No Design - Which points from 7.3.1 'Design And Development' are necessary to comply

I

Internet-101

#1
7.3.1 Design and Development

7.3.2.1 Product Design Input
7.3.3.1 Product Design Output

Are these points from Design Responsibility exclusively?

This is because other ones (7.3.1, 7.3.1.1, 7.3.2, 7.3.2.2, 7.3.2.3, 7.3.3, 7.3.3.2, 7.3.4, 7.3.4.1, 7.3.5, 7.3.6, 7.3.6.1, 7.3.6.2, 7.3.6.3 & 7.3.7) have part of Design and Development Process of the Product.
If some company hasn’t a Design Responsibility but they need to introduce new products according ISO/TS 16949:2K terms on 7.3 “Design and Development” Element.
Will they need to comply with (7.3.1, 7.3.1.1, 7.3.2, 7.3.2.2, 7.3.2.3, 7.3.3, 7.3.3.2, 7.3.4, 7.3.4.1, 7.3.5, 7.3.6, 7.3.6.1, 7.3.6.2, 7.3.6.3 & 7.3.7) points only without 7.3.2.1 & 7.3.3.1 points without any non-conformance about it?
Or Which points from 7.3.1 “Design And Development” are necessary to comply If some company doesn’t have Design Responsibility?

I hope you can understand the situation.

Thanks in advance for your help.


Smile is good for your health!

:bigwave: :bigwave:
 
Elsmar Forum Sponsor
T

tattva

#2
Taking advatnage of this new threat, I´ve another question I hope you can help me to fully understand.

We don´t design a thing not even the manufacturing process, so do I still have to deal with other points regarding design such as competence 6.2.2.1?

Thanks in advance:bigwave:
 

Manoj Mathur

Quite Involved in Discussions
#3
Tattva,

It is hard to digest for me that You don't design(Moderate & Regulate) the manufacturing processes or the elements of it.

May I know what is your product and How do you make it?

Manoj Mathur
 
#4
Internet,
This is a difficult element to take exceptions since it applies to both the "Design and Development" of the product. At what point do you separate the design from the development? You may not be designing a product, but you may be developing a product. What is meant by development? Only you can determine that point.
IMO all of 7.3 requirenments would apply, except those explicitly related to the design of the product,i.e., design review.

Tattva,
7.3 Design and development applies to "products" and "processes". You may not design a process, but you may develop one.
As stated above, IMO all of 7.3 would apply, except those explicitly related to the deisgn of the product.
 
T

tattva

#5
Thank You all VERY MUCH!

I was talking to quality manager yesterday, and told me we can not make aside point 7.3, ´cause we do develop de manufacturing process.

We pour polyurethane, and what I meant by " not designing a thing " is that all technology and equipment is designed by coorporate engineering. But as you say, developing the process is another thing,we do that.
:biglaugh:
 

Howard Atkins

Forum Administrator
Leader
Admin
#6
I have been pondering this problem and after some resarch including the thread at

http://Elsmar.com/Forums/showthread.php?s=&threadid=4602
I have to agree with Sam:
From ISO 9000:2000

3.4.4
design and development
set of processes (3.4.1) that transforms requirements (3.1.2) into specified characteristics (3.5.1) or into the
specification (3.7.3) of a product (3.4.2), process (3.4.1) or system (3.2.1)
NOTE 1 The terms “design” and “development” are sometimes used synonymously and sometimes used to define different
stages of the overall design and development process.
NOTE 2 A qualifier can be applied to indicate the nature of what is being designed and developed (e.g. product design and
development or process design and development).


3.4.1
process

set of interrelated or interacting activities which transforms inputs into outputs
NOTE 1 Inputs to a process are generally outputs of other processes.
NOTE 2 Processes in an organization (3.3.1) are generally planned and carried out under controlled conditions to add value.
NOTE 3 A process where the conformity (3.6.1) of the resulting product (3.4.2) cannot be readily or economically verified is
frequently referred to as a “special process”.


3.4.2
product
result of a process (3.4.1)
NOTE 1 There are four generic product categories, as follows:

— services (e.g. transport);
— software (e.g. computer program, dictionary);
— hardware (e.g. engine mechanical part);
— processed materials (e.g. lubricant).
7.3.1 must be in.
 
M

Mike Smith

#7
We are an electronics remanufacturing company. We did not develop or design the product that we remanufacture. We obtained a waiver from our customer stating that we did not have to comply to this requirement of the standard of QS9000. I am hoping to do the same for 16949 since we do not design the product or the MANUFACTURING process.
 

howste

Thaumaturge
Trusted Information Resource
#8
Mike, I would imagine that you do design/develop the remanufacturing processes. I believe you would need to include these design and development of these processes to meet 16949 requirements. From TS 16949 1.2:
The only permitted exclusions for this Technical Specification relate to 7.3 where the organization is not responsible for product design and development.
Permitted exclusions do not include manufacturing process design.
 
M

Mike Smith

#9
This brings up a good point. It also states in 1.2 "Where any requirement of this international standard cannot be applied due to the nature of an organization and its product, this can be considered for exclusion". Later in 1.2 it states "the only permitted exclusions for this TS relate to 7.3 where the organization is not responsible for product design and development". Does this contradict itself? First it states any requirement and then only requirement. Anyone?
 
#10
Mike S.
This is not a contradiction.
"Considered for exclusion" means you can ask for a waiver.
"Permitted exclusions", IATF permitted, means you do not have to ask for a waiver, only that you provide evidence to your CB that you do not design or develop product.
 
Thread starter Similar threads Forum Replies Date
V Handling open points in design reviews 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
R Understanding a few points on ISO 9001's Design and Development Planning ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
A Medical Device Electronic Systems Design - Points to be cleared EU Medical Device Regulations 2
A What factors must be considered during gauge design?what are VG points ? Design and Development of Products and Processes 1
B Validation of design for valve api 6d 25 edition ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 0
B Hi , everyone i need a procedure for validation of design prototype api 6d (valve manufacturing) Oil and Gas Industry Standards and Regulations 0
K Exclusion of 8.3 Design and Development for Design Companies ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
J Recognised/reputable courses for ISO 13485 Section 7.3 (Design & Development)? Other ISO and International Standards and European Regulations 1
J Design/Development. Oil and Gas Industry Standards and Regulations 6
O Regarding design and development requirements for a medical device EU Medical Device Regulations 6
W Rights of Supplier in Design & Build WP Manufacturing and Related Processes 0
K ISO 13485 Design Requirement with respect to "component" manufacturers ISO 13485:2016 - Medical Device Quality Management Systems 11
rob3027 ISO 13485 justification for design and development exclusion for medical packaging ISO 13485:2016 - Medical Device Quality Management Systems 18
K What constitutes design under 9001? Design and Development of Products and Processes 15
P Product Requirements Control during Design Changes Design and Development of Products and Processes 4
B Minitab factorial design regression equation center point term Manufacturing and Related Processes 1
J Design Input Verification 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
P Self Certified Class 1 MDR 2017/745 Design Change CE Marking (Conformité Européene) / CB Scheme 2
N Design V&V courses Career and Occupation Discussions 11
B AS9100 Design Authority questions AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
K Process FMEA responsible For "Make to Design Parts" (Inhouse or Suppliers ?) FMEA and Control Plans 3
K CE design Certificate EU Medical Device Regulations 3
M ISO 13485 consultants and auditors with design oriented focus ISO 13485:2016 - Medical Device Quality Management Systems 7
T Design Transfer & use of "prototype" components in production devices. ISO 13485:2016 - Medical Device Quality Management Systems 10
Q What are Quality Requirements for Design? Quality Management System (QMS) Manuals 8
G Mfr. Process Validation BEFORE Design Transfer? Other Medical Device and Orthopedic Related Topics 1
B Use of Statistical Techniques in Design Verification Design and Development of Products and Processes 18
X Design stage overview (Product specification) EU Medical Device Regulations 3
gohyl Fixture Setup Design Manufacturing and Related Processes 3
D IS0 13485 - Design perspective to regulatory requirement. ISO 13485:2016 - Medical Device Quality Management Systems 3
T Software item classification and Detailed Design IEC 62304 - Medical Device Software Life Cycle Processes 4
armani 7.1.5 and design and development of product ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
Sidney Vianna ISO 9001 News ISO 9001 Design Specification to be developed - May 2022 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
C Design output vs acceptance criteria ISO 13485:2016 - Medical Device Quality Management Systems 8
M Design Control Procedure Medical Device and FDA Regulations and Standards News 4
lisap Design falling outside ISO scope Misc. Quality Assurance and Business Systems Related Topics 11
S Link Between Essential Performance Requirements and Essential Design Outputs 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
G Calculating Ppk for Design Verification - Variable Sampling Design and Development of Products and Processes 15
G How many Design FMEAs do you have? FMEA and Control Plans 4
G Updating FMEAs after Design Updates FMEA and Control Plans 4
Stoic Details of Operational Qualification (OQ) test design for plastic extrusion processes, effect of material property noise, and GHTF/SG3/N99-10 US Medical Device Regulations 2
H R&D and Design procedures - can one procedure cover both activities Design and Development of Products and Processes 6
J Design Verification Testing and Statistics Reliability Analysis - Predictions, Testing and Standards 3
K Can I make an exclusion of Design and Development in ISO 13485:2016 if my product is not regulated ISO 13485:2016 - Medical Device Quality Management Systems 12
S Audit Finding - Design History File (DHF) Index: few (3 to 4) reports not identified ISO 13485:2016 - Medical Device Quality Management Systems 3
S Is the Design Service Provider required to be ISO 13485 certified? ISO 13485:2016 - Medical Device Quality Management Systems 13
M How to show the effect of the failure mode on the manufacturing process as a customer of product design process? FMEA and Control Plans 3
C MDR - Annex II _3. DESIGN AND MANUFACTURING INFORMATION_a EU Medical Device Regulations 3
C Items used for Design Verification Design and Development of Products and Processes 7
S Software design document NMPA guidance and consultant China Medical Device Regulations 4