can somebody explain what it means? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
J Ford Acronym - What the acronym "ym" means in Ford lingo? Customer and Company Specific Requirements 1
S How to Design the Means of Protection for Home Use Unit IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
M Definition Disposition Definition - Disposition means Prevention of Recurrence? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 24
M Surface mount resistor as a MOPP (Means Of Patient Protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
Wes Bucey Keeping your cool means keeping your job Career and Occupation Discussions 12
D Means Of Protection of the Video OUT of an Internally powered device IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
bobdoering From SME: What the Stimulus Plan Means to Manufacturers World News 1
P Public Access to Environmental Policy - Means and Methods Miscellaneous Environmental Standards and EMS Related Discussions 10
Stijloor What Your Boss Says and What He Really Means Coffee Break and Water Cooler Discussions 3
M Means to Achieve Objectives and Targets - ISO14001 4.3.4 Miscellaneous Environmental Standards and EMS Related Discussions 1
K Process monitoring, measuring and analyzing, does this means all company processes? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Marc Harley-Davidson: Strike means motorcycle shipments will be short World News 6
Jim Wynne 4th Edition PPAP manual - Bulk Materials - Can you guess what this means? APQP and PPAP 12
D New to PPAP and APQP - Don't really understand what all of it means APQP and PPAP 7
S Whether the word BETA means prototype samples or the Production samples IATF 16949 - Automotive Quality Systems Standard 1
T ESD test 50% flashover discharge - Do anyone know that 50% flashover discharge means? Reliability Analysis - Predictions, Testing and Standards 6
A Has anyone heard of "drop gauge policy"? If so, can you tell us what it means? General Measurement Device and Calibration Topics 17
M 7.6 - Validity of the previous measuring results - Is this what validity means? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
J AS9102 section 3 - Who can explain what ''standard catalog hardware '' means AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
K TS 16949 PPAP - What it means: SQAM Forms? Supplier Quality Assurance Manual? IATF 16949 - Automotive Quality Systems Standard 4