New Draft FDA Guidance to distinguish b/w Recall and Product Enhancement & Reporting
Chrisx
Re: New Draft FDA Guidance to distinguish b/w Recall and Product Enhancement & Report
It is an interesting guidance, but I think it is lacking when addressing reusable instruments in the orthopedic industry. Orthopedic devices are often sold in "loaner" cases that are returned to the manufacturer for replenishment of the consumed implants. Often, I would like to replace the instruments with ones that have undergone a product enhancement or simply have excessive wear. Is the replacement of components of these returned cases considered a recall? They are afterall in commercial distribution. Is replacing a worn bone screw driver a reduction in risk? I think yes, a worn driver has an increased risk of slipping in the screw head. According to this guidance FDA would classify this as a reportable recall, because I am reducing risk by replacing a worn instrument.
Often, I am replacing instruments that have a spare in the case. The reality is instruments do break. Surgeons hit them with excessive force with hammers that are larger then the ones provided in our case or otherwise misuse the instrument. For this reason, we always have spare instruments in the case. If I replace the instrument with a product enhacement to reduce the likelilhood of breakage, is this a reduction in risk? I'm not really sure. There is always a spare instrument in the case, so the risk is minimal. The procedure can be completed with the spare instrument. Does anyone else have any experience in managing this situation?
Thanks for any input
MIREGMGR
The above-mentioned Guidance is a draft. You might want to call or email Ron Brown and initiate a dialogue so that your concerns might be addressed in the final Guidance.
pkost
It is an interesting guidance, but I think it is lacking when addressing reusable instruments in the orthopedic industry. <snip>
I think the guidance addresses these issues quite well and with a quick read, to my satisfaction. It notes that if the level of safety performance was achieved, cleared and approved then the device is not in violation. (line 225-230)
Therefore, if you know that it is going to undergo wear and tear (which you should do) and that it as a service of life of x number of uses, then it is not a recall if you do not have to replace before x number of uses.
Chrisx
Certainly, the guidance can be applied to my products. I just think it could be a little clearer. Maybe an example in the guidance using orthopedic "loaner" cases. It is afer all found throughout the orthopedic industry and I am sure many people have the same question.
Unfortunately, answering all the "ifs" is not so simple. In many cases the instrument failure is not related to wear, but rather to excessive force applied by the surgeon. In this case, the number of uses has no bearing on whether the device failure makes the device violative.
Even wear is not an easy question to answer. We have many bone screw drivers that are being replaced with titanium nitride coated tips to reduce wear and extend life. These products are class I and therefore were never cleared and not subject to 510(k). It does not seem that such an activity should be construed as a recall, but the guidance would seem to suggest that it is a reportable recall.
What I probably need to do is right an HHE for each product enhancement documenting that the enhancement does not reduce risk because there is a spare instrument readily available to the surgeon. I would be curious to see if other people think this is a defensible argument.
pkost
Both of these points should have been considered and mitigated as best as possible in the inital design. The residual risk should the have been accepted.
When I developed orthopaedic instruments I always worked on the following premise....if a surgeon can hit it with a mallet, he/she will hit it with a mallet. Excess force was factored into our risk assessment and mitigated whereever possible. It is therefore known and predicted - therefore only an enhancement and not a recall.
The same can be said for wear and tear
Chrisx
Of courese, if everything had been done right the first time I wouldn't have to consider whether these are recalls or product enhancements. Unfortunately, this is what I inherited. To be sure, there is much improvement that is needed in risk management. Thanks for your advice.:thanx:
MIREGMGR
Certainly such actions on your part are removals or corrections, depending on circumstances. Have you discussed with your District's Recall Coordinator whether they should be considered recalls?
I don't think FDA is particularly eager to clog up their system by categorizing normal activity (replacement of not-reasonably-preventable breakage or wear-and-tear, or voluntary product upgrades) as Recalls.
Chrisx
I think it is not so simple. I do not think one can assume that because a reusuable instrument is replaced with a new design to address breakage, this automatically is a recall. I think if this were the case, there would be many more recalls in the orthopedic industry.
In some cases, product enhancements were made on the basis of a single instrument breakage. The complaint history over two years only has one repeat incident of breakage. We have not classified either breakage as an MDR. I think it is hard for me to envision this as a recall. I think they are at an acceptable level of risk, but this should be supported by the risk management file. For now, the best I can do is conduct an HHE.
The previous poster is correct. Our risk management file should spell out the acceptable risk for these events. This would support our decision that our actions are not recalls. To be sure, there is a lot of improvement needed in our risk management files.
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Draft Guidance – Breast Implants – Certain Labeling Recommendations to Improve Patient Communication Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Draft Guidance – Clinical Decision Support Software Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Draft Guidance – Providing Regulatory Submissions for Medical Devices in Electronic Format – Submissions Under Section 745A(b) of the Federal F Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Draft Guidance – Patient Engagement in Design and Conduct of Medical Device Clinical Investigations Medical Device and FDA Regulations and Standards News 0
M Informational FDA draft guidance – Safer Technologies Program for Medical Devices Medical Device and FDA Regulations and Standards News 0
M Informational US FDA draft guidance – Testing and Labeling Medical Devices for Safety in the Magnetic Resonance (MR) Environment Medical Device and FDA Regulations and Standards News 0
M Informational US FDA draft guidance. – Clinical Investigations for Prostate Tissue Ablation Devices Medical Device and FDA Regulations and Standards News 2
M Informational US FDA draft guidance – Mouse Embryo Assay for Assisted Reproduction Technology Devices Medical Device and FDA Regulations and Standards News 0
M Informational USFDA draft guidance – Nonbinding Feedback After Certain FDA Inspections of Device Establishments Medical Device and FDA Regulations and Standards News 0
Ronen E FDA releases new draft guidance on when to submit a 510(k) for device modification Other US Medical Device Regulations 10
Ronen E FDA issues draft guidance on form & content of UDI Other US Medical Device Regulations 0
Ronen E FDA issues draft guidance on additive manufacturing Other US Medical Device Regulations 1
M New FDA Draft Guidance on 510(k) Ownership and Use Other US Medical Device Regulations 3
R New Draft Guidance Document for FDA's current thinking on De Novo submission Other US Medical Device Regulations 1
S FDA Draft Guidance on HLA Kits 510(k) Submission 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
Y New Draft FDA Guidance for Applicability of GLP in Medical Device Submission Other US Medical Device Regulations 2
M FDA issues draft guidance to replace G95-1 (Application of ISO 10993) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
bio_subbu US FDA issues Draft Guidance on Premarket Approval (PMA) Acceptability Other US Medical Device Regulations 0
Ronen E New FDA Draft Guidance on Enteral Connectors Misconnection considerations Other US Medical Device Regulations 0
Ronen E FDA publishes Draft Guidance on Med Dev Pre-Submission Program and Meetings with FDA Other US Medical Device Regulations 0
Ronen E FDA distributes draft guidance related to X-ray Imaging Device 510(k)s Other US Medical Device Regulations 1
Ajit Basrur FDA issues draft guidance on nanotechnology US Food and Drug Administration (FDA) 1
bio_subbu FDA releases draft guidance on Providing Regulatory Submissions in Electronic Format Other US Medical Device Regulations 1
bio_subbu FDA issues draft PMA (Pre-Market Approval) Clinical Study Guidance 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
sagai New Draft Guidance from FDA - Factors considered for Risk/Benefit Determination Other US Medical Device Regulations 4
S FDA Draft Guidance on Mobile Medical Applications US Food and Drug Administration (FDA) 1
bio_subbu FDA issues first draft guidance document on Mobile Medical Applications 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
S New draft FDA Classification Guidance US Food and Drug Administration (FDA) 1
S New Draft FDA Guidance - Human Factors & Usability Engineering for Med Device Design US Food and Drug Administration (FDA) 0
S New Draft FDA Guidance on Enforcement of INDs - Safety Reporting Requirements US Food and Drug Administration (FDA) 0
S New US FDA Draft Guidance on Final Disclosure by Clinical Investigators US Food and Drug Administration (FDA) 0
Y Draft FDA Guidance on "Presenting Risk Information in Prescription Drug and Device" US Food and Drug Administration (FDA) 1
Ajit Basrur FDA releases new Draft Guidance - Process Validation - Gen Principles and Practices Qualification and Validation (including 21 CFR Part 11) 27
D Upcoming FDA ISO 13485 audit guidance (draft version) ISO 13485:2016 - Medical Device Quality Management Systems 19
M Informational From Medtech Insight – QSR/ISO 13485 Harmonization Update: FDA Enforcement Discretion Likely When New Rule Stands Up; Draft Reg Coming By Year’s End; Medical Device and FDA Regulations and Standards News 0
M Informational Several US FDA draft guidances, including some specific device guidances for the Safety and Performance Based Pathway Medical Device and FDA Regulations and Standards News 0
P Understanding FDA draft "Management of Cybersecurity in Medical Devices" Medical Information Technology, Medical Software and Health Informatics 3
M Medical Device News FDA Releases Draft Recommendations on Premarket Submissions for Management of Cybersecurity in Medical Devices Other US Medical Device Regulations 0
M Medical Device News FDA news - 05-09-18 - Draft - Uncertainty in Benefit-Risk Determinations Other US Medical Device Regulations 0
T Draft of New UK Regulations? UK Medical Device Regulations 2
C FMEA Process assessment In the Draft for the AIAG/VDA FMEA Manual is gone FMEA and Control Plans 0
M Informational IMDRF draft document – Principles and Practices for Medical Device Cybersecurity Medical Device and FDA Regulations and Standards News 0
M Informational IMDRF draft document – Requirements for Regulatory Authority Recognition of Conformity Assessment Bodies Conducting Medical Device Regulatory Reviews Medical Device and FDA Regulations and Standards News 0
M Informational EU draft act – Single-use medical devices – safety and performance requirements for reprocessing Medical Device and FDA Regulations and Standards News 0
M Informational Draft ANSM’S GUIDELINE Cybersecurity of medical devices integrating software during their life cycle Medical Device and FDA Regulations and Standards News 0
M Informational Draft standardisation request as regards medical devices in support of Regulation (EU) 2017/745 and in vitro diagnostic medical devices in support of Medical Device and FDA Regulations and Standards News 2
L Getting Notice of Revision and Accessing Public Draft Copies of ISO Standards Service Industry Specific Topics 2
M Informational EMA – Consultation on draft guideline on quality requirements for medical devices in combination products Medical Device and FDA Regulations and Standards News 0
M Informational USFDA draft guidance – Technical Performance Assessment of Quantitative Imaging in Device Premarket Submissions Medical Device and FDA Regulations and Standards News 0