CREED said:
DOES ANY ONE KNOW WHAT THE RECOMMENDATION FOR PV% SHOULD BE. ACCORDING TO THE MSA MANUAL, GRR% SHOULD BE UNDER 10% GOOD, 10%- 30% MAY BE ACCEPTABLE, OVER 30% NOT ACCEPTABLE. ndc SHOULD BE GREATER THAN 5. BUT IT DOES NOT GIVE ANY GUIDE LINES FOR PV%.
The guidelines are the same: 10% good, 10%-30% may be acceptable, over 30% not acceptable.

The difference between the metrics is %GRR is the acceptability of the gage for process control (or statistical/capability studies) and P/V is the acceptability of the gage for product acceptance (inspection).

See the following threads for more detail:
http://elsmar.com/Forums/showthread.php?p=124339#post124339
http://elsmar.com/Forums/showthread.php?p=125257#post125257
 
Elsmar Forum Sponsor
C

Chris Willis

#12
Well, this has just caught me out, we have been evaluating a variety of MSA forms that we use worldwide in out organisation.

Using three of the forms and the test data from MSA 3, we chose our form from the German side of the business (it had been adjusted for the new constant values). However when we input data from and old MSA 2 study into the MSA 3 form we found that we had a brilliant GRR.

At this point we were surprised, and almost framed the results. This new form was brilliant, GRR less than 1% of the overall tolerance, I'll have some of that please.

Then the old quality questions triggered in the brain, WHY has this happened. What is different. A line by line reading of the formulae reveals we have an MSA 2.5 form, the constants K1, K2 and K3 have been adjusted to MSA 3, the tolerence calculation has remained at MSA 2. The result, when based on a tolerance calulation we are out by a factor of SIX!

Why is this possible to miss, alas as the MSA manuals only show test data and results based on Total Variation and not Tolerance. So all of our testing showed that all of our calculations were correct and the error slipped through undetected.

Ooops,

Chris
 
A

Atul Khandekar

#13
Re: Msa Manual Third Edition

DOES ANY ONE KNOW WHAT THE RECOMMENDATION FOR PV% SHOULD BE. ACCORDING TO THE MSA MANUAL, GRR% SHOULD BE UNDER 10% GOOD, 10%- 30% MAY BE ACCEPTABLE, OVER 30% NOT ACCEPTABLE. ndc SHOULD BE GREATER THAN 5. BUT IT DOES NOT GIVE ANY GUIDE LINES FOR PV%.
The measurement system is supposed to be able to detect variation in parts (PV). Most of the observed study variation should be contributed by PV. This ability is reduced by the presence of GRR error. The standard acceptance criteria (%GRR/TV) are based on ratios of standard deviations. You should also consider the Variance Contribution (ratio of squares of standard deviations: Adds up to 100%). The worst case 30% GRR is roughly equal to 10% contribution which means 90% contribution by PV. A 10% GRR would mean variance contribution of 1% by GRR error and 99% PV.
 
K

khowell

#14
We are preparing gage R&R for Percent Tolerance, and have converted from MSA v2 to MSA v3. I have specifically referenced this forum to our customer for the changes between the versions.

I have received back a question that when using the v3 K1, K2 & K3 constants as they are found in the book, doesn't that provide a 1-sigma result and you still need to multiply by 6 to get to 6-sigma.

This is one of the last things holding up their acceptance of our results.
Any clarification would be greatly appreciated!
:confused:
 
K

khowell

#18
Thank you! I am not quite sure of the results columns and which would be the 6-sigma %GRR that would typically be reviewed as the 0-10%, 20-30% and over 30%.
Thanks you,
-Kerry:confused:
 

Miner

Forum Moderator
Staff member
Admin
#19
I have attached my spreadsheet analysis below. Your worksheet is close, but the appraiser variation is slightly high. Nothing major.

You are correct that the revised k constants give a 1 sigma result, but the formulae where these are used also removed the 6 from the equations, so that the result is the same. This was done to simplify the equations.
 

Attachments

Status
Not open for further replies.
Thread starter Similar threads Forum Replies Date
S MSA on a process using different fixtures Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
C PPAP Rejection for "MSA-GR&R are not using the latest AIAG format" APQP and PPAP 9
M Using an earlier MSA on a different but similar part Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
J Performing a MSA on a new gauge using different parts Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
M MSA 4 - Results of GR&R Study using Range & Average Method Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
Q Using Traceable Standards - AIAG MSA Manual Chapter II, Section B, 1st bullet Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Miner New MSA file using ANOVA - January 2010 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 33
T Non-Replicable MSA when using digital multimeters or oscilloscopes Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
B MSA (Measurement Systems Analysis) with One Operator using a CMM or Common Calipers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Miner Intro to MSA of Continuous Data – Part 7: R&R using Wheeler’s Honest Gage Study Imported Legacy Blogs 17
C Attribute MSA Study using Kappa - Why Kappa value is low Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 13
V To formulate an Excel .xls worksheet for MSA on attribute data using Kappa Analysis Excel .xls Spreadsheet Templates and Tools 14
Nihls When the MSA results show no operator influence Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
S MSA for attribute relation gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Nihls MSA Study Type 1 (CMM) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
M MSA Study Type 1 not capable. We are at the limit. And manufacturing wants to continue producing. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
I Overwhelmed with attribute MSA requirement for visual inspection IATF 16949 - Automotive Quality Systems Standard 8
H Need MSA 4th ed. compliant attribute MSA template General Measurement Device and Calibration Topics 4
P MSA for titration Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
P MSA - what exactly mean "system" and master sample Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
P MSA study for visual system with artifical inteligence Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
D MSA strategy Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
K Updated MSA with the Annual PPAP validations APQP and PPAP 8
E MSA for Push and Pull test Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
6 MSA for Calculated Clearance Dimension Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
Ron Rompen MSA on automated measurement system - Multiple Step Vision System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
G MSA check list to audit IATF 16949 - Automotive Quality Systems Standard 10
R MSA for ATE (Automatic Test Equipment Embedded Software) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
Bev D MSA Tools Misc. Quality Assurance and Business Systems Related Topics 3
R MSA studies for Heat Treatment Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
C Industrial scales and MSA (IATF 16949 requirement 7.1.5.1.1) IATF 16949 - Automotive Quality Systems Standard 30
M Difference between MSA and MSE? General Measurement Device and Calibration Topics 1
M Test method validation - Is MSA (MSA1, MSA2, MSA3 and linearity) a good solution? Medical Device and FDA Regulations and Standards News 1
A MSA results differences - Supplier results vs. My results Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
R Preliminary MSA and Sockets' correlation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
I MSA requirement for 5 Micrometers + CP changes need customer approval? IATF 16949 - Automotive Quality Systems Standard 2
T Correct MSA study for an automated camera system which makes attribute inspection Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
Bev D MSA Misc. Quality Assurance and Business Systems Related Topics 46
Emran.mi Measurement system analysis - Can you help me about implementation MSA for CMM device Manufacturing and Related Processes 2
N MSA Study for a Leak Testing Device Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
N Justifications for not performing MSA (Measurement System Analysis) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
D Do I need part variation while doing Destructive Variable Gage R&R MSA study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
Prashant G MSA Study - AS 9100 and and our customer want us to do MSA study for their parts dimension Reliability Analysis - Predictions, Testing and Standards 4
D Compression Spring Force MSA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
E MSA Study on MTS dynamic rate measurements Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
A The FILE recommended by AIAG MSA 4ed for LINEARITY Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
S IATF 16949 Audit MSA IATF 16949 - Automotive Quality Systems Standard 10
C Do Calibration Reference Standards require an MSA study? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
A MSA When an Instrument Measures More than One Parameter Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
G MSA Type 1 - Cg and Cgk ranges Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1

Similar threads

Top Bottom