Hello,
I work as a QC Chemist for a chemical manufacturing company. We're trying to obtain an ISO 9001:2008 certification.

NIST Traceable thermometers, purchased from vendors, have a stated uncertainty of +/- some value at various temps. Older lab thermometers that have been in use for many years are calibrated, by me, against high precision NIST Traceable Thermometers. These are the really expensive ones that are designed for calibration only purposes.

Question. If I have a thermometer that is minus -0.1C versus the NIST thermometer, do I report it as +/- 0.1C of do I report it as -0.1C? What is the correct protocol for reporting uncertainty for thermometers calibrated in the lab?
Thanks.

Bart
 
Elsmar Forum Sponsor

BradM

Staff member
Admin
#2
Hello Bart! :bigwave:

Just to clarify, are we speaking of uncertainty or accuracy? They are two different things.

Uncertainty is typically an estimation; a calculation of sorts.

Here is a link giving a brief definition of uncertainty:

http://www.l-a-b.com/content/measurement-uncertainty

So you have thermometers (called lab thermometers) and then the high precision thermometer. So you use the high precision ones to verify the accuracy of the lab thermometers. Correct?

So you have the high precision thermometers verified by a competent lab. This thermometer should be substantially higher accuracy than the lab thermometers.

You verify the lab thermometers and find a deviation. So if your high precision thermometer displays 0.1°C and the lab thermometers displays 0.3°C, then you can either generate a report with the actual values observed. You could also display the test value of 0.1°C and report the lab thermometer as +.2°C.

Not sure if you are estimating a true uncertainty. If so, that's another matter. If not, then you will want to assure the high precision thermometer has sufficiently better accuracy than the lab thermometers. A 4 to 1 ratio is typical.:)
 
B

b58chtec

#3
Brad,
Thanks for your reply.

Okay. Didn't realize there was a difference between uncertainty & accuracy. I'll check out the link you sent. I assumed they were interchangeable.

Anyway, the high precision thermometer(s) have values like +/- 0.06C at 20C or +/-0.03 at 30C & the regular thermometers came from the manufacturers with stated accuracies of +/- 1.5C or +/-1.0C. I think I'm ok there.

So, I don't have to report the regular thermometers as +/- value. Instead, I report the correction only of either + value or -value. Correct?

Thanks.

Bart
 

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#4
Brad,

Every lab is different. It's not quite that easy to take the difference in your temperature and just transfer it. I'm not a testing expert, but if you would like to give me a call next week when one of our testing program managers are in the office, we can tackle that question with the expert. <phone number removed> I'm the only Gary in the office.

Many people have questions about uncertainty budgets. We also have calculators you may find uselful in the resources tab on our website at <removed>
Last warning - This is not a place for you to come and drum up business. Your post has been reported and it's apparent you haven't read the ToS, or you have and are intentional violating it. You are making yourself look pretty bad - Who wants to do business with someone who either doesn't read the rules, or just decides not to follow the rules?
 

BradM

Staff member
Admin
#5
Anyway, the high precision thermometer(s) have values like +/- 0.06C at 20C or +/-0.03 at 30C & the regular thermometers came from the manufacturers with stated accuracies of +/- 1.5C or +/-1.0C. I think I'm ok there.

So, I don't have to report the regular thermometers as +/- value. Instead, I report the correction only of either + value or -value. Correct?

Thanks.

Bart
Well... I think I might have confused you and mixed two different things, here. If you are discussing the accuracy of the thermometer then it is usually expressed as a +/- value. The actual error that you observe while verifying the device is + value or -value.

You have two thermometers- S (standard) and B (bench thermometer).

S should have a stated accuracy that the thermometer should be able to meet. If it does not meet that accuracy, then it would be considered out of tolerance. It should be sufficiently more accurate than B. S should be calibrated by an outside lab, and be assessed against an identified specification (either your own, or mfg.).

Then you test B using S. B should have a stated specification also. This specification is what you expect the thermometer to meet while testing.

Let's say it is +/-5°C. When the S displays 50.0°C and B states 50.4°C, then you should list a +.4°C for the B thermometer @50°C.

However, that is still within your +/-5°C tolerance for the thermometer.

Now... as mentioned in this thread, the more correct method is to estimate the uncertainty of your measurement system. However, that tends to be a little more involved for the average individual. It's the better way; but not the simplest. :) So that is why I keep mentioning accuracy here.

I hope that explained things a bit better. :agree1:
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#6
BradM has been giving some great advice. Only one thing I would add at this point.

For the high accuracy thermometers, when there is +/- 0.06 C then that suggests that the value presented is the uncertainty. If so, the calibration cert should state if it is reported at k=2 to approximate 95% confidence, of if from a NMI then maybe k=3. If k=2, then 0.06 divides by 2 (assumes normal distribution) to return the value to standard uncertainty, necessary for calculation of your own uncertainty.
 
Thread starter Similar threads Forum Replies Date
F Nist traceable calibration certificates General Measurement Device and Calibration Topics 1
S Reference calibration tools (traceable to NIST) that are not used in production ISO 13485:2016 - Medical Device Quality Management Systems 4
C How to state that Calibration Services & Equipment are Traceable to NIST General Measurement Device and Calibration Topics 5
G Where to get a NIST Traceable Master Glass Refractive Index Sample General Measurement Device and Calibration Topics 5
R Does being traceable to NIST meet the requirements of ISO standards? General Measurement Device and Calibration Topics 16
R NIST - Primary Standard - Traceable to what? What exactly does that imply? General Measurement Device and Calibration Topics 10
R What is the correct way to refer to NIST? Calibration Frequency (Interval) 4
M NIST is one of the government services shut down General Measurement Device and Calibration Topics 2
G NIST SOP 29 - Assignment of Uncertainty - Question General Measurement Device and Calibration Topics 0
Marc NIST’s Trove of Scientific Instruments and Relics After Work and Weekend Discussion Topics 2
N Anyone working on NIST SP 800-171 (Network and Information Security)? Records and Data - Quality, Legal and Other Evidence 4
G Traceability - Checklist from NIST General Measurement Device and Calibration Topics 4
Raffy ISO 13485 NIST Traceability requirements for calibration of test equipment ISO 13485:2016 - Medical Device Quality Management Systems 3
A Is NIST traceability required if a Refractometer calibration is done in-house? Calibration Frequency (Interval) 6
D NIST HDBK 44 Table T.3. Class III Tolerance in Divisions Clarification General Measurement Device and Calibration Topics 4
A Calibration Service Suppliers refuse NIST reference numbers on C of C General Measurement Device and Calibration Topics 21
Marc CMM Training at NIST - August 28-29, 2013 General Measurement Device and Calibration Topics 1
M NIST Standard for CO2 Monitor Calibration General Measurement Device and Calibration Topics 5
N NIST Special Publications and Technical Notes inquiry General Measurement Device and Calibration Topics 4
J Unbroken Chain to NIST or a Physical Property - Request for Peer Review General Measurement Device and Calibration Topics 6
S Does NIST Certify Calibration Laboratories? General Measurement Device and Calibration Topics 4
T Outsourcing Calibration - NIST, OEM, and Calibration Laboratory General Measurement Device and Calibration Topics 2
S NIST equivalents for England and Germany Supplier Quality Assurance and other Supplier Issues 1
smryan Does the NIST certificate that came with my new ball-bar have any value? General Measurement Device and Calibration Topics 8
D Since NIST # can not be used to verify Traceability - What an Auditor should do AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
M Pipette Calibration & ISO 8655 - No NIST Standard(s) Traceability ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
AnaMariaVR2 New NIST reference materials available for popular diet supplement Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
S ISO 17025 requirements not enough to prove NIST Traceability? ISO 17025 related Discussions 10
Marc NIST - US Revamps Science, Technology Standard-Setting Efforts General Measurement Device and Calibration Topics 0
S NIST Traceability Numbers: Dense people, needle in a haystack General Measurement Device and Calibration Topics 16
N Thermocouple Drift Determination - Drift test done by NIST General Measurement Device and Calibration Topics 3
C Calibration Of Stop Watch using the NIST Clock General Measurement Device and Calibration Topics 4
C Steps to NIST - Measurement Traceability for a Certified Operator Class General Measurement Device and Calibration Topics 4
Marc NIST Handbook of Mathematical Functions reference text World News 0
R NIST Test Number - What does it mean? General Measurement Device and Calibration Topics 3
Jerry Eldred Presentation About NIST Traceability General Measurement Device and Calibration Topics 8
J NIST Traceability: OEM Products - New electronics measurement device General Measurement Device and Calibration Topics 5
ScottK NIST comes out against DRE voting machines. Coffee Break and Water Cooler Discussions 7
R Outside Calibration - Accredited, certified - Is NIST traceability guaranteed? General Measurement Device and Calibration Topics 8
Hershal Traceability - NIST Policy on Traceability, Section II.B.4 General Measurement Device and Calibration Topics 4
B Looking for Calibration / Reference Standards tracable to NIST General Measurement Device and Calibration Topics 15
Tim Folkerts Great NIST Stats Reference Statistical Analysis Tools, Techniques and SPC 2
S NIST Traceability - Calibrating a grid w/ laser interferonmeters and microscopes General Measurement Device and Calibration Topics 4
Marc Uncertainty ListServe - NIST TN 1297 Measurement Uncertainty (MU) 3
M Calibration of Timers to NIST - How are digital and analog timers calibrated? General Measurement Device and Calibration Topics 15
Marc NIST Traceability Numbers - Confused Yet? General Measurement Device and Calibration Topics 5
D Accreditation Scheme:relationships between ANSI, NIST, NCSL, ILAC, APLAC, NCCLS, etc. General Measurement Device and Calibration Topics 3
Marc NIST 'High Integrity Software Systems Assurance' resource (site semi-abandoned) Software Quality Assurance 2
G Mil-Std-130N, can calibration be a nationally traceable standard? Various Other Specifications, Standards, and related Requirements 0
N International / Traceable Standards - Procedures for Calipers, Micrometers, Etc. ISO 17025 related Discussions 2

Similar threads

Top Bottom