MSA Studies on each Equipment, each Measurement System or Family in Control Plan?

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#11
Re: MSA studies: on each Equipment, each Measurement System or Family in Control Plan

I don't know how long the tools have been in use. I don't know their understood variation over time.
I agree with you points, except that these items refer to calibration and stability - not gage R&R. A valid gage r&r has no time function - but calibration and stability definitely do!
 
Elsmar Forum Sponsor

qusys

Trusted Information Resource
#12
Re: MSA studies: on each Equipment, each Measurement System or Family in Control Plan

My organization just performed Gage R&R studies on 2 identical measuring tools. Both are under the "same" maintenance and calibration process. Both measure the exact same product. Both are used in the same way, in the same production environment.

One tool failed and the other passed.

Does this prove anything? No.

But consider if we had performed a Gage R&R on the tool that passed and applied the results to the "family" of 2 tools? What could the impact have been?

Or conversely, what if we had performed the Gage R&R on only the failed tool and applied the results to the family?

I am not sure if approaching all tools listed on a Control Plan or in a measurement system, as a family, is always the most effective approach.

Who would know? How would they know?

ISO TS does not mention statistical studies on each equipment on CP, so this could not be a requirement. Different thing if the requirement should come from the Customer , for example inside MSA studies in PPAP documentation. What about??
 

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#13
Re: MSA studies: on each Equipment, each Measurement System or Family in Control Plan

ISO TS does not mention statistical studies on each equipment on CP, so this could not be a requirement. Different thing if the requirement should come from the Customer , for example inside MSA studies in PPAP documentation. What about??
Yes, some customers have specific requirements. Ford appears to looking for Gage R&R for each device on the control plan - as well as capability and distribution of the characteristic. Others ask for Gage R&R annually - even though it is not a time function - calibration and stability are the time functions for gages. So...you have to keep your eyes open in TS - especially for customer specifics.
 
V

vanputten

#14
Re: MSA studies: on each Equipment, each Measurement System or Family in Control Plan

Bobdoering: Please notice that the original posting referred to MSA Studies. The original posting was not about only Gage R&R. Therefore, I included the time parameter in my thoughts because of stability studies.

Also, I am going to guess (Note - a total guess with no empirical data) that 20 year old identical tools, even if under the "same" PM and calibration controls, will exhibit more variation between the tools than brand new tools. I am going to guess that PM processes and calibration process may not be effective enough to control for all tool variation over time. This can be debated to infinity. I am simply providing this thought for thought.

Maybe we need an operational definition from the original poster as to what they are really asking. Are they asking about doing only gage r&r assessments for tools or are they asking about studies of their measurement system?
 
V

vanputten

#15
Re: MSA studies: on each Equipment, each Measurement System or Family in Control Plan

Qusys:

7.6.1 refers to "measurement systems referenced in the control plan."

What is a "measurement system?" The answer to this question may answer your original questions.
 

Jim Wynne

Staff member
Admin
#16
Re: MSA studies: on each Equipment, each Measurement System or Family in Control Plan

Bobdoering: Please notice that the original posting referred to MSA Studies. The original posting was not about only Gage R&R. Therefore, I included the time parameter in my thoughts because of stability studies.
MSA and calibration are separate categories. It's assumed that MSA is being performed on properly calibrated devices; thus the stability aspect is assumed.

Also, I am going to guess (Note - a total guess with no empirical data) that 20 year old identical tools, even if under the "same" PM and calibration controls, will exhibit more variation between the tools than brand new tools. I am going to guess that PM processes and calibration process may not be effective enough to control for all tool variation over time. This can be debated to infinity. I am simply providing this thought for thought.
What you're referring to is a problem with the calibration process, not the MSA process. If calibration intervals are set such that devices are found to be out of calibration at the end of the calibration period, someone needs to investigate. Note that calibration requirements apply to all measurement devices "when necessary to ensure valid results." The results of MSA are indicative of whether or not a give device (or type of device) is (when properly calibrated) appropriate to the task at hand. If the assumption is that measurement systems (device + operator + measured feature) in a given family might perform differently from one another in production, there's something wrong with the gage (addressed by calibration) or the operator (addressed by MSA).

Maybe we need an operational definition from the original poster as to what they are really asking. Are they asking about doing only gage r&r assessments for tools or are they asking about studies of their measurement system?
A GR&R is a study of a measurement system, although it's not the only type of study.
 

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#17
Re: MSA studies: on each Equipment, each Measurement System or Family in Control Plan

MSA and calibration are separate categories. It's assumed that MSA is being performed on properly calibrated devices; thus the stability aspect is assumed.
I have always interpreted (based on the AIAG book) MSA as the overall analysis of the measurement system, with the subcategories of: Location (or Calibration) Variation -Accuracy, Bias, Stability, Linearity; Width Variation - precision, gage R&R, Capability, Performance, Sensitivity, Consistency and Uniformity; and System Variation Capability, Performance and Uncertainty (AIAG MSA 4th ed page 5-8).

It is somewhat academic - as long as it is done properly and understood. But, one this is clear; MSA is not Gage R&R, and I hear that misconception a lot. Gage R&R is a subset of MSA.
 

qusys

Trusted Information Resource
#18
Re: MSA studies: on each Equipment, each Measurement System or Family in Control Plan

Bobdoering: Please notice that the original posting referred to MSA Studies. The original posting was not about only Gage R&R. Therefore, I included the time parameter in my thoughts because of stability studies.

Also, I am going to guess (Note - a total guess with no empirical data) that 20 year old identical tools, even if under the "same" PM and calibration controls, will exhibit more variation between the tools than brand new tools. I am going to guess that PM processes and calibration process may not be effective enough to control for all tool variation over time. This can be debated to infinity. I am simply providing this thought for thought.

Maybe we need an operational definition from the original poster as to what they are really asking. Are they asking about doing only gage r&r assessments for tools or are they asking about studies of their measurement system?
First thanks for response to you, Jim and BoB.
My original question is related to how to accomplish ISO TS for clause 7.6.1.
In the control plan there are reported several equipment measurement released , they belong to same technology and subjected to the same calibration and maintenance plan.
On this measurement equipment production material is measured by operator and equipment are completely automated. The operator charges the materials on one of them , the robot selects some sampled materials ( according to a sample plan) and measures some caractheristics according to a program .
The question is : to make G R&R for all of them or I can select some samples of them? For example I have 15 equipment of a certain technology, instead of doing 15 G R&R I could do 3 according to some criteria, for another technology I have 20 equiment , instead of doing 20 G R&R , I could do 5 ( according to some criteria that we could establish).
What about?
Which could be the criteria?:bigwave:
 

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#19
Re: MSA studies: on each Equipment, each Measurement System or Family in Control Plan

First thanks for response to you, Jim and Bob.
My original question is related to how to accomplish ISO TS for clause 7.6.1.
In the control plan there are reported several equipment measurement released , they belong to same technology and subjected to the same calibration and maintenance plan.
On this measurement equipment production material is measured by operator and equipment are completely automated. The operator charges the materials on one of them , the robot selects some sampled materials ( according to a sample plan) and measures some characteristics according to a program .
The question is : to make G R&R for all of them or I can select some samples of them? For example I have 15 equipment of a certain technology, instead of doing 15 G R&R I could do 3 according to some criteria, for another technology I have 20 equipment , instead of doing 20 G R&R , I could do 5 ( according to some criteria that we could establish).
What about?
Which could be the criteria?:bigwave:
Here is one of the best answers - ask your SQE. If your customer only wants GR&R for critical characteristics, fine. If they want Gage R and stability for automated gages, fine. If they want every gage Gage R&R reported - better to know before you submit PPAP, rather than have it rejected and spend a couple frantic weeks trying to make up for it.

The criteria for performing gage R&R is NOT the number of characteristics or machines you have, but the adequacy of the knowledge you have of the machines to assure they are appropriate for the job. You may have agreement with your SQE that the machines and their function are so similar that sampling is acceptable, but otherwise there is nothing that would support that - especially if the only supporting argument is time and effort. The other side of that argument is if you do not do the Gage R&R, how do you KNOW the machine is adequate? Guessing that they are kind of similar? Not good enough.
 

qusys

Trusted Information Resource
#20
Re: MSA studies: on each Equipment, each Measurement System or Family in Control Plan

Here is one of the best answers - ask your SQE. If your customer only wants GR&R for critical characteristics, fine. If they want Gage R and stability for automated gages, fine. If they want every gage Gage R&R reported - better to know before you submit PPAP, rather than have it rejected and spend a couple frantic weeks trying to make up for it.

The criteria for performing gage R&R is NOT the number of characteristics or machines you have, but the adequacy of the knowledge you have of the machines to assure they are appropriate for the job. You may have agreement with your SQE that the machines and their function are so similar that sampling is acceptable, but otherwise there is nothing that would support that - especially if the only supporting argument is time and effort. The other side of that argument is if you do not do the Gage R&R, how do you KNOW the machine is adequate? Guessing that they are kind of similar? Not good enough.
Good point!
Are matching studies among several equipment acceptable for this?
To return to your post, sometimes it is a question of time and effort. Sometimes special characteristics changes and your previous job of G R&R fades away or there are brand new special characteristics that are communicated in a second time!
To also consider that PTR is between 10% and 30% , this is communicated to the customer but imporvment programs are very difficult to implement, other you should change completely the equipment given that they are on leading edge.
 
Thread starter Similar threads Forum Replies Date
R MSA studies for Heat Treatment Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
P IATF 16949 MSA Studies for CMM machines IATF 16949 - Automotive Quality Systems Standard 3
L Number of Distinct Categories" or NDC Calculation in MSA Studies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
P Frequency and Sample Size Requirements for an MSA Studies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Ron Rompen MSA Studies of Instruments in a Paint Shop Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
A Is conducting of all MSA studies required, even if not required by the customer? IATF 16949 - Automotive Quality Systems Standard 23
B For a measuring instrument, should MSA be done first or the Capability studies? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
B Best Methods for conducting Shadowgraph MSA Studies Quality Manager and Management Related Issues 2
D Requiring a "bad" part for attribute MSA studies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
B How many MSA studies are required and at what interval(s)? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
J Analysis of GRR studies, MSA 4th Ed, Chapter III section B - Rigid Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
G Traceability Studies and Requirements for MSA (Measurement Systems Analysis) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
G Dealing with Excessive Within-Part Variation when doing MSA and Cpk studies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
B How many Samples are required in Attribute Data MSA Studies? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
S What is the Reference Value when we do Attribute MSA Studies? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 17
S The assessment of ndc figures in MSA studies. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
J ANOVA Method for MSA studies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
S TS 16949 MSA Study Requirements - MSA Studies by Family Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
P MSA Attribute Studies - Calculate Cross Tabs, Effectiveness, Miss Rate, False Alarm Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
Ron Rompen MSA to perform Linearity and Stability studies on Variable Gauges with Minitab Ver 15 Using Minitab Software 6
L Is it necessary to carry out all the MSA studies? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
M MSA Studies when not required by a customer Statistical Analysis Tools, Techniques and SPC 6
R "Gauge (Gage) R&R Studies per MSA Manual" - Customer Audit Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 13
D MSA - Can Regular Calibration Replace Bias, Linearity and Stability Studies? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
J What MSA studies do you include in PPAP level 3? APQP and PPAP 2
C What are the minimum required MSA studies for a TS 16949 PPAP? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
Courier MSA Tools - Excel templates for doing the Linearity & Bias, & Stability Studies Excel .xls Spreadsheet Templates and Tools 13
D Measurement System Analysis (MSA) - Requirement to also do gage stability studies? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
X MSA Studies on a scale that is used in the calculation of Specific Gravity (SG) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
Y MSA - Torque Wrenches - Statistical studies that cover requirements of MSA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 15
G MSA Studies on heat shock tests - Should we just be calibrating the oven? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
B PPK Requirements & MSA Studies for PPAP APQP and PPAP 10
T APQP Control Plans - MSA studies for everything listed on the plan? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Nihls MSA Study Type 2 (special conditions ) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
Nihls When the MSA results show no operator influence Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
S MSA for attribute relation gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
Nihls MSA Study Type 1 (CMM) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
M MSA Study Type 1 not capable. We are at the limit. And manufacturing wants to continue producing. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
I Overwhelmed with attribute MSA requirement for visual inspection IATF 16949 - Automotive Quality Systems Standard 8
H Need MSA 4th ed. compliant attribute MSA template General Measurement Device and Calibration Topics 4
P MSA for titration Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
P MSA - what exactly mean "system" and master sample Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
P MSA study for visual system with artifical inteligence Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
D MSA strategy Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
K Updated MSA with the Annual PPAP validations APQP and PPAP 8
E MSA for Push and Pull test Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
6 MSA for Calculated Clearance Dimension Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1