Need to convert Reference Only M&TE to a calibration schedule

Bluzman714

QMS Coordinator
Our company is in the process of becoming certified to ISO 9001:2015. In an internal audit we discovered we use a large number of "Reference Only" measuring and test equipment (M&TE) to inspect and approve final product. The M&TE were designated "Reference Only" by a previous QC Manager because they are sparsely used, and to reduce calibration costs. However, as I have learned, the general rule of thumb is to place M&TE used for inspecting and approving product on a calibration schedule to ensure their continued accuracy.

My questions are:
Do certain M&TE not require calibration/verification (e.g., pin gages, thread pitch gages)?

Is a 5 year interval too long for a device used an average of 2 or 3 times yearly? Is there an accepted standard for setting intervals?

What types of M&TE can be designated "Calibrate before use"?
 

howste

Thaumaturge
Trusted Information Resource
re: Need to convert Reference Only M&TE to a calibration schedule

My questions are:
Do certain M&TE not require calibration/verification (e.g., pin gages, thread pitch gages)?
No.

Is a 5 year interval too long for a device used an average of 2 or 3 times yearly? Is there an accepted standard for setting intervals?
Your company needs to understand and accept the risks. If at the end of a 5-year cycle you find that the equipment is out of tolerance, what would that mean in terms of nonconforming product shipped? You don't know when it went out. Is your company willing to deal with the consequences?

What types of M&TE can be designated "Calibrate before use"?
As far as I know, any equipment can be designated as calibrate before use. If you're not using the equipment then there's no risk of a bad measurement. Just make sure that you have the ability to calibrate/verify it in time so it doesn't impact schedules.
 
re: Need to convert Reference Only M&TE to a calibration schedule

We have a large amount of RO measurement tools on the floor, BUT, final inspection uses calibrated tools only, this is how we reduce calibration costs. Some is marked "CBU" but we have the ability to calibrate it as needed, as it is rarely used.
 

BradM

Leader
Admin
re: Need to convert Reference Only M&TE to a calibration schedule

To add to the other good responses.. the problem is. you don't have any history on these instruments. You don't know how they will perform.

Many times (not the norm) companies have a lot of instruments; many times more than their current processes dictate. I would first assess what you really need and which ones are going to be critical; need to be calibrated.

Then, I would take a sample of those and send for calibration. Before you spend too much time and money in enrolling them in a program and such, make sure they will pass calibration.

Like Howste stated, it really depends on a lot of factors as to how often you should calibrate. For starters, how often they are used, your risk, the instruments themselves. Saying... I personally set a much longer interval on a Fluke multimeter than say... an ABC cheapy meter I know nothing about. :)

So first, I would assess what equipment requires calibration, assume an annual calibration, determine annual cost and inform management as a start.
 
Thread starter Similar threads Forum Replies Date
K Need Help With IATF 16949 Audit NC IATF 16949 - Automotive Quality Systems Standard 6
F Defining the need to perform the Proximity magnetic fields immunity tests (IEC 61000-4-39) according to the IEC 60601-1-2 ed.4.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D FDA combo product-need to define direct & indirect impact regarding safety,quality & regulatory.Can anyone point me in right direction-SupplierMgmt 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
C Does a Pass or Fail IP test need UoM on the report? Measurement Uncertainty (MU) 0
P Do Barcode scanners need validation in Medical device manufacturing? Manufacturing and Related Processes 6
S ISO 9001: 2015 MR Outputs - "any need for changes to the quality management system' Management Review Meetings and related Processes 13
S Need guidance for Software validation Qualification and Validation (including 21 CFR Part 11) 5
R Do I need to get calibration certificate from ISO 17025 for IATF Audit IATF 16949 - Automotive Quality Systems Standard 8
H Electronic Replacement parts need CE? CE Marking (Conformité Européene) / CB Scheme 0
G Need for update of Declaration of Conformity EU Medical Device Regulations 4
D Do I need MDSAP? ISO 13485:2016 - Medical Device Quality Management Systems 5
bryan willemot Wanted: in need of start up Lock out Tag out program in Excel or Word, your help is much appreciated Occupational Health & Safety Management Standards 3
Q Does a gage used for a visual verification need to be calibrated? Manufacturing and Related Processes 7
B Hi , everyone i need a procedure for validation of design prototype api 6d (valve manufacturing) Oil and Gas Industry Standards and Regulations 1
J Do we need to raise concessions for every noted deviation under AS9100 Manufacturing and Related Processes 3
M Need to set up a "crisis management and business continuity plan" AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
D MDR technical file: does it need to contain duplicates of controlled documents EU Medical Device Regulations 2
R Customer Returns / RMA's & the need for NCR's AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
T Need advice on what inspection equipment to buy. General Measurement Device and Calibration Topics 2
Sam.F Do heat treatment ovens need calibration? General Measurement Device and Calibration Topics 25
T Need 5 why’s help AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 15
H Need guidance on importation of Custom-made devices into Japan Japan Medical Device Regulations 1
T Need help understanding AS6174 clause 3.1.5. AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
B Spare parts of medical device need CE mark? EU Medical Device Regulations 0
F Change in address for critical supplier - does NB need to be informed? CE Marking (Conformité Européene) / CB Scheme 20
leftoverture REACH Folks - We Need Common Sense! RoHS, REACH, ELV, IMDS and Restricted Substances 23
B Whose waste carrier licence do I need? ISO 14001:2015 Specific Discussions 8
I I need some fun examples of teaching the 5 Whys Lean in Manufacturing and Service Industries 43
R First Time Managing Calibration and Measurement System. Need Help. General Measurement Device and Calibration Topics 19
M CE marked medical device - do they need UKAS certified calibration provider ISO 17025 related Discussions 4
C I need some clearing on requirements of Asuring the validity of results ISO 17025 related Discussions 2
C Need help in determining applicable clause for an audit finding (based on AS9120B) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
K Subcontractors Providing Services Under MDD or MDR need ISO 13485 from EU Notified Body? CE Marking (Conformité Européene) / CB Scheme 8
E Need some peer advice Quality Manager and Management Related Issues 18
S Need clarification on UKCA and "Approved Body" UK Medical Device Regulations 7
S What types of changes need to be notified to NBs? Registrars and Notified Bodies 1
B Do I need a Cleanroom to manufacture Medical Devices? ISO 13485:2016 - Medical Device Quality Management Systems 2
J CMO in Canada do they need MDSAP/13485 certification? ISO 13485:2016 - Medical Device Quality Management Systems 1
M Attempting to Poke Yoke a wire crimping process...need help Design and Development of Products and Processes 3
C I need your help to find the action plan for these quality objectives Misc. Quality Assurance and Business Systems Related Topics 23
ChrisM Class 1 non-sterile device, need certified QMS or not? EU Medical Device Regulations 2
Steve Prevette I need a template to solve world hunger, a world class Quality Program, and make lots of money Coffee Break and Water Cooler Discussions 5
C Need to understand difference between an actual Calibration vs. a Calibration check General Measurement Device and Calibration Topics 3
S How many tester quantity we need on the line based on the cycle time and peak volume Manufacturing and Related Processes 3
R Need help on calibration result analysis Measurement Uncertainty (MU) 17
M Do i need to have equipment validation if 100% testing is completed? Qualification and Validation (including 21 CFR Part 11) 6
G Need Help with Run @ Rate for Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 0
F Need help in IMS management review Management Review Meetings and related Processes 3
S I need help in corrective action plan Nonconformance and Corrective Action 15
A Do clinical performance studies for IVDs need to be conducted in a member state? EU Medical Device Regulations 2

Similar threads

Top Bottom