ndc=0 and GRR=14% ??? Interpretation - basic question
ricardo.qre
Hi everyone, this is my first post:
Can anyone help me to interpret the following:
Spec: 6.887 - 6.913 mm (diameter)
Eq: Air Gauge
Eq. Resolution: .0005mm
GRR = 14%
ndc= 0
How can a process like this and its measuring system can be interpreted if GRR is between 10-30% that means the system variation is much lower than the part to part variation and could be approved, BUT ndc=1.41(PV/GRR) = 0 ??? which implicates the system variation is higher than the part to part variation?
PV=.00039
GRR=.00062
I´m confused, file is attached
Best regards and thanks everyone
bobdoering
Stop X-bar/R Madness!!
You are looking at GRR as a % of tolerance, whereas ndc is looking at GRR as a function of the variation you supplied to represent the variation of the process. If that variation is, for example, only 30% of the tolerance, you ndc will look bad. For the ndc to be accurate, the variation you supplied for the GRR must be representative of the variation you will see across the life of the process. That is a very important detail. Many times people only supply parts from one lot...and that generally does not represent the life of the part.
ricardo.qre
Thanks a lot. I´ve just taken 10 parts ramdonly that represents the total variation (3 operator/3 trials) and the total variation they show is extraodinarly small so the GRR es even less than 1% (this is actually the variation in the process I checked it our X-R charts and history data), but as you mention the process variation is so small that the ndc=0...looks bad
do I ignore this ndc = 0 value... what does this represent?
this time, the spec. is 84.412mm - 84.488mm
instrument values gives readings of .0001mm !! 10:1 rule is OK
bobdoering
Depends on what you are using the gage for. If it is pass/ fail, it is not a big deal. If it is used for controling the process - especially SPC, it is a big deal.
10:1 is rule of thumb. ndc is "statistical resolution" or the true resolution in usage when error is applied.
ricardo.qre
it is actually to control the process thru X-R, the Cpk=17
So, even the gauge resolution 10:1 is ok, and considering the variation is that small, can I get a final conclusion for this? Is the measurement system and its gauge (in variables) acceptable?
no need for controlling the process with X-R? right?
bobdoering
For X-bar R to be valid, to have enough resolution to see signals, you should have an ndc of at least 10.
It is acceptable to determin if the parts meet specification, and woefully inadequate for any use with charting. That is the bottom line!
That is a totally different question. You may need to have "report card' charts to meet customer requirements, there is a lot more to ponder before walking away from any benefits the charting may provide.
A lot depends on the use made of the gauge.
  • If the gauge is used for inspection, use the %Tolerance. It is 6.85%, which means the gauge is perfectly acceptable to inspect parts to tolerance.
  • If the gauge is used for process control, use %Study Variation and/or ndc. %SV is 95.95% and ndc = 0, which can mean one of two things. If the part variation represents the actual process variation, it means the gauge is not acceptable to control the process. Since your Cpk is > 9, it is probably same to assume that this is the situation. The second thing it could mean is that the part variation is less than the process variation and the %SV is meaningless.
  • If the gauge is used for both inspection and process control, use all three metrics.
  • %Contribution is used to focus your improvement efforts. 92.07% of the variation is caused by the gauge repeatability.
Okay, that was the official line per AIAG. You appear to be in a Wikipedia reference-linkCatch-22 (logic) situation. The better you make the process capability, the worse your gauge appears. At some point, you have to decide that loss in the ability to control your process is low risk and not worth the expense of new gauges.
supreecha
  • If the gauge is used for process control, use %Study Variation and/or ndc. %SV is 95.95% and ndc = 0, which can mean one of two things. If the part variation represents the actual process variation, it means the gauge is not acceptable to control the process. Since your Cpk is > 9, it is probably same to assume that this is the situation. The second thing it could mean is that the part variation is less than the process variation and the %SV is meaningless.%.
I agree with them.:applause:
Gage R&R for Air Gauge
Gage R&R Study - XBar/R Method
Gage name: Air Gauge
Date of study: Jan 13-2012
Reported by:
Tolerance: 6.887 - 6.913 mm (diameter)
Misc: Eq. Resolution: .0005mm
%Contribution
Source VarComp (of VarComp)
Total Gage R&R 0.0000004 71.56
Repeatability 0.0000002 39.76
Reproducibility 0.0000002 31.80
Part-To-Part 0.0000002 28.44
Total Variation 0.0000005 100.00
Process tolerance = 0.026
Study Var %Study Var %Tolerance
Source StdDev (SD) (6 * SD) (%SV) (SV/Toler)
Total Gage R&R 0.0006207 0.0037244 84.59 14.32
Repeatability 0.0004627 0.0027761 63.05 10.68
Reproducibility 0.0004138 0.0024828 56.39 9.55
Part-To-Part 0.0003913 0.0023480 53.33 9.03
Total Variation 0.0007338 0.0044028 100.00 16.93
Number of Distinct Categories = 1
Q Contingency plan - IATF Sanctioned Interpretation 17 IATF 16949 - Automotive Quality Systems Standard 4
U MDD Article 12/ MDr Article 22 interpretation EU Medical Device Regulations 2
R MDR, Annex I, 23.1 Interpretation - IFU on the website EU Medical Device Regulations 5
M MDR Article 22 Interpretation EU Medical Device Regulations 3
I Interpretation of the MDR GSPR 23.4 (u) EU Medical Device Regulations 2
D IATF16949 - Interpretation of Customer Requirements clauses IATF 16949 - Automotive Quality Systems Standard 3
A Interpretation of GMP Requirements for class 1 medical device manufacturer (device GMP exempt, only General controls applicable) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
P MDR Rule 10 interpretation - Active Device EU Medical Device Regulations 6
Q % Study variation low, % tolerance high - GR&R Interpretation help Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
A Interpretation with regards to Ppk > Cpk Capability, Accuracy and Stability - Processes, Machines, etc. 14
A OHSAS 18001 external auditor finding personal interpretation? Occupational Health & Safety Management Standards 5
A OEM branding - My interpretation of the LVD Directive makes us a manufacturer CE Marking (Conformité Européene) / CB Scheme 3
P Average Peel Strength - Interpretation of BS EN 868-5:2018 and ASTM F0088/F0088M Other Medical Device Related Standards 2
B ISO 50001 Interpretation of section 3.3.9 (Outsourcing) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
K Interpretation of significant change - material change Canada Medical Device Regulations 3
R MDR Software Rule 11 Formal Interpretation EU Medical Device Regulations 7
A IATF Sanctioned Interpretation No. 7 - Type and Extent of Control (supplemental) IATF 16949 - Automotive Quality Systems Standard 4
F Interpretation of MDR 2017/745 Article 23 - CE Mark Requirements EU Medical Device Regulations 8
U Hand-Held dosing device has no PATIENT - Interpretation of the PATIENT definition IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
G Medical Device "Immediate Container" Interpretation of Definition US Food and Drug Administration (FDA) 2
A Interpretation of Article 16 (2b) - Packaging, samples and Certificate EU Medical Device Regulations 10
D EU MDR Corrigendum Interpretation EU Medical Device Regulations 3
M Informational MDCG 2019-3 Interpretation of Article 54(2)b – Pre- market clinical evaluation consultation procedure with the involvement of expert panels Medical Device and FDA Regulations and Standards News 0
G EU MDR 2017/745 Rule 11 interpretation - Re-classification of a Software as Medical Device Other Medical Device Related Standards 0
JoshuaFroud Interpretation of Clause 5.5.2 in ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 2
qualprod P x I = Value interpretation for residual risk? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
T CSA Z299.3-85 Nuclear interpretation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
D VDA Trigger Matrix in the VDA 6.2 Manual - Interpretation and Use VDA Standards - Germany's Automotive Standards 3
B Interpretation of Customer Specific Requirements of Continental - Records Retention Customer and Company Specific Requirements 6
S Interpretation or Definition of ‘Once Every 5 Days’ ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
N What is the interpretation I-MR-R chart in this question? Statistical Analysis Tools, Techniques and SPC 1
E ME (Medical Equipment) Systems - IEC 60601-1 Clause 16.1 Interpretation IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
P IATF 16949 Cl. 7.1.5.2.1 "Calibration/Verification Records" Interpretation IATF 16949 - Automotive Quality Systems Standard 12
B Clarification on interpretation of some EN ISO 14971:2012 & IEC 62304:2006 req's ISO 14971 - Medical Device Risk Management 46
M IATF 16949 - 7.1.3.1 Plant, Facility, and Equipment Planning - Interpretation IATF 16949 - Automotive Quality Systems Standard 8
R Location Interpretation on Drawing - Hole to Hole? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
T TS 16949 Clause 7.4.1.2 and Sanctioned Interpretation IATF 16949 - Automotive Quality Systems Standard 12
D Interpretation of new IAQG ruling - Audit duration for nonconformance verification AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
K Interpretation of IEC 60601-1-2 Electromagnetic Compatibility – Requirements & tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Heat Treatment - Soaking time interpretation Manufacturing and Related Processes 5
D Interpretation of DOE interaction plot Using Minitab Software 8
A Training material for interpretation & understanding Part 11 requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
L Interpretation of "Any or All"'? IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
G Interpretation of phrase - "Direct Authority" EASA and JAA Aviation Standards and Requirements 5
pittmatj Brazil: Interpretation of RDC 185 Other Medical Device Regulations World-Wide 6
D What is your understanding or interpretation of TS16949 7.4.1.2 IATF 16949 - Automotive Quality Systems Standard 6
M Interpretation of Plus/Minus Draft - Injection Molded Parts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 10
E FDA Significant Change Interpretation - Guidance Document EU Medical Device Regulations 2
0 ISO 9001:2008 interpretation not the same per country? (Netherlands and France) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Multiple Standard Interpretation Queries Occupational Health & Safety Management Standards 4