New EMC IEC60601-1-2 4.0 Ed. Requirements

N

nikolaos

Hi All,

I try to understand the IEC60601-1-2:2014 4.0 Ed. new requirements and I have started a gap analysis for some of our ME.

According IEC60601-1-2:2014:
Annex I ? immunity PASS/FAIL criteria can specify degradations that are acceptable because they do not result in unacceptable RISK? and also ?the ME passes the test if it continues to provide its BASIC SAFETY and ESSENTIAL PERFORMANCE?

According the Risk Management of our ME there are NO ESSENTIAL PERFORMANCES of a clinical function, other that related to BASIC SAFETY, where loss or degradation beyond the limits specified results in an unacceptable RISK.

Does this mean that whatever happens during IMMUNITY test, the test is PASS if the ME continues to provide BASIC SAFETY?

Checking with the 2014/30/EU EMC Directive:
?When placing their apparatus on the market, manufacturers shall ensure that they have been designed and manufactured in accordance with the essential requirements set out in Annex I. :

Equipment shall be so designed and manufactured, having regard to the state of the art, as to ensure that:


(b)it has a level of immunity to the electromagnetic disturbance to be expected in its intended use which allows it to operate without unacceptable degradation of its intended use.

These are the FUNCTIONAL PERFORMANCES but the IEC60601-1-2:2014 4.0 Ed. refers only to the ESSENTIAL PERFORMANCES which is quite different.


Example:

If during the ESD test (16KV) the display becomes black or gets broken, the test is PASS for IEC60601-1-2:2014 if the ME continues to provide its BASIC SAFETY but FAIL according the 2014/30/EU EMC Directive.

What is missing me? What is the correct interpretation?


Thanks in advance
 

Stijloor

Leader
Super Moderator
Re: EMC IEC60601-1-2 4.0 Ed. New Requirements

A Quick Bump!

Can someone help?

Thank you very much!!
 
N

nikolaos

Re: EMC IEC60601-1-2 4.0 Ed. New Requirements

No help or any comments? The 4.0 Ed. is very recent and there is no much experiance with test labs.
Thanks
 
W

wally4u

I'm somewhat hesitant to classify any medical device with "no essential performance", but if your NB accepts that (ours doesn't) great :eek:!

The way I would read the MDD section would that if you can reasonably expect any EMC interference you should shield your device in such a way that it will function appropriately as stated in the intended use.

So your example would be correct, you can pass the 2-40 but still fail the MDD requirements.
 
N

nikolaos

A ME must not necessaty have "essential performances", it depends on the type of the ME and what is an "unacceptable risk" for the manufacturer.

Regarding the 4.0 Ed., to define the pass/fail criteria to the EMD one must follow the MDD and not the 4.0 Ed.!! Something is missing here.
 
B

Bryce S

My first thought is that the EMC Directive is not needed - you are exempt under the MDD. As long as your test results match what you specify as pass/fail, then you pass. It will be up to your NB to accept or not.

Is your NB asking for 4th ed EMC already? I don't even see it in the EU OJ list of Standards.

good luck...
 
C

cdewitt

It is difficult for me to believe your product has no ?Essential Performance?. That is essentially saying it does nothing. You need to ask the question what is the medical function that it performs. After you?ve answered that you ask what are the parameters it must operate on or within to do that function. You then have a function and limits that define its operation that you can use to define Essential Performance.
 
N

nikolaos

[FONT=&quot]Hi cdewitt,[/FONT]
[FONT=&quot]According the 3rd Ed. of IEC60601-1 ESSENTIAL PERDORMANCE is:[/FONT]
[FONT=&quot]"performance of a clinical function, other than that related to [/FONT][FONT=&quot]BASIC SAFETY[/FONT][FONT=&quot], where loss or[/FONT]
[FONT=&quot]degradation beyond the limits specified by the [/FONT][FONT=&quot]MANUFACTURER [/FONT][FONT=&quot]results in an unacceptable [/FONT][FONT=&quot]RISK"[/FONT]
What you consider "essential performance" in my opinion are the "functional performance" which is another story.
[FONT=&quot][/FONT]
[FONT=&quot]
[/FONT]
 
Thread starter Similar threads Forum Replies Date
C IEC60601-1-2 :2020 EMC Reliability Analysis - Predictions, Testing and Standards 2
L IEC60601-1-11 Home Healthcare + IEC60601-1-2 EMC (Functional Earth allowed?) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
H EMC & Safety Testing of the product with the same Sr. No. for EC Declaration CE Marking (Conformité Européene) / CB Scheme 3
D IEC 60601-1-2: Is EMC immunity testing required for a device without essential performance? IEC 60601 - Medical Electrical Equipment Safety Standards Series 25
H Regarding the validity of EMC & Safety test reports without NABL & ILAC-MRA mark. CE Marking (Conformité Européene) / CB Scheme 2
H Regarding EMC & Safety retesting requirements of the product CE Marking (Conformité Européene) / CB Scheme 2
S What EMC 60601-1-2 version to use with 60601-2-47 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Essential performance and EMC immunity testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC 60601-1-2 EMC Tables IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
M EMC Directive and product safety standards CE Marking (Conformité Européene) / CB Scheme 1
B Evaluation of Basic Safety during EMC Immunity or Climate Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B Does EMC Directive 2014/30/EU cease to apply when the Machine Directive 2006/42/EC is applicable? CE Marking (Conformité Européene) / CB Scheme 4
H External power supply - requirements in safety/EMC test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
F EMC testing for a system that is provided with a computer IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
O EMC - EFT (Electrical Fast Transients) on SIP couples erroneously with other cable CE Marking (Conformité Européene) / CB Scheme 1
K Is a 3V torch classified as medical electrical equipment needing EMI/EMC testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Arguing with the test lab about failing in an EMC test IEC 60601 - Medical Electrical Equipment Safety Standards Series 12
E Airborne Medical Equipment EMI/EMC IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Portugal ISO 13485 Certification and EMC Test Houses Other Medical Device and Orthopedic Related Topics 3
T Question: Do we consider 60601-1-2 (4th ed) EMC testing a mode for a diagnostic port? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
O Medical Device EMC Risk Management CE Marking (Conformité Européene) / CB Scheme 4
O Guide on EMC for Functional Safety CE Marking (Conformité Européene) / CB Scheme 3
T CE Marking & Compliance to MDD, EMC, RoHS, and GPSD REACH and RoHS Conversations 4
J IEC 60601-2-2 - EMC Requirements - Surgical instrument manufacturer IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
O EMC testing per IEC 60601-1-2 for considering new cables CE Marking (Conformité Européene) / CB Scheme 1
O EMC testing under RED and IEC 60601-1-2 CE Marking (Conformité Européene) / CB Scheme 1
D EMC Testing for a Production Fixture CE Marking (Conformité Européene) / CB Scheme 0
Q IVD vs EMC OJ (Official Journal of the European Union) Harmonised Standard Conflicts EU Medical Device Regulations 1
A EMC Emission Issues - Communication between PC and Monitor IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Is EMC + Radio Re-Test Required? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Particular standards, essential performance in EMC Immunity Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F EMC (Electromagnetic Compatibility) Conformity Charts - IFU Requirements Misc. Quality Assurance and Business Systems Related Topics 3
G CE Marking - Low Voltage Directive and EMC Directive CE Marking (Conformité Européene) / CB Scheme 4
T Firmware version in EMC report / DoC CE Marking (Conformité Européene) / CB Scheme 1
W EMC, ESD for battery operated apparatus Other Medical Device and Orthopedic Related Topics 1
B What is our requirement for maintaining the EMC series of standards? IEC 61000 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P EMC requirements for an accessory to a medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T EMC for Medical Devices in Home Environment Other US Medical Device Regulations 4
J EMI/EMC Test differences for Radiography Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M EMC Testing for a US-only, Home-Based Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Safety & EMC certificate updated when MCU (MIcro-Controller) is changed? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
T EMC Tests - While certified power supply is part of device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B EMC Repeat Justification - Reducing NRTL testing required after modification of ME IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C EMC Test Report Requirements to IEC 60601-1-2:2009 ISO 13485:2016 - Medical Device Quality Management Systems 4
O CE TCF (Technical Construction File) Template for LVD/EMC/Pressure Directive CE Marking (Conformité Européene) / CB Scheme 5
P New EMC standard IEC 60601-1-2:2014 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
T IEC 60601-1-2 EMC and WiFi IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
T How to fulfill IEC 60601-1-2 EMC Testing Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S IEC 60601-1-2 EMC Testing Optional in US? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
S Does Australia/New Zealand require RoHS, WEEE, REACH and EMC? RoHS, REACH, ELV, IMDS and Restricted Substances 4

Similar threads

Top Bottom