New to Calibration - Four Questions

A

AndrewHanna77

#1
Hello all, and thanks for having me in the forum.

As the thread title suggests I am very new to calibration. I work in the QA department an automotive engineering company in the UK and my management has decided they want to try and bring gauge calibration in house. Historically all calibration has been done by sending gauges to external companies. I have been tasked with researching the calibration processes in order to move this project forward. After trawling through endless information on the internet on various sites my head is starting to spin :confused:

Initially we want the in house calibration to cover thread gauges for machined tapped holes (M6x1.0, M8x1.25, M10x1.5 etc), go-no go plug gauges for un-tapped machined holes and digital vernier calipers. The plan is to have a set of externally calibrated master equipment which will then be used to calibrate all other gauges in house.

My question is can someone please give advise on:
  • what aspects of each of the mentioned gauges would need to be checked in order to meet correct calibration requirements? ie for a thread gauge checks should include major/ minor dia, pitch dia, depth step (where applicable) etc
  • what equipment would be required to calibrate the above mentioned gauges?
  • what documentation/ logs/ procedures would be required?
  • any other advise you can give someone starting out in calibration
Apologies for the length of the post. Basically I need to set up the calibration of these gauges from scratch and any/ all advise is much appreciated

Thanks in advance

Andrew
 
Elsmar Forum Sponsor

NikkiQSM

Quite Involved in Discussions
#2
In regards to the documentation - you will want a document explaining the calibration process of the items along with the allowable tolerance.

For the calibration report - you will want it to include:
* Calibration Date
* Next Calibration Due Day
* Name of person conducting the calibration
* How the unit was found - Was it intolerance of the previous calibration when you conducted the calibration?
* If any adjustments were needed - what were they?
* The calibration status of the equipment used to conduct the calibration.

Hope this helps.
 

Ron Rompen

Trusted Information Resource
#3
Just based on my own previous experience, plug gauges (GO-NOGO) are fairly simple as long as you are allowing yourself sufficient tolerance - you need something an order of magnitude 'better' than the tolerance you are using (example - if you are allowing a tolerance of ±0.0005" on a plug gauge, you need something with an ACCURACY of greater than 0.00005").
As for thread gauges, I would strongly recommend continuing to outsource them. Besides the obvious evaluations (pitch diameter, etc) there are other things that a good calibration lab can verify for you that you would be unable to do without investing in a lot of equipment and training.
 
B

BoardGuy

#4
The very basic start with calibration of threaded ring and plug gages would be to purchase the specifications to which they are made. Usually these specification will provide temperature requirements for inspection, tread forms and tolerance info etc.

Equipment wise you’re looking at a super bench micrometer, master set plugs for ring gages, and other calibration standards for plug gages. Work environment will need to be clean, temperature regulated and, typically for super mics, no air movement that is detectible.

Based on your customer requirements your calibration system may need to comply with ISO 17025 or be certified to 17025. Look at your national standards for calibration labs and UK’s very own National Physical Laboratory at Teddinton, Middlesex for guidance documentation. If I remember correctly NPL at one time had information of calipers, micrometers etc. free to the public via their website.
 

dwperron

Trusted Information Resource
#5
Andrew,

Welcome to the wonderful world of calibration!
First, having been both in the third party and internal lab calibration environments I would alert you to the pitfalls of bringing the calibrations in house. You will not only need to obtain the calibration equipment and a properly controlled environment, you will also need to create an entire calibration program. This will include setting up a calibration recall program, establishing a calibration history database, maintaining the calibration - and traceability - of your lab standards, producing and storing calibration records, training programs, a recall system for out of tolerance standards, etc. All the things that your current calibration providers are doing in the background. And you need to do this well enough to achieve whatever quality and program requirements that your products are committed to meet, especially to the point of being "audit proof". A good amount of time, effort and money will need to be dedicated not only to establish such a program, but also to maintain it.

A lot of this will depend on the volume of work involved, and available equipment. To calibrate the thread plugs and plug gauges you will need, at a minimum, a supermicrometer / horizontal measuring machine, a set of thread wires, and the accessories needed to make these measurements. If you need to purchase these you will require a large work volume to justify the high costs involved. These are also time consuming calibrations, the thread plugs requiring an advanced skill level. Plan on the labor costs involved as well. For the calipers you will need, at a minimum, a set of gage blocks that cover the length range of your calipers, a surface plate, ring(s) for the inside jaws, etc. All of this needs to be done in an adequately controlled environment.

In the end, the practicality of taking on the calibrations yourself will boil down to the work volume, equipment required, labor costs, and setting up a calibration program. I know that here in the States that mechanical calibrations tend to be inexpensive, so in most cases it is not economically feasible to do them yourself. But if it makes financial sense in your situation then got for it!
 
A

AndrewHanna77

#6
Hello all, and thank you for the replies. All good information to take on board. As mentioned I have no experience in calibration and am starting off from scratch so all information is welcome.

At the moment Im trying to put together the information to show the feasibility of launching such a project in house and my head is spinning with information regarding 10:1 rules, 3 wire methods, uncertainty etc

With regards to an environment, we have a large QA lab which houses our CMMs along with other equipment such as tensile/ hardness testers, profile projector, metal spectrometer, and is clean, temperature controlled etc. It needs to be to carry out the appropriate dimensional checks on our products. The plan is to dedicate an area within this lab to calibration equipment and carry out all calibration measurements there.

The big thing will obviously be the cost associated with getting the right equipment. As a large engineering company we have any amount of micrometers, but whether any of them are accurate enough to preform calibration is another thing :confused:
 
Thread starter Similar threads Forum Replies Date
T Torque tool calibration Oil and Gas Industry Standards and Regulations 2
E Calibration-Conditional Pass Measurement Uncertainty (MU) 2
S Calibration Interval calculation for ECG Machine Calibration Frequency (Interval) 4
Mikey324 External calibration - Finding in our 3rd party audit General Measurement Device and Calibration Topics 58
L A certificate of conformance used to support a calibration General Measurement Device and Calibration Topics 6
A AS9100 - 7.1.5.2 In-House Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 25
B UV-A and Illuminance meter Calibration Setup General Measurement Device and Calibration Topics 0
S Calibration Related Posters General Measurement Device and Calibration Topics 2
F Quality Control in Electrical Calibration ISO 17025 related Discussions 0
C Narrow definition of IMTE or calibration possible? ISO 13485:2016 - Medical Device Quality Management Systems 8
dhuot1990 In-House Calibration Program - Biotech Industry General Measurement Device and Calibration Topics 3
M Reference part calibration or verification General Measurement Device and Calibration Topics 12
M Calibration of equipment IATF 16949 - Automotive Quality Systems Standard 4
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 5
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 5
S Calibration Due date-Critical Connie General Measurement Device and Calibration Topics 16
Q Request Calibration Info Calibration and Metrology Software and Hardware 16
R Calibration Log - Archive Manufacturing and Related Processes 2
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 6
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 32
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0

Similar threads

Top Bottom