Hi All,
I try to understand the IEC60601-1-2:2014 4.0 Ed. new requirements and I have started a gap analysis for some of our ME.
According IEC60601-1-2:2014:
Annex I ? immunity PASS/FAIL criteria can specify degradations that are acceptable because they do not result in unacceptable RISK? and also ?the ME passes the test if it continues to provide its BASIC SAFETY and ESSENTIAL PERFORMANCE?
According the Risk Management of our ME there are NO ESSENTIAL PERFORMANCES of a clinical function, other that related to BASIC SAFETY, where loss or degradation beyond the limits specified results in an unacceptable RISK.
Does this mean that whatever happens during IMMUNITY test, the test is PASS if the ME continues to provide BASIC SAFETY?
Checking with the 2014/30/EU EMC Directive:
?When placing their apparatus on the market, manufacturers shall ensure that they have been designed and manufactured in accordance with the essential requirements set out in Annex I. :
Equipment shall be so designed and manufactured, having regard to the state of the art, as to ensure that:
(b)it has a level of immunity to the electromagnetic disturbance to be expected in its intended use which allows it to operate without unacceptable degradation of its intended use.
These are the FUNCTIONAL PERFORMANCES but the IEC60601-1-2:2014 4.0 Ed. refers only to the ESSENTIAL PERFORMANCES which is quite different.
Example:
If during the ESD test (16KV) the display becomes black or gets broken, the test is PASS for IEC60601-1-2:2014 if the ME continues to provide its BASIC SAFETY but FAIL according the 2014/30/EU EMC Directive.
What is missing me? What is the correct interpretation?
Thanks in advance
I try to understand the IEC60601-1-2:2014 4.0 Ed. new requirements and I have started a gap analysis for some of our ME.
According IEC60601-1-2:2014:
Annex I ? immunity PASS/FAIL criteria can specify degradations that are acceptable because they do not result in unacceptable RISK? and also ?the ME passes the test if it continues to provide its BASIC SAFETY and ESSENTIAL PERFORMANCE?
According the Risk Management of our ME there are NO ESSENTIAL PERFORMANCES of a clinical function, other that related to BASIC SAFETY, where loss or degradation beyond the limits specified results in an unacceptable RISK.
Does this mean that whatever happens during IMMUNITY test, the test is PASS if the ME continues to provide BASIC SAFETY?
Checking with the 2014/30/EU EMC Directive:
?When placing their apparatus on the market, manufacturers shall ensure that they have been designed and manufactured in accordance with the essential requirements set out in Annex I. :
Equipment shall be so designed and manufactured, having regard to the state of the art, as to ensure that:
(b)it has a level of immunity to the electromagnetic disturbance to be expected in its intended use which allows it to operate without unacceptable degradation of its intended use.
These are the FUNCTIONAL PERFORMANCES but the IEC60601-1-2:2014 4.0 Ed. refers only to the ESSENTIAL PERFORMANCES which is quite different.
Example:
If during the ESD test (16KV) the display becomes black or gets broken, the test is PASS for IEC60601-1-2:2014 if the ME continues to provide its BASIC SAFETY but FAIL according the 2014/30/EU EMC Directive.
What is missing me? What is the correct interpretation?
Thanks in advance