MSA for Push and Pull test

ElenR

Registered
#1
We received a Minor NC due to no MSA conducted on our Push and Pull Test from our IATF Surveillance audit.
We told the auditor that the bonding force is not classified as special characteristics on the control plan thereby the push and pull tester will not require mandatory MSA. But the auditor said that MSA applies to any type of measurement, inspection, and testing referenced in the Control Plan. It's just that those with SC needed to be prioritized. Is it valid that the push n pull tester need to undergo MSA?

Our bonding force check is sampling check to ensure our ferrite core will not be detached from its base after it undergone curing process.
 
Last edited:
Elsmar Forum Sponsor

Ninja

Looking for Reality
Trusted Information Resource
#2
Howdy ElenR,

I won't answer your question directly, though I'm sure others will.
I will however walk down a train of thought...

"Our bonding force check is sampling check to ensure our ferrite core will not be detached from its base after it undergone curing process. "
Do you care about the results of the test?
"No" = stop doing the test
"Yes" = next question:

Do you care if the test is accurate?
"No" = stop doing the test
"Yes" = next question:

How do you know your test is accurate?

In automotive..."Accuracy" in a customer or auditor's eyes is almost always answered by GR&R.
I'm sure there are other valid answers to the question...by why not just answer it in the parlance of the field?
This is one of the (many) areas that I have found it cheaper and more efficient to just comply rather than to prove that another method is valid.
You can make a stand, and prove your point (over and over and over and over and over) ... or just do the GRR.

HTH
 

Miner

Forum Moderator
Leader
Admin
#3
@Ninja I agree with the intent of your post, but want to make a correction:
  • Accuracy - also known as bias or a measure of the location of the measurements. This is measured by a bias, linearity or stability study and is controlled by calibration.
  • Precision - a measure of the variation or spread of the measurements. This is measured by an R&R study.
To quote another cover's tagline: "Because words have meaning."
 

Ninja

Looking for Reality
Trusted Information Resource
#4
You are completely correct.

I used the word "Accuracy" since it is the only word I've ever heard automotive personnel use when asking about a vendor test method "Is it good enough to meet our needs".

Words have meaning...and it's fairly typical for folks to use them incorrectly.
Thank you for the clarification.
 
#5
@Ninja I agree with the intent of your post, but want to make a correction:
  • Accuracy - also known as bias or a measure of the location of the measurements. This is measured by a bias, linearity or stability study and is controlled by calibration.
  • Precision - a measure of the variation or spread of the measurements. This is measured by an R&R study.
To quote another cover's tagline: "Because words have meaning."
If I understood it right, the nonconformity statement should not No MSA since our Push and Pull Test undergone calibration but we failed to conduct R&R study, By the way, thank you for the clarification.
 

Ninja

Looking for Reality
Trusted Information Resource
#6
Hello again ElenR,

Calibration and MSA study are two different things.

"Calibration" is verifying to a known good value (typically NIST traceable) that the equipment has the ability to measure accurately (See Miner's definition above).

MSA study considers both the equipment and how you use it (the complete test method).

If you've only calibrated the gage but then use it poorly, or if it is simply not the right tool for the job, you can totally fail MSA with a newly calibrated tool.

I've seen many tools that were calibrated, then used with no idea (no GRR) if the results meant anything or not.
 
#7
We received a Minor NC due to no MSA conducted on our Push and Pull Test from our IATF Surveillance audit.
We told the auditor that the bonding force is not classified as special characteristics on the control plan thereby the push and pull tester will not require mandatory MSA. But the auditor said that MSA applies to any type of measurement, inspection, and testing referenced in the Control Plan. It's just that those with SC needed to be prioritized. Is it valid that the push n pull tester need to undergo MSA?

Our bonding force check is sampling check to ensure our ferrite core will not be detached from its base after it undergone curing process.

What kind of MSA study did you end up using?
 

WEAVER

Involved In Discussions
#8
@Ninja I agree with the intent of your post, but want to make a correction:
  • Accuracy - also known as bias or a measure of the location of the measurements. This is measured by a bias, linearity or stability study and is controlled by calibration.
  • Precision - a measure of the variation or spread of the measurements. This is measured by an R&R study.
To quote another cover's tagline: "Because words have meaning."
Are BIAS, LINEARITY AND STABILITY STUDIES required for every calibration? Like in some of our calibrated instruments there is only one test point -- we are not required to calibrate it in other measurement range since it's application is that test point only; so we don't have the data for linearity. Are we now required to calibrate it in it's full range since we need to evaluate its linearity?
How can we confirm (and prove to auditors and customers) that our calibration plan conforms to these three parameters?
Thanks in advance,
 

Miner

Forum Moderator
Leader
Admin
#9
I recommend that you perform a risk assessment for the risk of not performing a linearity study for an instrument with a single test point. If the risk is low, skip the linearity evaluation and include your risk assessment with the records for this instrument.
 

WEAVER

Involved In Discussions
#10
I recommend that you perform a risk assessment for the risk of not performing a linearity study for an instrument with a single test point. If the risk is low, skip the linearity evaluation and include your risk assessment with the records for this instrument.
Thank you so much, sir, this has been most helpful!
 
Thread starter Similar threads Forum Replies Date
C Part selection for Faro arm MSA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
M Spot Weld Nugget - MSA Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
P Type of MSA study according to measurement equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
M MSA on wire pull & ball shear Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
S MSA annual plan Capability, Accuracy and Stability - Processes, Machines, etc. 5
S If 4 operate to measure, what is the value of k2?(MSA gage R&R) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
8 MSA--Should I Use Parts or Gage Pins? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
W Conducting an MSA Test on scales Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
A MSA samples Preparation. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
N MSA Frequency related literature Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
A %AV=0 in the MSA GRR Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
M MSA - Dealing with inherent within-part variation in assemblies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
T Ford Q1 MSA - Virtual Site Visit Hardware Manufacturing and Related Processes 14
R MSA on set of parts with low variation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 20
E MSA Reference Manual 2nd Edition Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
W Medical Device - fixture qualification and MSA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
V MSA - Using a R&R study for a Tape Measure Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
M Options when MSA acc. to VDA 5 is NOK (Qms > 30%) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
F MSA for non-contact CMM - ISO10360 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
T Thermometers and Thermocouple or Transducer's require MSA? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
R MSA Studies for chemical analytical instruments in QC Lab Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
A IATF MSA study Type 1 General Measurement Device and Calibration Topics 7
Nihls MSA after the device move Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
M MSA angle issue - Nylon parts using Aberlink 3D Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
Ashland78 MSA clause or wording needed. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
S MSA clause 7.1.5.1.1 - Statistical studies on counting weight scale Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
Nihls MSA Study Type 2 (special conditions ) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
Nihls When the MSA results show no operator influence Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
S MSA for attribute relation gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
Nihls MSA Study Type 1 (CMM) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
M MSA Study Type 1 not capable. We are at the limit. And manufacturing wants to continue producing. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
I Overwhelmed with attribute MSA requirement for visual inspection IATF 16949 - Automotive Quality Systems Standard 11
H Need MSA 4th ed. compliant attribute MSA template General Measurement Device and Calibration Topics 4
P MSA for titration Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
P MSA - what exactly mean "system" and master sample Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
P MSA study for visual system with artifical inteligence Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
D MSA strategy Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
K Updated MSA with the Annual PPAP validations APQP and PPAP 8
6 MSA for Calculated Clearance Dimension Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
Ron Rompen MSA on automated measurement system - Multiple Step Vision System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
G MSA check list to audit IATF 16949 - Automotive Quality Systems Standard 10
R MSA for ATE (Automatic Test Equipment Embedded Software) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
Bev D MSA Tools Misc. Quality Assurance and Business Systems Related Topics 3
R MSA studies for Heat Treatment Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
C Industrial scales and MSA (IATF 16949 requirement 7.1.5.1.1) IATF 16949 - Automotive Quality Systems Standard 30
M Difference between MSA and MSE? General Measurement Device and Calibration Topics 1
M Test method validation - Is MSA (MSA1, MSA2, MSA3 and linearity) a good solution? Medical Device and FDA Regulations and Standards News 2
A MSA results differences - Supplier results vs. My results Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
R Preliminary MSA and Sockets' correlation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
I MSA requirement for 5 Micrometers + CP changes need customer approval? IATF 16949 - Automotive Quality Systems Standard 2