No Design - Which points from 7.3.1 'Design And Development' are necessary to comply

I

Internet-101

#1
7.3.1 Design and Development

7.3.2.1 Product Design Input
7.3.3.1 Product Design Output

Are these points from Design Responsibility exclusively?

This is because other ones (7.3.1, 7.3.1.1, 7.3.2, 7.3.2.2, 7.3.2.3, 7.3.3, 7.3.3.2, 7.3.4, 7.3.4.1, 7.3.5, 7.3.6, 7.3.6.1, 7.3.6.2, 7.3.6.3 & 7.3.7) have part of Design and Development Process of the Product.
If some company hasn’t a Design Responsibility but they need to introduce new products according ISO/TS 16949:2K terms on 7.3 “Design and Development” Element.
Will they need to comply with (7.3.1, 7.3.1.1, 7.3.2, 7.3.2.2, 7.3.2.3, 7.3.3, 7.3.3.2, 7.3.4, 7.3.4.1, 7.3.5, 7.3.6, 7.3.6.1, 7.3.6.2, 7.3.6.3 & 7.3.7) points only without 7.3.2.1 & 7.3.3.1 points without any non-conformance about it?
Or Which points from 7.3.1 “Design And Development” are necessary to comply If some company doesn’t have Design Responsibility?

I hope you can understand the situation.

Thanks in advance for your help.


Smile is good for your health!

:bigwave: :bigwave:
 
Elsmar Forum Sponsor
T

tattva

#2
Taking advatnage of this new threat, I´ve another question I hope you can help me to fully understand.

We don´t design a thing not even the manufacturing process, so do I still have to deal with other points regarding design such as competence 6.2.2.1?

Thanks in advance:bigwave:
 

Manoj Mathur

Quite Involved in Discussions
#3
Tattva,

It is hard to digest for me that You don't design(Moderate & Regulate) the manufacturing processes or the elements of it.

May I know what is your product and How do you make it?

Manoj Mathur
 
#4
Internet,
This is a difficult element to take exceptions since it applies to both the "Design and Development" of the product. At what point do you separate the design from the development? You may not be designing a product, but you may be developing a product. What is meant by development? Only you can determine that point.
IMO all of 7.3 requirenments would apply, except those explicitly related to the design of the product,i.e., design review.

Tattva,
7.3 Design and development applies to "products" and "processes". You may not design a process, but you may develop one.
As stated above, IMO all of 7.3 would apply, except those explicitly related to the deisgn of the product.
 
T

tattva

#5
Thank You all VERY MUCH!

I was talking to quality manager yesterday, and told me we can not make aside point 7.3, ´cause we do develop de manufacturing process.

We pour polyurethane, and what I meant by " not designing a thing " is that all technology and equipment is designed by coorporate engineering. But as you say, developing the process is another thing,we do that.
:biglaugh:
 

Howard Atkins

Forum Administrator
Staff member
Admin
#6
I have been pondering this problem and after some resarch including the thread at

http://Elsmar.com/Forums/showthread.php?s=&threadid=4602
I have to agree with Sam:
From ISO 9000:2000

3.4.4
design and development
set of processes (3.4.1) that transforms requirements (3.1.2) into specified characteristics (3.5.1) or into the
specification (3.7.3) of a product (3.4.2), process (3.4.1) or system (3.2.1)
NOTE 1 The terms “design” and “development” are sometimes used synonymously and sometimes used to define different
stages of the overall design and development process.
NOTE 2 A qualifier can be applied to indicate the nature of what is being designed and developed (e.g. product design and
development or process design and development).


3.4.1
process

set of interrelated or interacting activities which transforms inputs into outputs
NOTE 1 Inputs to a process are generally outputs of other processes.
NOTE 2 Processes in an organization (3.3.1) are generally planned and carried out under controlled conditions to add value.
NOTE 3 A process where the conformity (3.6.1) of the resulting product (3.4.2) cannot be readily or economically verified is
frequently referred to as a “special process”.


3.4.2
product
result of a process (3.4.1)
NOTE 1 There are four generic product categories, as follows:

— services (e.g. transport);
— software (e.g. computer program, dictionary);
— hardware (e.g. engine mechanical part);
— processed materials (e.g. lubricant).
7.3.1 must be in.
 
M

Mike Smith

#7
We are an electronics remanufacturing company. We did not develop or design the product that we remanufacture. We obtained a waiver from our customer stating that we did not have to comply to this requirement of the standard of QS9000. I am hoping to do the same for 16949 since we do not design the product or the MANUFACTURING process.
 

howste

Thaumaturge
Trusted Information Resource
#8
Mike, I would imagine that you do design/develop the remanufacturing processes. I believe you would need to include these design and development of these processes to meet 16949 requirements. From TS 16949 1.2:
The only permitted exclusions for this Technical Specification relate to 7.3 where the organization is not responsible for product design and development.
Permitted exclusions do not include manufacturing process design.
 
M

Mike Smith

#9
This brings up a good point. It also states in 1.2 "Where any requirement of this international standard cannot be applied due to the nature of an organization and its product, this can be considered for exclusion". Later in 1.2 it states "the only permitted exclusions for this TS relate to 7.3 where the organization is not responsible for product design and development". Does this contradict itself? First it states any requirement and then only requirement. Anyone?
 
#10
Mike S.
This is not a contradiction.
"Considered for exclusion" means you can ask for a waiver.
"Permitted exclusions", IATF permitted, means you do not have to ask for a waiver, only that you provide evidence to your CB that you do not design or develop product.
 
Thread starter Similar threads Forum Replies Date
V Handling open points in design reviews 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
R Understanding a few points on ISO 9001's Design and Development Planning ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
A Medical Device Electronic Systems Design - Points to be cleared EU Medical Device Regulations 2
A What factors must be considered during gauge design?what are VG points ? Design and Development of Products and Processes 1
Watchcat REGULATORY WATCHCAT De Novo Post-Mortem 4 – Design Controls Other US Medical Device Regulations 0
D Finding Optimum Design Parameters using Taguchi method? Using Minitab Software 1
S ISO13485:2016 7.3.9 design and development change ISO 13485:2016 - Medical Device Quality Management Systems 3
J Managing design verification regression testing of design changes Design and Development of Products and Processes 1
K Design History File - Looking for Examples of DHF's. 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
Watchcat Software validation vs design V&V? Other US Medical Device Regulations 27
S Design & Development records for Medical Devices Packaging and Labelling ISO 13485:2016 - Medical Device Quality Management Systems 8
D Using electronic lab notebooks in the design control process ISO 13485:2016 - Medical Device Quality Management Systems 3
L Essential Design Outputs for SAMD 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
T Justification of "Design Excluded" in IATF 16949 Design and Development of Products and Processes 5
H Contract Manufacturer as Design Owner ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
B Drawings - Design Rules Inspection, Prints (Drawings), Testing, Sampling and Related Topics 26
R Role of QA during Design and Development of Products and Processes Design and Development of Products and Processes 4
B Design Responsibilities for Mature Acquired Product Lines ISO 13485:2016 - Medical Device Quality Management Systems 3
I RWE Studies and Design Controls US Food and Drug Administration (FDA) 1
G How to record Regulations and Standards as Design Inputs? Design and Development of Products and Processes 15
T Design Traceability - Labelling / Instructions for Use ISO 13485:2016 - Medical Device Quality Management Systems 5
B How to exclude empty rows in full factorial design analysis? Using Minitab Software 0
M Handling design changes with use-as-is inventory disposition Quality Manager and Management Related Issues 6
M Design Control - Management of Electronic CAD files Design and Development of Products and Processes 14
S Examples of FDA acceptable Software Design Specification (SDS) Medical Device and FDA Regulations and Standards News 6
A Design Change/ECO Related Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
R Using R package to implement Bayesian phase I/II dose-finding design for three outcomes ISO 13485:2016 - Medical Device Quality Management Systems 6
C ISO 9001:2015 8.3.2. h) Design and Development Planning - What is required? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
Z Software for design control ISO 13485:2016 - Medical Device Quality Management Systems 5
E Design and Development file Procedure ISO 13485:2016 - Medical Device Quality Management Systems 0
DuncanGibbons Section 8.3 relevant for design organisations AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
P DFMEA - Machinery Design Best Practices FMEA and Control Plans 0
R Is a FAIR required on parts that we design? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
U API Spec Q1 - 5.6.1.2 C (3) - Design software Oil and Gas Industry Standards and Regulations 3
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
A 8.6 Release of products and services, 8.3 Design and development - evidence required ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
C Stress / Challenge Conditions for Design Verification Testing to Reduce Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
J Significant change related to design and intended use EU Medical Device Regulations 3
S Traceability of requirements to design and risk Design and Development of Products and Processes 3
U NOC - What is considered a "design change" EU Medical Device Regulations 5
Q PPT used as Design Review ISO 13485:2016 - Medical Device Quality Management Systems 3
D Design Verification Sample Size vs Repeats Statistical Analysis Tools, Techniques and SPC 9
A Design and development procedure for API Spec Q2 Oil and Gas Industry Standards and Regulations 6
D Design controls - Inputs, outputs, V&V, DHF, DMR ISO 13485:2016 - Medical Device Quality Management Systems 10
LostLouie Manufacturer divorced from Design process, is he justified in design process deficiencies? ISO 13485:2016 - Medical Device Quality Management Systems 9
R DFA & DFM - Examples for Design for assembly and design for manufacturability Lean in Manufacturing and Service Industries 2
D Using Laboratory Notebooks in R&D and Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 - 7.3.6 Design and development verification - Do most folks create a separate SOP? ISO 13485:2016 - Medical Device Quality Management Systems 6
K Joint approval between OEM and Manufacturer on Design Documents ISO 13485:2016 - Medical Device Quality Management Systems 4
M API 4F/7K/8C Design Package Validation Oil and Gas Industry Standards and Regulations 2

Similar threads

Top Bottom