I am in the process of developing a design verification document for HIV rapid test device, class III IVD. The general concept of this document is to provide a reasonable justification through testing that all the claims made in the design output file have been verified through lab testing and a good way of presenting this would be in a tabular matrix form.

I am confused regarding the scope/topics to be addressed in this document and how to go about with things. Where should I begin with and what are the sections which I should address in this documentation. for ex. justifications for the reagents and their concentrations used in the device, temperature/specificity/sensitivity claims made on the device, testing for temperature extremes for the device and so on.

I have tried reading through design control guidances but feel very lost in a heap of information at the moment. I will hugely appreciate any pointers that you guys have through experience that could assist me.

Thanks,
Aniket
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#2
Hi Aniket,

It would help clarify / focus if you told us what is the regulatory context of your query. If it's FDA oriented, it could also help if you specified the FDA ProCode.

Beside all that, I'd say that your company QS design Control SOPs would be a good place to start from.

Cheers,
Ronen.
 
R

redknight07

#3
Thank you for going through my query, Ronen. I am a graduate student who just stated interning at a medical device company making III IVDs (HIV Rapid test device). Current company acquired the technology from another company and does not have the design control documents in the proper format as required by FDA.

I recently compiled design input and output documents for the device using the existing documentation according to design control guidance document requirements and the templates provided on this forum and have to prepare the design verification document next.

In the input file I have included: performance requirements, functional requirements, product requirements and user interface requirements. In this section we have made claims such as the accuracy, sensitivity, specificity needs to be a certain %. there should be no cross reactions, the storage conditions/shelf life should be within a particular range and so on.

In the output document I have included: specifications of every raw material used (pH range, grade state of matter of the materials, pKa, impurities etc from the data I already present at the company), dimensions and specifications of the sub-assembly and final device including engineering drawings of all the parts, labeling, packaging and so on.

I understand that verification is the process by which the lab verifies that the established performance claims of an IVD test or product as mentioned in the input and output files can be replicated in the lab before patient testing. How do I tackle these issues when writing the verification file, what protocols to use, what format to use? Couldn't find any templates regarding this. I do know that it will need evidence and justification to every claim made in the input-output files.

Phew!!
 

Ronen E

Problem Solver
Staff member
Moderator
#4
Hi,

You have done some valuable work, but I suspect some of the fundamentals aren't there yet.

You didn't specifically said otherwise, so I'm going to assume that we are discussing all of this from an FDA perspective.

You really need to get the classification sorted first. Namely, identify the ProCode. This will start solidifying what you actually need to do / have.

Further, the fact that the technology is acquired / licensed doesn't provide any concessions. All applicable Design Control requirements have to be met, one way or another. Access to relevant information is typically a requirement in the acuisition agreements.

Before jumping on actually preparing the Design Control docs (input, output etc.), your company has to have it's relevant Quality System (QS) Standard Operating Procedures (SOP) in place. In general, it has to comply with 21 CFR 820. Those SOPs should tell you what to do, and how, including what reports are needed and what they should include.

Specifically, Design Verification is the process of making sure that Design Outputs are established, are accessible and properly address the Design Inputs (i.e. all pre-defined requirements are met). Verification is about demonstration; proof belongs to the Validation stage. This is true for all devices.

Cheers,
Ronen.

P.S. Perhaps you should start here, and follow through:

http://www.fda.gov/ MedicalDevices/DeviceRegulationandGuidance/HowtoMarketYourDevice/default.htm
 
R

redknight07

#5
Ronen, thank you for your assessment of my query. I am just about to graduate from University and am doing an internship at this company since September. I aspire to have a career in Regulatory Affairs and this is my first job as a regulatory professional.

You really need to get the classification sorted first. Namely, identify the ProCode. This will start solidifying what you actually need to do / have.

Further, the fact that the technology is acquired / licensed doesn't provide any concessions. All applicable Design Control requirements have to be met, one way or another. Access to relevant information is typically a requirement in the acuisition agreements.


The device in question is HIV Rapid test to detect HIV 1/2 antibodies using blood and serum/plasma samples. I was not able to find the product code for the device which is really strange. Never had such problems when had to look out for codes for devices. Under CLIA, since the device uses both blood and plasma, the complexity is moderate.


You didn't specifically said otherwise, so I'm going to assume that we are discussing all of this from an FDA perspective.

PMA submission is being considered for the coming year.


Before jumping on actually preparing the Design Control docs (input, output etc.), your company has to have it's relevant Quality System (QS) Standard Operating Procedures (SOP) in place. In general, it has to comply with 21 CFR 820. Those SOPs should tell you what to do, and how, including what reports are needed and what they should include.

The company does have a quality system in place which cover most sections of 820 but there are no guidance's on how to go about with design control documentation related to .


Specifically, Design Verification is the process of making sure that Design Outputs are established, are accessible and properly address the Design Inputs (i.e. all pre-defined requirements are met). Verification is about demonstration; proof belongs to the Validation stage. This is true for all devices.

I have tried my best to use all the information at hand regarding the device and compiled the input-output documents with the information that needs to be in these documents.

With verification, the approach I have taken here is to state every claim made in the input files and to provide evidence and justification for the same in the output and verification files.

A
 

yodon

Staff member
Super Moderator
#7
As Ronen noted, you have grasped the basic intent of the regulation.

But there's much more to V&V than just verifying your claims. For example, you need to do a hazard analysis / risk assessment. Controls to mitigate hazards need to be verified for implementation and effectiveness.

Without (hopefully) confusing matters for you too much, the 'V model' is a good approach (http://en.wikipedia.org/wiki/V-Model). Customer requirements / claims are validated and specifications (including risk controls) are verified.
 

Ronen E

Problem Solver
Staff member
Moderator
#8
As Ronen noted, you have grasped the basic intent of the regulation.

But there's much more to V&V than just verifying your claims. For example, you need to do a hazard analysis / risk assessment. Controls to mitigate hazards need to be verified for implementation and effectiveness.

Without (hopefully) confusing matters for you too much, the 'V model' is a good approach (http://en.wikipedia.org/wiki/V-Model). Customer requirements / claims are validated and specifications (including risk controls) are verified.
As much as I agree on a general level, FDA's requirements are not that clear / elaborate WRT risk management. In that context I would do 2 things to satisfy the FDA:

1. Include formal risk management in Design Validation.
2. Adopt a general approach of recurring risk management through various stages of device lifecycle.

The question of whether to run a full blown risk management session at Design Verification is - IMO - a question of benefit to resources ratio, and should be answered on a case by case basis. In my understanding Design Verification and Design Validation should be separate, clearly distinct activities, and since Verification is more about demonstration than about proof, I think risk management would add much more value at the Design Input and Design Validation stages.

Cheers,
Ronen.
 
R

redknight07

#9
Thank you Yodon and Ronen for your inputs.

I do understand that risk assessment needs to start from the design input stage and the controls used to mitigate need to be verified against implementation. At this moment I am planning to keep verification and validation as separate files.

Cheers,
Aniket
 
Thread starter Similar threads Forum Replies Date
N Defining Security Interfaces for Scope for ISMS - Need help IEC 27001 - Information Security Management Systems (ISMS) 10
S Need help in defining a World Class Certified Workcentre Manufacturing and Related Processes 3
E Defining Special Characteristics and Control Plan - Welding Operation - Need help Document Control Systems, Procedures, Forms and Templates 7
C ISO/ IEC 17021 Resource requirement (need help) Document Control Systems, Procedures, Forms and Templates 5
SocalSurfer AS9100 new certificate, but need QMS software, help Quality Assurance and Compliance Software Tools and Solutions 2
S Need help with analysing a survey on minitab Using Minitab Software 1
K A proposal for the model Quality Management - I need help for the project ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
R Probability - Need a help to solve the below question Statistical Analysis Tools, Techniques and SPC 5
8 Need Help - Runout - Function Gage Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
S New to FAIR, need help in filling it out AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
R Pls help --- Need expert advice on Video Measurement Measurement Uncertainty (MU) 0
Ajit Basrur Need help to understand HIPAA requirements ISO 13485:2016 - Medical Device Quality Management Systems 17
Z I need help with getting a CFS in China Manufacturing and Related Processes 2
S Need help wrapping my head around confidence vs beta error Statistical Analysis Tools, Techniques and SPC 2
P Received a minor for not having good measureables/goals. Need help with KPIs. IATF 16949 - Automotive Quality Systems Standard 52
A Failed the AS9120B Exam Twice.....Need Help AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
Ajit Basrur Need help in Conditional Formatting for the entire row Excel .xls Spreadsheet Templates and Tools 3
H Need help on Quality Improvement Plan (QIP) Quality Tools, Improvement and Analysis 5
S Help: Need Control Self Assessment Questionnaire for HR Departmental Functions Process Audits and Layered Process Audits 0
M Need help in Calibration of Directional Coupler General Measurement Device and Calibration Topics 2
L Need HELP with Internal Audit Program ISO 13485.2003 Quality Management System (QMS) Manuals 3
J Supporting Processes - Internal Audits - Need help settling a debate IATF 16949 - Automotive Quality Systems Standard 4
A Cpk Formula seems off, need help!? Capability, Accuracy and Stability - Processes, Machines, etc. 11
A Need Help With AEA Auditor Related Situational Questions AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
E Transformer Overload Test - Need help IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C I Need Help with Drafting Conventions!!! Design and Development of Products and Processes 4
M Need help on QMS Development in an AS9100C certified organization AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
R Need Help on Analysis: How to know potential causality model from historical data Problem Solving, Root Cause Fault and Failure Analysis 5
B Need help tracking training for my business Quality Assurance and Compliance Software Tools and Solutions 3
B Need help with a Control Plan for Encapsulation Mold / Trimming / Forming Tooling FMEA and Control Plans 9
S I need help from all of you bros Oil and Gas Industry Standards and Regulations 1
LRE67 Need help identifying this "thing" Coffee Break and Water Cooler Discussions 9
G Need help on Device Description in the Acceptance Checklist for Traditional 510(k) Other US Medical Device Regulations 2
K Need help finding a book on Documentation Book, Video, Blog and Web Site Reviews and Recommendations 8
J Need help/advice on video editing software etc..... Coffee Break and Water Cooler Discussions 8
L Need Help with Case Study Questions Manufacturing and Related Processes 16
W Help explaining the need for the Process Approach AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
Ron Rompen Unusual problem with Excel - Need some help Excel .xls Spreadsheet Templates and Tools 7
S Need help on new TS 16949 Rules IATF 16949 - Automotive Quality Systems Standard 5
M Need help with Network Analyzers and which DUT can be Calibrated with it Calibration and Metrology Software and Hardware 2
J Help I need your opinion on a News Letter Misc. Quality Assurance and Business Systems Related Topics 15
N Need Help Preparing for Medical Device Quality Engineer Interview Career and Occupation Discussions 2
B Need help on an application we are working on Coffee Break and Water Cooler Discussions 4
S Need help with Project Charter for my Six Sigma Green Belt Project Six Sigma 1
Q Need help on MSA study on a torque analyzer Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
B Need help in adding second location to FDA Registration Other US Medical Device Regulations 4
T A survey on problems during ISMS implementation - need help IEC 27001 - Information Security Management Systems (ISMS) 2
Z My first Gage R&R - Paper/Poly/Foil Material Bond Test, need help Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 15
R Need help on cost analysis for copper tube manufacturing Manufacturing and Related Processes 1
T SPSS for Analysis of Medical Data - Need help Statistical Analysis Tools, Techniques and SPC 12

Similar threads

Top Bottom