Our company is releasing a new electronics measurement device. Let me know if the following has any traceability gaps. I feel uneasy with this....

1. We are the OEM for item "A" (DUT)
2. Item "A" requires calibration
3. Our engineering dept. designed a calibration device (standard) to calibrate item "A" (DUT)
4. This calibration device (standard) obtains power from item "A" (DUT)
(CAN WE CALIBRATE EQUIPMENT WE MANUFACTURE WITH EQUIPMENT WE MANUFACTURE?)
What does Fluke do?
I value being able to utilize established standards such as oscilloscopes/dmm's/function generators but our engineers are trying to eliminate the need for such. Does the above "plan" fit within NIST Traceability requirements?
Your opinions and comments are extremely welcome!

thanks!
 
Last edited by a moderator:
Elsmar Forum Sponsor

Jerry Eldred

Forum Moderator
Super Moderator
#2
Many instruments have a "SELF CALIBRATION" feature. HOWEVER (and it is a BIG however), this is not a substitute for a NIST traceable calibration. The Agilent (HP) 3458A for example, has an AUTOCAL feature which linearizes range to range relationships RATIOMETRICALLY (important term) against its internal 10 VDC reference, 10 Kilohm input and electrically established zero offset cal factors.

This is only a confidence check, and not at all a true calibration. There is a principle in metrology called the principal of uncertainty growth. That is, from the moment you adjust/verify a measurand, the uncertainty of what it's true value grows at some predictable rate. The 3458A above will drift away from it's actual values, and the AUTOCAL function will ratiometrically, re-establish all ranges, etc. against those drifted values. So by a years time (statistically, anyway), that 3458A will have drifted out toward it's tolerance limits. I have high respect for the 3458A, so don't anyone interpret this as negative toward it.

The only way to re-establish the 3458A against known accuracies, is by verifying it's relationship to NIST (or other) traceable values. The purpose of traceability is to assure everyone is on the same page (so to speak).

I can pretty well assure that if you design an instrument that measures NIST traceable parameters, and engineeers design it only to "SELF-CALIBRATE" (an oxymoron), it will, over some predictable time, drift away from true nominal.

So No, you can not self-calibrate any instrument. About the only exceptions (crazy, but for the sake of illustration) would be physical constants. If you built a Cesium Beam or a GPS receiver into a unit, it would not need external frequency cal. If you built a Triple Point of Water (not really feasible) into a unit, it would not need zero degree Celsius cal. But these things are not feasible.

Let me throw in a qualifier on my above posting. If your engineers design a "GOLDEN UNIT" to calibrate your instrument, this may potentially simplify the process greatly. The GOLDEN UNIT would require NIST traceability. If your GOLDEN UNIT adequately covers function/ranges of your instrument, it may do what they require. The question is assuring the GOLDEN UNIT has adequate coverage. It does not replace NIST traceable calibration, it simplifies it.

I hope this helps.
 
J

joecalibratio

#3
Thank you so much for the response. I battle with our engineers because they resist the point of "adequately covers function/ranges of your instrument" by argueing that it is all digital and a failure would not result in an oot condition but a total failure (no signal/data versus and improper signal).

Joel
 

Jerry Eldred

Forum Moderator
Super Moderator
#4
I've had numerous instruments over the years where the OEM said it didn't need calibration (for some reason, none of the models currently come to mind). As a Metrologist for nearly thirty years, you can accomplish a lot ratiometrically. However, in my humble opinion, no quantitative instrument which measures a traceable type parameter can function accurately for an indefinite period without benefit of external tracability.

The only exception (anomaly) I can think of is the Fluke 77. One company years ago had a very large inventory of Fluke 77 handheld multimeters (can't remember their name; but I believe it was a major defense contractor type company). With thousands of the meters in their inventory, over something like ten years of history, they had zerro out-of-tolerances. So they removed calibration intervals based on that. However, I have had to adjust even those, I have had dFluke 77's with deteriorated range switches (that cause intermittend ohms), and other problems. So although one of the most robust handheld multimeters ever made, it is not completely above the need for external calibration.

Bottom line... no instrument is completely exempt from calibration (if you want what it measures to be quantitatively accurate as relates to the rest of the world.
 

BradM

Staff member
Admin
#5
Awesome posts, Jerry! Thank you.

Ok, you have device D.U.T. that works on it's own. You have this "standard" device that becomes integral with the D.U.T. Is the standard just using power from the D.U.T.? What type of algorithm is the standard utilizing to assure the D.U.T. is working properly? Is there some measurement verification of some kind?

If the customers purchase the D.U.T., will they have to buy the standard also? Can you re-certify the standard?

Are these devices constant voltage (120 vac), or battery operated? Do they have low voltage shutoff?

What kind of verification will be performed on the standard to assure it's continued success?

Jerry is right on. In my shortened life I have seen multiple setups, due to necessity. Somewhere in that system you will need a N.I.S.T. verification to assure that nothing has gone haywire.
:topic:
I don't know about you folks, but I wished everyone did things like Fluke. First, their equipment is so good. Too, most people believe their stated tolerances are quite conservative.
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#6
Good question and some good answers.....

As for what Fluke does, their NVLAP accredited lab has a Josephson standard that provides as direct as possible to SI.

If the manufacturer in question here has a similar arrangement, then the traceability change can be easily established.....otherwise the calibration standards need to be sent out.....

One other thing to consider.....most manufacturers of test and measurement equipment are not calibrated.....this manufacturer should consider accreditation to ANS/ISO/IEC 17025 for its calibration operation.....

Hershal
 
Thread starter Similar threads Forum Replies Date
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
G Traceability - Checklist from NIST General Measurement Device and Calibration Topics 4
Raffy ISO 13485 NIST Traceability requirements for calibration of test equipment ISO 13485:2016 - Medical Device Quality Management Systems 3
A Is NIST traceability required if a Refractometer calibration is done in-house? Calibration Frequency (Interval) 6
D Since NIST # can not be used to verify Traceability - What an Auditor should do AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
M Pipette Calibration & ISO 8655 - No NIST Standard(s) Traceability ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
S ISO 17025 requirements not enough to prove NIST Traceability? ISO 17025 related Discussions 10
S NIST Traceability Numbers: Dense people, needle in a haystack General Measurement Device and Calibration Topics 16
C Steps to NIST - Measurement Traceability for a Certified Operator Class General Measurement Device and Calibration Topics 4
Jerry Eldred Presentation About NIST Traceability General Measurement Device and Calibration Topics 8
R Outside Calibration - Accredited, certified - Is NIST traceability guaranteed? General Measurement Device and Calibration Topics 8
Hershal Traceability - NIST Policy on Traceability, Section II.B.4 General Measurement Device and Calibration Topics 4
S NIST Traceability - Calibrating a grid w/ laser interferonmeters and microscopes General Measurement Device and Calibration Topics 4
Marc NIST Traceability Numbers - Confused Yet? General Measurement Device and Calibration Topics 5
W 17025 and NIST handbook relationship (need advice) ISO 17025 related Discussions 8
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
R What is the correct way to refer to NIST? Calibration Frequency (Interval) 4
M NIST is one of the government services shut down General Measurement Device and Calibration Topics 2
G NIST SOP 29 - Assignment of Uncertainty - Question General Measurement Device and Calibration Topics 0
S Reference calibration tools (traceable to NIST) that are not used in production ISO 13485:2016 - Medical Device Quality Management Systems 4
Marc NIST’s Trove of Scientific Instruments and Relics After Work and Weekend Discussion Topics 2
N Anyone working on NIST SP 800-171 (Network and Information Security)? Records and Data - Quality, Legal and Other Evidence 4
D NIST HDBK 44 Table T.3. Class III Tolerance in Divisions Clarification General Measurement Device and Calibration Topics 4
C How to state that Calibration Services & Equipment are Traceable to NIST General Measurement Device and Calibration Topics 5
G Where to get a NIST Traceable Master Glass Refractive Index Sample General Measurement Device and Calibration Topics 5
A Calibration Service Suppliers refuse NIST reference numbers on C of C General Measurement Device and Calibration Topics 21
Marc CMM Training at NIST - August 28-29, 2013 General Measurement Device and Calibration Topics 1
M NIST Standard for CO2 Monitor Calibration General Measurement Device and Calibration Topics 5
N NIST Special Publications and Technical Notes inquiry General Measurement Device and Calibration Topics 4
J Unbroken Chain to NIST or a Physical Property - Request for Peer Review General Measurement Device and Calibration Topics 6
S Does NIST Certify Calibration Laboratories? General Measurement Device and Calibration Topics 4
T Outsourcing Calibration - NIST, OEM, and Calibration Laboratory General Measurement Device and Calibration Topics 2
S NIST equivalents for England and Germany Supplier Quality Assurance and other Supplier Issues 1
smryan Does the NIST certificate that came with my new ball-bar have any value? General Measurement Device and Calibration Topics 8
B NIST Traceable Thermometer Uncertainty Measurement Uncertainty (MU) 5
AnaMariaVR2 New NIST reference materials available for popular diet supplement Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
Marc NIST - US Revamps Science, Technology Standard-Setting Efforts General Measurement Device and Calibration Topics 0
R Does being traceable to NIST meet the requirements of ISO standards? General Measurement Device and Calibration Topics 16
N Thermocouple Drift Determination - Drift test done by NIST General Measurement Device and Calibration Topics 3
C Calibration Of Stop Watch using the NIST Clock General Measurement Device and Calibration Topics 4
Marc NIST Handbook of Mathematical Functions reference text World News 0
R NIST Test Number - What does it mean? General Measurement Device and Calibration Topics 3
ScottK NIST comes out against DRE voting machines. Coffee Break and Water Cooler Discussions 7
B Looking for Calibration / Reference Standards tracable to NIST General Measurement Device and Calibration Topics 15
R NIST - Primary Standard - Traceable to what? What exactly does that imply? General Measurement Device and Calibration Topics 10
Tim Folkerts Great NIST Stats Reference Statistical Analysis Tools, Techniques and SPC 2
Marc Uncertainty ListServe - NIST TN 1297 Measurement Uncertainty (MU) 3
M Calibration of Timers to NIST - How are digital and analog timers calibrated? General Measurement Device and Calibration Topics 15
D Accreditation Scheme:relationships between ANSI, NIST, NCSL, ILAC, APLAC, NCCLS, etc. General Measurement Device and Calibration Topics 3
Marc NIST 'High Integrity Software Systems Assurance' resource (site semi-abandoned) Software Quality Assurance 2

Similar threads

Top Bottom