Never put an Expiration (Due) Date on a Calibration Certificate
#31
I was really expecting to see a funny face after that question, Jim.
Or, wait, this is better....:sarcasm:
I'll bite anyway....
Yes, we refer to items that are in our lab to be calibrated as artifacts. That is, in written procedures and statements such as the one above. Not in daily conversations, of course... We're not that crazy.:bonk:
Sarcasm? Moi??? I've just never seen the word used that way. Carry on.
Anerol C
#33
Good morning,
What I can do if we bought several mesuring devices to a distribuitor and the calibration certificates are expired, the devices are brand new, Do I need to create a letter stating the date that we started using the devices and use date date as our calibration date; Do I need to include purchar order as evidence/back up in order to justify why using a differerent date vs certificate date? Please advise, I don't know whant is the best way to deal with this situation.
Thanks for your help.
Do you have such type of letter to share?
AC
Coury Ferguson
Moderator here to help
#34
Good morning,
AC
What I would do in this situation, is: Send them out for Calibration.
One thing I have learned is not to always accept the OEM's certificate of Calibration.
Jerry Eldred
#35
There are times when you can get away with using a delayed start date for a cal interval. However, generally, in an audit, you would have the burden of proof to demonstrate to the auditor why during the intervening time the accuracy would not have degraded in the instrument.
Items which change value strictly due to use versus passage of time, that may be acceptable. In a previous employer, we used high temperature thermocouples which have a life span of about 6 months when used at around 1000 to 1200 Degrees Celsius. However, that aging was due to the high temperature. So we had a document in our quality system stating that our procedure was to place them in environmentally controlled storage. Then, when they were given to the user, the calibration interval began. This could have been true for items which only age/drift/change during use, and definitely do not change in controlled storage.
But for everything else, the cal date is when it was actually calibrated.
My recommendation would be as in the previous post: to either live with the existing cal date/due date, or get them re certified.
As to the validity of the OEM certificate, some are good and some are not. I am currently in a high volume lab with pretty strict guidelines in that regard. We review every one of the OEM certificates on new instruments. Quite often they are fine (from the large, reputable OEMS (such as Fluke, Tektronix, Agilent, Mitutoyo, etc.). But some are not. I would weigh the validity of the particular certificate as to whether to accept it.
Sorry if any of this is bad news. Just have to shoot as straight as I can, and try to help.
Wes Bucey
#36
Good morning,
AC
I'm not exactly clear about what timetable events followed.
  1. Were the certificates expired when you obtained the devices? (If yes, you should have gotten a BIG discount, which would cover cost of updating calibration) - many folks have operations where devices are for "intermediate measuring activity" and only the FINAL devices are kept calibrated - not exactly best practice, but often acceptable, so they are natural market for discounted, out-of-date devices.
  2. If the devices were NOT out-of-date when you acquired them, it is your organization's responsibility to keep them calibrated with up-to-date certificates if your own QMS (Quality Management System) requires calibrated devices.
If these devices are used to approve product, they probably should have current certificates of calibration. If they have NEVER been used, odds are they can merely be rechecked by the original certificate signer and updated, without requiring costly adjustments or refurbishing to bring into calibration - price "could" be much more reasonable than going to a new entity to obtain a certificate.
Good luck! Tell us what you ultimately do and whether the pricing was fair or exorbitant.
Bob the QE
#37
I strongly agree with Wes's first point. Whether you believe the cert to be accurate or not (even if it up to date). That cert is part of your purchase and thus you have paid for it. If you went to buy another type of equipment or device wouldn't you expect it to be in working condition for its intended use? Of course you would, so the same would hold true in this case. I would guess the use of this equipment is verification and thus calibration is a major part of that. I would also ask your distributor what their FIFO policy is. I don't know that this would be a concern but the answer would be interesting.
The points on your organization are accountable to audit criteria regardless of the root cause also appear to be strong.
Good Luck
Hershal
Metrologist-Auditor
#38
One thing that has been mentioned (thanks AndyN) is accreditation under ISO/IEC 17025.....if a cal provider is not accredited or is ISO 9K, there is no real oversight with respect to the due date issue. In fact, if the lab is only ANSI/NCSL Z540-1-1994 which is self-declared, that standard requires the due date be provided.....otherwise the customer (user of the instrument) has to monitor and control this.....in my experience there are few Metrology professionals at end user organizations and especially at smaller organizations, hence there may be insufficient knowledge to interface with the cal provider.....remember, Quality and Metrology are related but are not the same!
Under ISO/IEC 17025, Clause 5.10.4, due dates cannot be put on without prior agreement by the customer, and the ABs oversee this.
Hope this helps clarify what is or is not allowed.
Wayne
Gage Crib Worldwide
#39
What I can do if we bought several measuring devices to a distributor and the calibration certificates are expired, the devices are brand new, Do I need to create a letter stating the date that we started using the devices and use date as our calibration date; Do I need to include purchase order as evidence/back up in order to justify why using a different date vs certificate date? Please advise, I don't know what is the best way to deal with this situation.
Anerol C; I would point you back to the first post that started this thread:
... the basic fact that a certification is only valid at the moment the instrument is being checked. By the time the certification has been typed the instrument may already be out of calibration....
In this specific situation, you know by virtue of the provided calibration certificate, that it has been some time since the tools were manufactured. It is for you to determine if they deteriorate with the passage of time. You also know that the tools are “brand new”; that they are in original packaging and appeared to no signs of use or abuse.
It is up to you, as the local Quality Professional to determine if the tools need calibration. If you decide not to have the tool checked, it would be best to attach to the certificate a note describing your decision process, with any appropriate documentation, because you may need to substantiate the logic behind your decision.
Basically I would say: When in doubt; send them out.
What I would do in this situation, is: Send them out for Calibration. One thing I have learned is not to always accept the OEM's certificate of Calibration.
Again, remembering the first post, calibration expiration dates are worth little. Whenever you purchase a new tool, regardless of the supplied calibration certificate, you should evaluate the tool to determine if local verification is required.
... you should have gotten a BIG discount, which would cover cost of updating calibration ...
... That cert is part of your purchase and thus you have paid for it.... I would guess the use of this equipment is verification and thus calibration is a major part of that.
As long as the type of tool is such that it does not deteriorate with age, and knowing that any certification is only really valid when the tool was verified, regardless of any applied “expiration date”, sending a new tool to your approved laboratory upon arrival is just prudent. The initial calibration of the new tool will give you a base-line calibration, and the cost of calibration should not that BIG of an issue. It will be an expense that you will incur several times during the life of the tool.
Anerol C
#40
Yes some of the certificates were expired when we received the equipment.
What I think I will do is get a deviation to use them while we can send them out to calibration, the issue is that some of them are the only equipment in the whole plant, so I think we need to get a new one then send the equipment with expired certificate to calibration. I think this is the only thing that I can do if we are going to have the registrar coming to our plant in about one month.
AC
supadrai Possible to Submit a 510(k) if you have never, do not and cannot produce the device for which you're submitting the application? US Food and Drug Administration (FDA) 4
hockeyhead Existing process never included in AS9100 scope of certification AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
Jim Wynne Some things never change Coffee Break and Water Cooler Discussions 2
N Variable Data Test Never Fails Capability, Accuracy and Stability - Processes, Machines, etc. 9
K 510K statement Questions 21 CFR 807.93 - Discontinued or never released for sale US Food and Drug Administration (FDA) 0
P Retired equipment that was never validated Qualification and Validation (including 21 CFR Part 11) 2
M Is calibration from long ago still valid on never used pipettes? General Measurement Device and Calibration Topics 8
A Receiving paper from the customer and never knowing when it will arrive Quality Manager and Management Related Issues 1
J Validating a stopped test due to never finding any defects Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
R KPI Review Meetings never discusses previous meeting action plan status Quality Manager and Management Related Issues 5
G 10 Things You Should Never Say at the Office Coffee Break and Water Cooler Discussions 7
W How to withdraw a cleared (but never sold in the US) 510(k) Medical Device? Other US Medical Device Regulations 7
Wes Bucey Never give up! Career and Occupation Discussions 1
Marc Lascaux Cave Paintings - Rare And Never-Seen Photos Coffee Break and Water Cooler Discussions 4
sagai Why never to ask favours from the designers Funny Stuff - Jokes and Humour 8
Marc "Prompt Corrective Action" is never prompt and rarely truly corrective Nonconformance and Corrective Action 48
S Why should we never re-use an identification number for a replacement M&TE? General Measurement Device and Calibration Topics 7
Jim Wynne Why You Should Never Depend on Babel Fish for Language Translations Funny Stuff - Jokes and Humour 6
SteelMaiden Some people never learn. Coffee Break and Water Cooler Discussions 4
bio_subbu Never happen in life again??!!!! Coffee Break and Water Cooler Discussions 4
Marc Security Firms Fined Over Never-Ending Subscriptions After Work and Weekend Discussion Topics 0
K Never been this close to a plane crash... Coffee Break and Water Cooler Discussions 38
GStough REACH and ISO 9001:2000/13485:2003 - Never the Twain Shall Meet? RoHS, REACH, ELV, IMDS and Restricted Substances 4
J What to do if equipment IQ and OQ were never signed off ? Nonconformance and Corrective Action 4
Jim Wynne Never believe early news reports Coffee Break and Water Cooler Discussions 7
B Retiring Early or Never? Coffee Break and Water Cooler Discussions 65
Marc Never Give Up.... Funny Stuff - Jokes and Humour 6
C Acceptance sampling should never be used for processes in statistical control Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
M Never though ordering breakfast was so difficult! Funny Stuff - Jokes and Humour 0
D Unique user ID under ISO 27001 - ID can never ever be reissued to another user? Other ISO and International Standards and European Regulations 7
R LOST...the TV show - Why have we never discussed "LOST"? Coffee Break and Water Cooler Discussions 23
A Never question a drunk Funny Stuff - Jokes and Humour 4
I How many fail ALL ISO 9001 registration audits - How many never make it? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 46
C Are Purchase Orders required for 7.4? Company never issues POs and never has. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
I Records that never were a document...? Shipping Labels and Computer 'Screens'? Records and Data - Quality, Legal and Other Evidence 7
B Logistics FMEA - Never Have done One - Need Ideas FMEA and Control Plans 12
S Approval For Standard OEM Parts - I am never quite sure how we should submit approval Customer and Company Specific Requirements 3
D 6.2.2 e - Minimum education requirements for workers? Some never went to school! ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
R When can you stop root cause analysis - Is the answer never? Problem Solving, Root Cause Fault and Failure Analysis 6
Marc ITS - Intertek - Fraud? NEVER!!! Philosophy, Gurus, Innovation and Evolution 4
A Can a US manufacturer put export only products on a CFG? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M Hiring Decisions - How much effort do you put into the hiring process these days? (7/2020) Misc. Quality Assurance and Business Systems Related Topics 6
C Is it required to put"Rx only" on the home page of an app? Medical Device and FDA Regulations and Standards News 4
V What plans have been put in place to achieve quality objectives? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
G What information to put on measurement Dimensional Results APQP and PPAP 7
bryan willemot Looking for a good CoC to put with my certification sheets when we send parts Document Control Systems, Procedures, Forms and Templates 1
Z Do we still have to put UDI if we have DI and lot number seperately on our product ? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
D QMS Documentation Process - I want to put all the Process SOPs to a QMS system Document Control Systems, Procedures, Forms and Templates 3
C What are rules to notify customers if IATF cert is put on hold? IATF 16949 - Automotive Quality Systems Standard 2
C IATF 16949 : 2016 - How put in place a Containment for this kind of Major NC IATF 16949 - Automotive Quality Systems Standard 14