Dear all,
I'm a hardware development engineer for polygraphic signal acquisition devices. New to this forum, so apologies up front if I have missed a thread that covered this already.
Identifying hazards frequently feels like kicking in open doors, but nothing surprises me anymore as even microwave manufacturers have to state that their equipment is not to be used for drying ones wet dog or cat...
Sub clause 9.1 has table 19, this is the only place in IEC-60601-1 where Entanglement HAZARD is mentioned. Trying to be thorough and equipped with a wild imagination I can foresee situations where people have ExG leads connected, or other applied parts, which could be highly uncomfortable when wrapped around ones neck, reducing blood circulation in limbs can be hazard enough.
The chances of entanglement with leads are not that remote, consider moving infants in incubators or with sleep diagnosis. So I would expect some warnings in the accompanying documentation of every lead manufacturer. But to my surprise I could not find any warning against entanglement hazards within the documentation of all competitors/colleagues/suppliers.
Is it because entanglement is not considered a hazard? Someone offered an explanation by stating that ?every medical professional gets instruction during their education how to use band aid to fixate applied parts, just to prevent getting hooked by a loose cable and to prevent entanglement. So plainly nobody sees it as a hazard?.
(But then I would expect a clear statement that ExG leads are only to be applied by well instructed personnel, which isn?t always the case either)
So what is it?
? No, applied part leads do not pose this mechanical hazard
? Yes, we should address this hazard in the RMF, and maybe in the user documentation.
I?d very much welcome your expert opinion.
I'm a hardware development engineer for polygraphic signal acquisition devices. New to this forum, so apologies up front if I have missed a thread that covered this already.
Identifying hazards frequently feels like kicking in open doors, but nothing surprises me anymore as even microwave manufacturers have to state that their equipment is not to be used for drying ones wet dog or cat...
Sub clause 9.1 has table 19, this is the only place in IEC-60601-1 where Entanglement HAZARD is mentioned. Trying to be thorough and equipped with a wild imagination I can foresee situations where people have ExG leads connected, or other applied parts, which could be highly uncomfortable when wrapped around ones neck, reducing blood circulation in limbs can be hazard enough.
The chances of entanglement with leads are not that remote, consider moving infants in incubators or with sleep diagnosis. So I would expect some warnings in the accompanying documentation of every lead manufacturer. But to my surprise I could not find any warning against entanglement hazards within the documentation of all competitors/colleagues/suppliers.
Is it because entanglement is not considered a hazard? Someone offered an explanation by stating that ?every medical professional gets instruction during their education how to use band aid to fixate applied parts, just to prevent getting hooked by a loose cable and to prevent entanglement. So plainly nobody sees it as a hazard?.
(But then I would expect a clear statement that ExG leads are only to be applied by well instructed personnel, which isn?t always the case either)
So what is it?
? No, applied part leads do not pose this mechanical hazard
? Yes, we should address this hazard in the RMF, and maybe in the user documentation.
I?d very much welcome your expert opinion.