Multiple Site ISO 13485 Certification

T

TX Mfg

#1
Has anyone had experience in moving from multiple single-site certifications to a single multi-site certification with ISO 13485? My company is in the process of consolidating 10+ single certificates into 1 multiple-site certificate, but I need some feedback from others who have completed this process.

1. Did your NB require a re-audit of each site prior to adding to your multiple site certificate?

2. Do you certify your EU sales offices?

3. As all of our sites are currently certified, our plan is to allow them to continue to run their QMS systems as they are with the exception of requiring documented reporting to Corporate of internal audits, CAPAs, complaints, and mgmt rvw. We have corporate wide stds, which are general enough in nature to allow each site to tailor to their needs, but are we missing other important items?

4. Do each of your sites maintain their own quality manual or do you have an overall corporate quality manual covering all sites? We will have an overall corporate manual, but have not decided if each site will keep their own manuals.

Thank you for your feedback,

Greg
 
Elsmar Forum Sponsor

jkuil

Quite Involved in Discussions
#3
The general idea with multi-site certification is that all site are governed by the same quality system. How would a NB otherwise be able to provide a multi-site certificate without auditing all the sites. I can not tell from your question if your general corporate QS documents are specific enough in their requirements to meet the multi-site QS requirement. If each site has their own quality manual it becomes highly doubtfull that you will meet the single QS requirement.
The other requirement is that all sites are managed by the same general management, which you seem to have in place.
 
M

MIREGMGR

#4
The general idea with multi-site certification is that all site are governed by the same quality system. How would a NB otherwise be able to provide a multi-site certificate without auditing all the sites.
I didn't know there is such a thing as a multi-site certification for which some of the sites are not audited. How could the NB, or anyone relying on the certificate, be certain that all of the sites were compliant? Does this approach rely on a sampling concept?
 

Sidney Vianna

Post Responsibly
Staff member
Admin
#6
That document is almost obsolete, now.

See IAF MD 1:2007 Certification of Multiple Sites Based on Sampling
(Issue 1, version 2, issued on 20 November 2007; Application from 15 September 2008)
This document is mandatory for the consistent application of Clause 9.1.5. of ISO/IEC 17021:2006 and is based upon guidance previously provided in IAF GD2: 2005 Annex 3 and IAF GD6:2006, clause G.5.3.5 - G.5.3.13. All clauses of ISO/IEC 17021:2006 continue to apply and this document does not supersede any of the requirements in that standard. This mandatory document is not intended exclusively for Quality Management Systems (QMS) and Environmental Management Systems (EMS) and may be used for other management systems. However, relevant standards may provide specific requirements for multiple sites or preclude the use of sampling (eg. ISO/IEC 27006, ISO/TS 22003).

Keep in mind that certain sector schemes don't allow for sampling of sites. For example, the IATF does not allow sampling during TS-16949 certification audits, and the IAQG does not allow for sampling of AS/EN 9100 and 9110 audits.

An AB might prohibit sampling related to ISO 13485 certification audits.
 
M

MIREGMGR

#7
It'll be interesting to see how the FDA (which AFAIK does not utilize a sampling approach in either QSIT or for-cause inspection prioritization) will react to a request for recognition of an ISO 13485 audit report under the long-discussed and supposedly still-upcoming inspection-recognition program, if the establishment in question is multi-site and the NB utilized a sampling approach.
 
T

TX Mfg

#8
Thanks to everyone who has responded.

Let me clarify my question on multiple sites...Each of our sites currently gets audited every year based upon their own certificate, so we are simply trying to consolidate these many certificates into one single multiple-site certficate. Our NB stated they would need to re-audit each location prior to adding them to a multiple-site certificate even if they had already audited that site during the current year. Keep in mind, that no significant changes are being made at the sites as the QMS is already set up a global process (i.e. the system consists of standards required for each site in accordance with the requirements of 13485).

For example, a German facility gets audited say in January in support of their individual site certificate. In August, we want to add Germany to a new multiple site certificate and remove their individual cert. Our NB stated that they would need to return to Germany to cover this portion with that location before they could add them to the certificate even if no significant changes had occurred. This appears to be over-kill (or "over-charge") in my opinion.
 

Sidney Vianna

Post Responsibly
Staff member
Admin
#9
Our NB stated they would need to re-audit each location prior to adding them to a multiple-site certificate even if they had already audited that site during the current year.
Do you want to know the magic words? *

"I think we will have to put this out for competitive bidding, then...."

Back tracking is faster than count Dracula towards his casket, at the first sight of dawn....Try it. And give us feedback....
*Just make sure that you talk to someone high enough in the organization, and not just a sales person, whom might see this as a chance for another quote....In other words, make sure you talk to someone who would be concerned with the loss of your business.
 
Last edited:
R

Roland Cooke

#10
My basic advice on this is don't do it for the financial savings. There may be some, but they will generally be minor in the scheme of things. Do what makes most sense for your business, and if you save some money at the corporate level, that's a nice bonus.

I deal with a few companies that have a multi-site ISO13485 certificate. The main site is typically the Corporate HQ, that does regulatory, sales, marketing etc.
There are usually then one or two manufacturing facilities described on the certificate.

All sites have to be covered by a single top-level quality system, covering things like management review, document control etc, even if certain other individual processes, and the procedures to control those processes are very different from site to site.

The audit time at the manufacturing facilities will typically be reduced if there is sufficient overlap of procedures/processes to justify that.


In terms of sampling of sites, that's pretty rare. The only example I can think of is a company that has an HQ, one or two manufacturing facilities, and about 30 sales offices, all in the same country, all doing the same simple thing. We don't audit every sales office every year, I don't know the sampling regime.


I can't comment on the need to do re-audits of everywhere, that certainly sounds excessive, but I can think of some situations where that might be the case. You should certainly consider getting a more detailed explanation from your Notified Body as to why that might be required.
 
Thread starter Similar threads Forum Replies Date
M ISO 13485 Multiple Site Registration - Corporate HQ is listed as a subsidiary ISO 13485:2016 - Medical Device Quality Management Systems 4
V Site Master File for multiple business units within same premises/facility Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
T Multiple site certification vs Single site certification IATF 16949 - Automotive Quality Systems Standard 4
E Multiple Registrations at One Site - Operations were Separated - Four different QMS's ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 18
B Multiple site Certifications ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
V Software as medical device (SaMD) replicated for multiple clients through APIs IEC 62304 - Medical Device Software Life Cycle Processes 5
C Multiple owners for one listed device Medical Device and FDA Regulations and Standards News 1
A CE-IVD Technical File with Multiple IFUs EU Medical Device Regulations 4
K One CE Mark holder - multiple manufacturing sites EU Medical Device Regulations 3
K Multiple function device products EU Medical Device Regulations 0
O Are power distribution units Multiple Socket Outlets (MSOs)? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Ron Rompen MSA on automated measurement system - Multiple Step Vision System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
shimonv FDA News FDA guidance on Multiple Function Device Products (8/2020) Other US Medical Device Regulations 1
C Contract Review with Multiple Line items ISO 13485:2016 - Medical Device Quality Management Systems 7
I Cannot Export multiple sections from Excel to PDF without losing pagination Excel .xls Spreadsheet Templates and Tools 23
Q Legal Manufacturing Address Change – multiple registrations at same address 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
B QMS question in regards to multiple medical devices/products and N/A activities Other Medical Device Related Standards 12
Proud Liberal Power Query to combine multiple tables of unequal number of columns Statistical Analysis Tools, Techniques and SPC 1
D Reduction of software class based on multiple external risk controls IEC 62304 - Medical Device Software Life Cycle Processes 5
G Gauge R&R on multiple dimensions using 3D measurement system Capability, Accuracy and Stability - Processes, Machines, etc. 6
S Samples for GR&R - Can we get the 10 samples from multiple models Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
Proud Liberal Cp / Cpk on position using multiple MMC bonuses Capability, Accuracy and Stability - Processes, Machines, etc. 2
M Measuring Capability of Process with Multiple Specifications Capability, Accuracy and Stability - Processes, Machines, etc. 6
C FDA Registration Requirements for Multiple Sites 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
C Can a manufacturer (in South of Korea) have multiple Brazil Registration Holder (BRH) for same brend name? Other Medical Device Regulations World-Wide 1
S Multiple Authorized Reps in the EU for the same device? EU Medical Device Regulations 3
S Providing IFU (Instructions for Use) in multiple languages EU Medical Device Regulations 0
C FMEA - Multiple function failures considerations FMEA and Control Plans 3
B AS9100 Multiple Distributor warehousing locations under same Company AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 12
W SOP for Vigilance reporting covering multiple requirements Document Control Systems, Procedures, Forms and Templates 3
E Informational Internal Audits - Wear multiple hats what can and can't I audit (so I'm not auditing my own work) Internal Auditing 149
A Doubt on multiple inspection carrying out for same balloon number during Stage Inspection and Final Inspection? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
J DoE (Design of Experiments) - Multiple responses with different factors Using Minitab Software 2
L Control Plan - Multiple Process Steps FMEA and Control Plans 8
K Internal Auditing - Umbrella QMS and Multiple Standards Oil and Gas Industry Standards and Regulations 4
M Multiple missed POs grounds for SCAR? ISO 13485:2016 - Medical Device Quality Management Systems 32
G Single DFU for multiple medical devices in one box Other Medical Device Regulations World-Wide 0
J Brazil GMP license work with multiple BRH? Other Medical Device Regulations World-Wide 3
E Multiple socket-outlets - Medical Device Electrical Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
F Marketing a single medical device with multiple indications Other US Medical Device Regulations 4
M IATF 16949 - Multiple Locations - 3rd party audit scope IATF 16949 - Automotive Quality Systems Standard 1
Q IATF 16949 Scope - Producing multiple products IATF 16949 - Automotive Quality Systems Standard 5
S Language of procedures when operating in multiple countries ISO 13485:2016 - Medical Device Quality Management Systems 4
A Working for multiple CB's (Certification Bodies) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 17
K IEC 60601-1 Testing for Multiple Socket Outlet (MSO) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
H CE Marking on Multiple-Components Medical Products EU Medical Device Regulations 8
O How to deal with Multiple Datums - Position with respect to Multiple Datum Feature Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
D Implementing 5S for Multiple Shop Floor Layout Workstations Lean in Manufacturing and Service Industries 2
P Multiple Certifications - Is there an order of precedence? Other ISO and International Standards and European Regulations 3
M FDA UDI with Multiple Packaging Configurations Other US Medical Device Regulations 3

Similar threads

Top Bottom