MSA after the device move

Nihls

Registered
#1
Hello,
We will move a form measuring device from one laboratory to another in the company. We will do an MSA study 1 before and after the move. Is an MSA Study 3 required after MSA Procedure 1?
If you could explain, I would be very happy.
Thanks in advance.
 
Elsmar Forum Sponsor

Miner

Forum Moderator
Leader
Admin
#2
The Type 1 R study will assess whether there was any impact to the device itself from the move. If you think that the difference in operators running the equipment may be significant, run the R&R study. If they are equally skilled, it will probably not be necessary.
 

Welshwizard

Involved In Discussions
#3
Hi Nihls,

Its tempting to constrain an answer to your question to MSA tools, but in doing so it will limit learning if this situation arises again in the future.

You have a measurement system which needs to be moved elsewhere. The sources of variation during operation will be largely centered around people, methods, the equipment and the environment.

So, if you don't currently have a study which flushes out and can quantify the effects of these variables in your current equipment situation you should. When the equipment moves it may be subjected to a new environment, new less experienced people for example so you need to ascertain whether these sources affect your measurement result.

A type 1 GRR is effectively a run chart of the measurand measured multiple times with the repeated values plotted against tolerances with some statistics computed. The value of this could be debated at length but its clear its one person measuring one part multiple times.

If your new location uses the same operator measuring the same feature on the same machine, same method with the same control on the environment then another Type 1 is logical. If however you know or suspect you will use different operators or other aspects have changed or might change it would makes sense to do the full GRR to flush out the effects.

If the machine is delicate there may be an effect to the accuracy as well as the repeatability from the move so you it makes sense to run your Type 1 with a calibrated artefact before and after which the study takes into account.

If you have to run a full GRR after the move I would advise that you have one to reference before the move too!

Hope this helps
 
Thread starter Similar threads Forum Replies Date
W Medical Device - fixture qualification and MSA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
Emran.mi Measurement system analysis - Can you help me about implementation MSA for CMM device Manufacturing and Related Processes 2
N MSA Study for a Leak Testing Device Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
K Measuring Device -Timer MSA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
S MSA (Measurement System Analysis) for Medical Device Manufacturers Other US Medical Device Regulations 3
C MSA for resistance measurement Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
A MSA and Poison (Red Rabbit) Test Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
C Part selection for Faro arm MSA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
M Spot Weld Nugget - MSA Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
P Type of MSA study according to measurement equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
M MSA on wire pull & ball shear Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
S MSA annual plan Capability, Accuracy and Stability - Processes, Machines, etc. 5
S If 4 operate to measure, what is the value of k2?(MSA gage R&R) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
8 MSA--Should I Use Parts or Gage Pins? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
W Conducting an MSA Test on scales Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
A MSA samples Preparation. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
N MSA Frequency related literature Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
A %AV=0 in the MSA GRR Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
M MSA - Dealing with inherent within-part variation in assemblies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
T Ford Q1 MSA - Virtual Site Visit Hardware Manufacturing and Related Processes 14
R MSA on set of parts with low variation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 20
E MSA Reference Manual 2nd Edition Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
V MSA - Using a R&R study for a Tape Measure Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
M Options when MSA acc. to VDA 5 is NOK (Qms > 30%) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
F MSA for non-contact CMM - ISO10360 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
T Thermometers and Thermocouple or Transducer's require MSA? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
R MSA Studies for chemical analytical instruments in QC Lab Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
A IATF MSA study Type 1 General Measurement Device and Calibration Topics 7
M MSA angle issue - Nylon parts using Aberlink 3D Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
Ashland78 MSA clause or wording needed. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
S MSA clause 7.1.5.1.1 - Statistical studies on counting weight scale Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
Nihls MSA Study Type 2 (special conditions ) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
Nihls When the MSA results show no operator influence Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
S MSA for attribute relation gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
Nihls MSA Study Type 1 (CMM) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
M MSA Study Type 1 not capable. We are at the limit. And manufacturing wants to continue producing. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
I Overwhelmed with attribute MSA requirement for visual inspection IATF 16949 - Automotive Quality Systems Standard 11
H Need MSA 4th ed. compliant attribute MSA template General Measurement Device and Calibration Topics 4
P MSA for titration Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
P MSA - what exactly mean "system" and master sample Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
P MSA study for visual system with artifical inteligence Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
D MSA strategy Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
K Updated MSA with the Annual PPAP validations APQP and PPAP 8
E MSA for Push and Pull test Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
6 MSA for Calculated Clearance Dimension Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
Ron Rompen MSA on automated measurement system - Multiple Step Vision System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
G MSA check list to audit IATF 16949 - Automotive Quality Systems Standard 10
R MSA for ATE (Automatic Test Equipment Embedded Software) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
Bev D MSA Tools Misc. Quality Assurance and Business Systems Related Topics 3
R MSA studies for Heat Treatment Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1

Similar threads

Top Bottom