"New" MDR implant card for absorbable substances

Wolf.K

Involved In Discussions
#1
Hi,

A friend of mine asked me if I already read the new MDR (yes) and if I noticed chapter 18 ("Implant card and information to be supplied to the patient with an implanted device". As he works for a company which manufactures hemostats and adhesion prevention products (like oxidized cellulose products) he has products class III, implants for transient, short or long term duration of use - depending how fast the substances are absorbed by the body. Some within days, some can still be found after more then 30 days (if the doctor put too much into the patient).

We both read the above mentioned article 18, also definition # 5 "implantable device", and now we wonder: to be MDR complient, does he has to add implant cards to each small bellows bottle of powder or each sheet of cellulose??

Until now, this was not a requirement, only for certain implants (like metal hip joint replacements, heart pace makers and so on).

Article 18 states a lot.
(1) The manufacturer of an implantable device shall provide together with the device the following... (a)(information e.g. serial number)... to (d)...

"In addition, the manufacturer shall provide the information referred to in point (a) of the first subparagraph on an implant card delivered with the device".

(2)"...together with the implant card, which shall bear their identity".

(3) The following implants shall be exempted from the obligations... sutures, staples, dental fillings and else

So, how shall we interprete this article 18???

:confused:
 
Elsmar Forum Sponsor

Pads38

Trusted Information Resource
#2
re: "New" MDR implant card for absorbable substances

I would suggest that your 'temporary' implants would fall into the excluded list.

The purpose of an Implant Card is to provide the patient with clear identification of their implant and any precautions that they need to be aware of going into the longer term. (MRI safety, airport scanners).

Where the implant is absorbed this would be largely irrelevant.
 

Wolf.K

Involved In Discussions
#4
Yes, for absorbable substances an implant card seems a bit strange, but the text of the MDR might include some mistakes, right?

I wonder where he could get "official" information on this issue?
 
#5
The following point in the MDR sheds some light, although not strictly applicable to your case, it is addressing a similar conundrum.

Page 2, para 8

It should be the responsibility of the Member States to decide on a case-by-case basis whether or not a product falls within the scope of this Regulation. In order to ensure consistent qualification decisions in that regard across all Member States, particularly with regard to borderline cases, the Commission should be allowed to, on its own initiative or at the duly substantiated request of a Member State, having consulted the Medical Device Coordination Group (‘MDCG’), decide on a case-by-case basis whether or not a specific product, category or group of products falls within the scope of this Regulation. When deliberating on the regulatory status of products in borderline cases involving medicinal products, human tissues and cells, biocidal products or food products, the Commission should ensure an appropriate level of consultation of the European Medicines Agency (EMA), the European Chemicals Agency and the European Food Safety Authority, as relevant.

So I guess, you might need to check with the CA of member state where you plan to sell the product.
 

Wolf.K

Involved In Discussions
#6
I found this leaflet published by the Australian Therapeutic Goods Administration:
https://www.tga.gov.au/sites/defaul...opean-medical-device-regulatory-framework.pdf

Australia is planning to align their regulation with the new European regulation. The leaflet is about implantable meshes, which are now all classified as class III, and they now need a patient implant card.

Therefore, I guess, all implantable devices will need an implant card, including hemostats (if left in the body and absorbed), tissue glues, adhesion prevention products, and else.
 

Wolf.K

Involved In Discussions
#7
Thread starter Similar threads Forum Replies Date
T Clarification on MDR - Article 18(d) - Implant Card EU Medical Device Regulations 14
B Clinical Evaluation Expert Panels - MDR EU Medical Device Regulations 1
M EU MDR - Labilities when a device is discontinued from the Market EU Medical Device Regulations 1
S National registrations for MDR certified devices in absence of EUDAMED EU Medical Device Regulations 1
Z MDR Class Ir deadline (Reusable surgical instruments) EU Medical Device Regulations 27
B MDR equvalent device - can you use a medicine as an equivalent 'device' EU Medical Device Regulations 8
N Harmonized Standards for MDR Certification EU Medical Device Regulations 6
K Is UDI (Unique Device Identifier) symbol mandatory under EU MDR? EU Medical Device Regulations 1
K EU MDR Annex 1 Chapter III: Information in the Instructions for Use-23.4 (e) the performance characteristics of the device; EU Medical Device Regulations 1
R Requirements for MDR Article 22 Systems EU Medical Device Regulations 1
H EU MDR Economic Operator notifications? EU Medical Device Regulations 1
M MDR EU Distributor for our CE Mark product in the for Netherlands EU Medical Device Regulations 6
R Declaration of conformity as per EU MDR (2017/745) CE Marking (Conformité Européene) / CB Scheme 5
K Class III Kit Box - Are translations mandatory under EU MDR? EU Medical Device Regulations 5
B Herbal ingredients MDR EU Medical Device Regulations 0
S MDR GSPR Clause 17 - Software Requirements EU Medical Device Regulations 2
N Which EN ISO 17664 version compliance to EU MDR? Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 3
D Hospital Decontamination and the MDR - Technical Files, UDI and PMS EU Medical Device Regulations 3
Kuldeep Singh Submission of MDR File - Partner Companies - Computed Tomography (CT) CE Marking (Conformité Européene) / CB Scheme 10
B How to classify a medical device based on MDR? EU Medical Device Regulations 3
A MDR GSPR Checklist wanted EU Medical Device Regulations 3
Rincewind MDR - GSPR - Annex I - Chapter II EU Medical Device Regulations 2
N EU MDR - Applicability Article 22 Systems and Procedure Packs CE Marking (Conformité Européene) / CB Scheme 4
B EU MDR Perscription product guidance EU Medical Device Regulations 1
cscalise Suggestions for MDR Auditing tools EU Medical Device Regulations 1
M Example of statement for procedure pack MDR Article 22? Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 0
H MDD CM, MDR manufacturer. 2020. Is it ok? EU Medical Device Regulations 1
M MDR, RED and LVD - Should our device comply with them? EU Medical Device Regulations 2
E UK MDR brexit out of sync with EU MDR deadline? Other Medical Device Regulations World-Wide 4
S MDR Delay - MDD design Change? (before new MDR DOA) EU Medical Device Regulations 8
S Class I MDR Article 10 – QMS EU Medical Device Regulations 1
dgrainger Medical Device News MDR postponed - link to Official Journal of the European Union (24-4-2020) EU Medical Device Regulations 2
K Supply Agreement Regarding MDR 2017/745 - Manufacturer-supplier cooperation Other Medical Device Related Standards 0
Ed Panek Medical Device News MDR Delayed 1 year (Posted 4/2020) EU Medical Device Regulations 9
dinaroxentool Former Class I device that is upscaled to IIa if the MDR is delayed EU Medical Device Regulations 2
B FDA-Medical Device Reporting (MDR )procedure compliant with 21CFR section 803 US Food and Drug Administration (FDA) 0
J PPE, the MDR and EU 2016/425 EU Medical Device Regulations 4
R An indication that the device is a medical device (MDR, Annex 23.2q) - applicable for accessories? EU Medical Device Regulations 5
R Clinical evaluation without clinical data - MDR Article 61(10) EU Medical Device Regulations 6
dgrainger Informational Postponement of MDR - adopted by the Commission - April 2020 Medical Device and FDA Regulations and Standards News 3
R Applicability of new non-harmonized standards (MDD/MDR) EU Medical Device Regulations 8
F The EU Drafts Bill to Postpone MDR by a Year (3/2020) Medical Device and FDA Regulations and Standards News 2
T Commission proposal being drawn up for postponement for date of application of MDR (2017/745) EU Medical Device Regulations 7
A MDR Article 22 applicability - Legal manufacturer EU Medical Device Regulations 6
M Estimating the benefit-risk ration under MDR EU Medical Device Regulations 1
F Will the EU postpone 2017/745 MDR Implementation - COVID-19 Impact Medical Device and FDA Regulations and Standards News 11
Ajit Basrur Physical Address on Product Labels (EU MDR perspective) EU Medical Device Regulations 2
B Notified Bodies for Software (MDR) EU Medical Device Regulations 1
P MDR Rule 10 interpretation - Active Device EU Medical Device Regulations 3
JoCam Certified Body Audit of MDR requirements EU Medical Device Regulations 4

Similar threads

Top Bottom