NIST Traceability - Calibrating a grid w/ laser interferonmeters and microscopes
S
Sentient
#1
Hello
We're making a large glass grid with targets on it, and planning to calibrate this grid on a large X,Y,Z stage system with a laser interferonmeter/s and microscopic optics.
For this process, what steps would need to be taken to achieve NIST Traceability on the large grid?
I am most appreciative for your informed replies.
Elsmar Forum Sponsor
Jerry Eldred
Forum Moderator
Super Moderator
#2
I am not familiar with the specifics of your equipment. But for it to be NIST traceable, it must somehow be compared with NIST traceable length. If your XYZ stage system has been calibrated with NIST traceable standards, that will provide traceability. If your XYZ system is not calibrated, you will need to compare its accuracy against traceable standards.
Sorry for the redundancy. I think this should be straightforward. Hope that is of help.
G
Graeme
#3
I was briefly involved in a similar project about 8-9 years ago. The client then had a CMM system that used a glass grid as the reference standard, and then the CMM made measurements of other large, flat objects.
My recollection from then is that NIST (Gaithersburg, MD) does have a calibration service for these grids. If you are manufacturing the grids, you may want to consider having NIST calibrate your master grid and use that to transfer the measurements. If you are a user, you may want to check with the manufacturer of the grid for traceability.
All of your dimesional measuring devices need to be calibrated, of course. That is, as I recall, a fairly routine and easy process with the laser interferometer. (At least, with the brand we were using then.)
Graeme C. Payne
ASQ Certified Quality Engineer
[email protected]
[This message has been edited by Graeme (edited 28 February 2001).]
M
metrologyguy
#5
NIST traceability has recently been redefined. In the good old days if you had a cert with a NIST number, it was assumed to be traceable. For the real definition, I recommend you check out www.a2la.org.
Thread starter Similar threads Forum Replies Date
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
G Traceability - Checklist from NIST General Measurement Device and Calibration Topics 4
Raffy ISO 13485 NIST Traceability requirements for calibration of test equipment ISO 13485:2016 - Medical Device Quality Management Systems 3
A Is NIST traceability required if a Refractometer calibration is done in-house? Calibration Frequency (Interval) 6
D Since NIST # can not be used to verify Traceability - What an Auditor should do AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
M Pipette Calibration & ISO 8655 - No NIST Standard(s) Traceability ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
S ISO 17025 requirements not enough to prove NIST Traceability? ISO 17025 related Discussions 10
S NIST Traceability Numbers: Dense people, needle in a haystack General Measurement Device and Calibration Topics 16
C Steps to NIST - Measurement Traceability for a Certified Operator Class General Measurement Device and Calibration Topics 4
Jerry Eldred Presentation About NIST Traceability General Measurement Device and Calibration Topics 8
J NIST Traceability: OEM Products - New electronics measurement device General Measurement Device and Calibration Topics 5
R Outside Calibration - Accredited, certified - Is NIST traceability guaranteed? General Measurement Device and Calibration Topics 8
Hershal Traceability - NIST Policy on Traceability, Section II.B.4 General Measurement Device and Calibration Topics 4
Marc NIST Traceability Numbers - Confused Yet? General Measurement Device and Calibration Topics 5
W 17025 and NIST handbook relationship (need advice) ISO 17025 related Discussions 8
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
R What is the correct way to refer to NIST? Calibration Frequency (Interval) 4
M NIST is one of the government services shut down General Measurement Device and Calibration Topics 2
G NIST SOP 29 - Assignment of Uncertainty - Question General Measurement Device and Calibration Topics 0
S Reference calibration tools (traceable to NIST) that are not used in production ISO 13485:2016 - Medical Device Quality Management Systems 4
Marc NIST’s Trove of Scientific Instruments and Relics After Work and Weekend Discussion Topics 2
N Anyone working on NIST SP 800-171 (Network and Information Security)? Records and Data - Quality, Legal and Other Evidence 4
D NIST HDBK 44 Table T.3. Class III Tolerance in Divisions Clarification General Measurement Device and Calibration Topics 4
C How to state that Calibration Services & Equipment are Traceable to NIST General Measurement Device and Calibration Topics 5
G Where to get a NIST Traceable Master Glass Refractive Index Sample General Measurement Device and Calibration Topics 5
A Calibration Service Suppliers refuse NIST reference numbers on C of C General Measurement Device and Calibration Topics 21
Marc CMM Training at NIST - August 28-29, 2013 General Measurement Device and Calibration Topics 1
M NIST Standard for CO2 Monitor Calibration General Measurement Device and Calibration Topics 5
N NIST Special Publications and Technical Notes inquiry General Measurement Device and Calibration Topics 4
J Unbroken Chain to NIST or a Physical Property - Request for Peer Review General Measurement Device and Calibration Topics 6
S Does NIST Certify Calibration Laboratories? General Measurement Device and Calibration Topics 4
T Outsourcing Calibration - NIST, OEM, and Calibration Laboratory General Measurement Device and Calibration Topics 2
S NIST equivalents for England and Germany Supplier Quality Assurance and other Supplier Issues 1
smryan Does the NIST certificate that came with my new ball-bar have any value? General Measurement Device and Calibration Topics 8
B NIST Traceable Thermometer Uncertainty Measurement Uncertainty (MU) 5
AnaMariaVR2 New NIST reference materials available for popular diet supplement Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
Marc NIST - US Revamps Science, Technology Standard-Setting Efforts General Measurement Device and Calibration Topics 0
R Does being traceable to NIST meet the requirements of ISO standards? General Measurement Device and Calibration Topics 16
N Thermocouple Drift Determination - Drift test done by NIST General Measurement Device and Calibration Topics 3
C Calibration Of Stop Watch using the NIST Clock General Measurement Device and Calibration Topics 4
Marc NIST Handbook of Mathematical Functions reference text World News 0
R NIST Test Number - What does it mean? General Measurement Device and Calibration Topics 3
ScottK NIST comes out against DRE voting machines. Coffee Break and Water Cooler Discussions 7
B Looking for Calibration / Reference Standards tracable to NIST General Measurement Device and Calibration Topics 15
R NIST - Primary Standard - Traceable to what? What exactly does that imply? General Measurement Device and Calibration Topics 10
Tim Folkerts Great NIST Stats Reference Statistical Analysis Tools, Techniques and SPC 2
Marc Uncertainty ListServe - NIST TN 1297 Measurement Uncertainty (MU) 3
M Calibration of Timers to NIST - How are digital and analog timers calibrated? General Measurement Device and Calibration Topics 15
D Accreditation Scheme:relationships between ANSI, NIST, NCSL, ILAC, APLAC, NCCLS, etc. General Measurement Device and Calibration Topics 3
Marc NIST 'High Integrity Software Systems Assurance' resource (site semi-abandoned) Software Quality Assurance 2
Similar threads
Top Bottom