Non-conformances whilst in the Design and Development stage of a project

neash83

Starting to get Involved
#1
Hello,

I am looking to get some thoughts and opinions on how people manage nonconforming product whilst in the Design and Development stage of a project.

At a design review stage if there is a problem with the design would you raise an NCR? Say it is decided the dimensions on the drawings need to be amended as two parts don?t quite fit together correctly (from prototypes or samples produced), how do people go about managing this change? Via an NCR? Separate actions from the design review? As the product isn?t technically ?live? does it constitute a non-conformance as the specification/drawing is yet to be released?

Thanks
 
Elsmar Forum Sponsor
P

PaulJSmith

#2
Re: Non-conformance's whilst in Design and Development

Absolutely not ... unless you have a bunch of unaccounted-for time you wish to spend on excessive and unnecessary paperwork.

Note it in the review as action needed, and move on to more important things.
 

somashekar

Staff member
Admin
#3
Re: Non-conformance's whilst in Design and Development

Hello,

I am looking to get some thoughts and opinions on how people manage nonconforming product whilst in the Design and Development stage of a project.

At a design review stage if there is a problem with the design would you raise an NCR? Say it is decided the dimensions on the drawings need to be amended as two parts don?t quite fit together correctly (from prototypes or samples produced), how do people go about managing this change? Via an NCR? Separate actions from the design review? As the product isn?t technically ?live? does it constitute a non-conformance as the specification/drawing is yet to be released?

Thanks
The intent of the review (at various stages of design and development) is to correct any errors and confirm that it works as intended at that stage.
So correction does happen at several stages, including inputs. However if you have set out to design a Camel and you are seeing Elephant, then more than a corrective action, you must see if competent persons are involved in the design and development phases.
 

NikkiQSM

Quite Involved in Discussions
#4
Re: Non-conformance's whilst in Design and Development

No NCR is needed. Design & Development is the process where adjustments, and failures can be expected. It is the process used to ensure there will be no failures / adjustments needed when the project does go live.

I would make a note of the changes / adjustments needed and proceed.
 
V

Vthouta

#5
Re: Non-conformance's whilst in the Design and Development stage of a project

Hi,

During the design and development stage there is no need for issuing NCR officially as the product is not been completely qualified for production.

But, I do notice that the review should be documented for future purposes. I would suggest to use a action register on simple spread sheet that captures Date, Issue found, who found, Action item status -open/closed.

This log could be useful while making the validation plans in future.

Vikram
 

kreid

Involved In Discussions
#6
If the drawings are formally controlled (have been issued / have a configuration ident) then I would say that some sort of formal change process is needed.

But as there other replies say, it does not constitute a non-conforming product as the product does not formally exist yet.
 
#7
Hello,

I am looking to get some thoughts and opinions on how people manage nonconforming product whilst in the Design and Development stage of a project.

At a design review stage if there is a problem with the design would you raise an NCR? Say it is decided the dimensions on the drawings need to be amended as two parts don?t quite fit together correctly (from prototypes or samples produced), how do people go about managing this change? Via an NCR? Separate actions from the design review? As the product isn?t technically ?live? does it constitute a non-conformance as the specification/drawing is yet to be released?

Thanks
Testing in a lab - prototype builds etc usually have some type of "build log" and even a a non-conformance report. It's going to depend on how many people are involved etc how "formal" this becomes. In my book, which was written by Xerox btw, errors are kept track of because it all erodes the development budget and time/materials are booked against that budget.
 
Q

QA_Newbie2000

#8
I agree with the consensus that a NCR is not required. As the name indicates a design review is just that, a review of the design to determine if it is correct. What we do, assuming the drawing/design is controlled /released is raise an Action Request (AR) that details what the problem is. The AR is evaluated and eventually becomes an Engineering Change Order (ECO). The intent of the ECO is to resolve the problem identified by the AR
 
#9
I agree with the consensus that a NCR is not required. As the name indicates a design review is just that, a review of the design to determine if it is correct. What we do, assuming the drawing/design is controlled /released is raise an Action Request (AR) that details what the problem is. The AR is evaluated and eventually becomes an Engineering Change Order (ECO). The intent of the ECO is to resolve the problem identified by the AR
If the details of the failure are recorded the same way, for input into the design function for them to take action, then what it's called is just a name. It's the principles of reporting failures etc. In fact, if we get into details, testing - which is likely to reveal such failures - plus any actions taken should become inputs to the design reviews. Rarely are people that sophisticated, so Engineering often make the same mistakes over and again.

AR/NCR doesn't matter what it's called, as long as you follow the right process.
 
Thread starter Similar threads Forum Replies Date
A Non-Conformances Found After 3rd Party Sorting Supplier Quality Assurance and other Supplier Issues 12
G Addressing Non-Conformances from an Internal Audit that are not product related ISO 13485:2016 - Medical Device Quality Management Systems 11
G Question about Non-conformances during New Product Introduction Nonconformance and Corrective Action 14
qualprod Internal Audits - Categories of non conformances Internal Auditing 12
Ed Panek Are audit non conformances also risk based? ISO 13485:2016 - Medical Device Quality Management Systems 1
E Example of 9001:2015 audit report with non-conformances ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
M Corrective action on Non-conformances from a long time ago ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
N Non-conformances due to Human Error Nonconformance and Corrective Action 64
D Consequences of Major Non-Conformances during a Registration Audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
E Non-Conformances and the Relevant ISO 9001 Paragraph Nonconformance and Corrective Action 4
P Losing Accreditation - JAS-ANZ has changed the conditions of non-conformances ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
T How many complaints, defects and non-conformances are typical? Customer Complaints 5
P Complaints vs. Non Conformances - When is a Complaint a Nonconformance? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
P Objecting to Registrar Audit Non-conformances General Auditing Discussions 43
K Cost of Supplier Non-conformances Nonconformance and Corrective Action 5
S Root Cause Analysis Taxonomy - Categorising the reasons for non-conformances ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
N Supplier Non Conformances - Format for a Supplier Rejection Form ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
L Do ALL non conformances have to be documented? Nonconformance and Corrective Action 1
V Management Review Meeting - Statistics for Customer Complaints and Non Conformances ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
C Internal Audit Concerns including Escalation of Non-Conformances General Auditing Discussions 6
D Non-Conformances in a Service based industry with no product Service Industry Specific Topics 9
P Interpreting the Process Gap - Open Non-Conformances & ISO Clause ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
M Pharmaceutical Non conformances SOP - Solid oral dose start up company Nonconformance and Corrective Action 2
B Internal Audits save you from Non-Conformances on external audits? General Auditing Discussions 16
B 19011 - Reporting Non-conformances: Grading of findings ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
S Examples of Non-conformances during ISO/TS 16949 pre/final audits IATF 16949 - Automotive Quality Systems Standard 2
S Priorizing Non-conformances - What is the best way? Nonconformance and Corrective Action 6
J QS-9000 System redesign - Several Major non-conformances - Where to start? QS-9000 - American Automotive Manufacturers Standard 8
D Clause 8.7 Control of Non conforming outputs AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
M Non conformance/complaint and CAPA ISO 13485:2016 - Medical Device Quality Management Systems 4
F Product classification according to the 60601-2-57: Particular requirements [...] of non-laser light sources [...] IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
Q API Spec Q1 9th Ed 4.4.2 Non-conformance Oil and Gas Industry Standards and Regulations 8
T Do I need to add non-product related service providers to my ASL? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 12
J ISO 13485- 8.3.1 Non-conforming material high volume ISO 13485:2016 - Medical Device Quality Management Systems 4
F MSA for non-contact CMM - ISO10360 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
M Using a Non Certified vendor Supplier Quality Assurance and other Supplier Issues 5
somashekar Justification for Non-applicable clause at Stage 2, in clause 4. ISO 13485:2016 - Medical Device Quality Management Systems 11
Z Swiss Authorized representative & non-medical device regulations Other Medical Device Regulations World-Wide 0
R Lead auditor scenario (Need help) is this non conforming to ISO13485 Manufacturing and Related Processes 2
D Help Me. Non conformitty in External Audit IATF 16949 - Automotive Quality Systems Standard 13
S Electronic Signatures - Non-Conformance - ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 30
S Do you follow your QMS for non-device software features? Medical Information Technology, Medical Software and Health Informatics 4
J Can we register non-device clinical decision support software under draft guidance? Other US Medical Device Regulations 5
D IATF 16949 SI 10, External non-accredited lab IATF 16949 - Automotive Quality Systems Standard 4
M Selling non-CE marked devices for evaluation EU Medical Device Regulations 4
T Non API products need to comply to API Q1? Oil and Gas Industry Standards and Regulations 3
L "Shelf-Life" Class I Non-Sterile Products Expiration Date? CE Marking (Conformité Européene) / CB Scheme 3
B Process / Procedure - Radiographic (X-Ray) Non-Film Document Control Systems, Procedures, Forms and Templates 0
N How to find column number for the last non-empty column? Using Minitab Software 0
A Need to calculate tolerance Intervals with a set of non-normal data and 3-Parameter Weibull distribution Using Minitab Software 0

Similar threads

Top Bottom