I have question about the number of lots required for stability testing of a Class II device in the 510(k). We have a request from the reviewer of our last submission for 3 lots used in testing shelf life. I cannot find anything to support his request other than the Drug guideline. Has anyone ever had a question about in relation to the number of lots needed for stability report in a 510(k)?
 
Elsmar Forum Sponsor

planB

Super Moderator
#2
Lilybef,

as you correctly pointed out, lot-based shelf life testing is a typical concept of pharmaceutical products. Is your product a combination device or a degradable device? Has the reviewer given you any indication why potential lot-based variances should be tested?

HTH,
Gerhard
 

Sam Lazzara

Trusted Information Resource
#3
Until now, I have not heard of FDA/ODE reviewers are suggesting that shelf life (stability) testing should be performed on multiple lots. I have seen some FDA guidance documents that suggest testing multiple lots for characteristics such as coating particulates, corrosion and a few other characteristics, but not in the context of shelf life testing.

Some "seemingly dated" FDA medical device guidance documents mention testing multiple lots for shelf life testing. For example:

FDA 1991 Guidance: Shelf Life of Medical Devices (attached)
- This is the only FDA device shelf life guidance that broadly covers all types of devices
- It refers to 1985 and 1989 FDA guidance documents for contact lenses that mention testing 3 lots, and 2 to 3 lots, respectively

More recent FDA device guidance documents I checked that discuss shelf life testing but do not mention how many lots to test:
- 2019 PTCA Catheters Guidance
- 2010 Non-Clinical Engineering Tests and Recommended Labeling for Intravascular Stents and Associated Delivery Systems Guidance
- 2018 Coronary, Peripheral, and Neurovascular Guidewires – Performance Tests and Recommended Labeling DRAFT Guidance
- 2018 Peripheral Vascular Atherectomy Devices - Premarket Notification [510(k)] Submissions DRAFT Guidance

The FDA-recognized pharma guideline for testing 3 lots/batches is mentioned in ICH Q1A
STABILITY TESTING OF NEW DRUG SUBSTANCES AND PRODUCTS

Also: Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations)

Data from formal stability studies should be provided on at least three primary batches of the drug substance.

Other documents I checked that do not mention testing multiple lots/batches:
- ASTM F2914 2012 SG for Identification of Shelf-life Test Attributes for Endovascular Devices (interesting since it provides examples of attributes typically not impacted by aging (e.g.. stent/hard plastic dimensions, MR compatibility, marker radiopacity, corrosion resistance)
- EN ISO 11607-1 2017 Packaging for terminally sterilized medical devices-Part 1-Requirements for materials, sterile barrier systems and packaging systems (ISO 11607-1 2006+A1 2014)

Note: "Stability" is a term that is mostly used in the pharma industry, whereas "shelf life" is the term I have seen primarily used for medical devices. However, 11607-1 does use the term "stability" and does not use "shelf life".
 

Attachments

Last edited by a moderator:

lilybef

Involved In Discussions
#4
Hi Sam and Plan B,

My medical device is not a combination product nor is made of degradable materials. My team is baffled on the requirement for 3 lots for stability/shelf life.
Thanks for the inputs!
 
Thread starter Similar threads Forum Replies Date
Jane's Like-for-like critical raw material change qualification - type of testing/ number of lots required ISO 13485:2016 - Medical Device Quality Management Systems 4
D Number of Manufacturing Lots for Process Validation Manufacturing and Related Processes 10
M Zero Acceptance Number Sampling Plans for Isolated Lots Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
F Determining the Number of Samples for receiving large lots of wire? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
B Document Issued with incorrect Document number code Document Control Systems, Procedures, Forms and Templates 2
M Are you allowed to reuse the same asset number if the device was replaced? General Measurement Device and Calibration Topics 10
D Control Number for Class III and IV devices Canada Medical Device Regulations 0
K How to handle GTINs for different configurations of one device with one part number? Other Medical Device Related Standards 0
K How can Acceptance Number and Reject Number be larger than Sample Size on Z1.4 Table? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
D Determining the the maximum number of reprocessing cycles of attachments CE Marking (Conformité Européene) / CB Scheme 2
K Why 'FD&C act section number' and 'section number' in the title of the act are different? US Food and Drug Administration (FDA) 4
A Risk Number for each software requirement IEC 62304 - Medical Device Software Life Cycle Processes 7
K Adding the Notified Body Number CE Marking (Conformité Européene) / CB Scheme 2
C Quantifying risk in choosing the number of parts, operators and replicates in a GR&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Q LOT or Serial Number Symbol not used when the information is contained in the UDI? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Proud Liberal Power Query to combine multiple tables of unequal number of columns Statistical Analysis Tools, Techniques and SPC 1
Q How is Medical Device Number (MDL) assigned to companies? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
D How to get EUDAMED Single Registration Number (SRN) EU Medical Device Regulations 17
M Informational US FDA – Modifications to the List of Recognized Standards, Recognition List Number: 052 Medical Device and FDA Regulations and Standards News 0
M Nonconformity for missing form number on the job description document Nonconformance and Corrective Action 1
I Number of decimals in equation calculated by MiniTab Using Minitab Software 0
F 2017/745 Article 31 Single Registration Number Medical Device and FDA Regulations and Standards News 5
Z Do we still have to put UDI if we have DI and lot number seperately on our product ? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
Z Two Payment Identification Number (PIN) for the same order in DFUF website 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Ed Panek GTIN (Global Trade Item Number) for non CE Mark Other Medical Device Related Standards 4
A The latest revision number and changes in automotive core tools IATF 16949 - Automotive Quality Systems Standard 1
G Belgium - How do I get an enterprise CBE Number Other Medical Device Regulations World-Wide 0
M Document Control - Revision Number in Document Names Document Control Systems, Procedures, Forms and Templates 4
M Informational USFDA – FDA Modernization Act of 1997: Modifications to the List of Recognized Standards, Recognition List Number: 051 Medical Device and FDA Regulations and Standards News 0
S How to number and control translated SOPs - Three languages Document Control Systems, Procedures, Forms and Templates 4
qualprod Calculate number of people in processes, considering cycle time/takt time = quantity of people Lean in Manufacturing and Service Industries 9
A Doubt on multiple inspection carrying out for same balloon number during Stage Inspection and Final Inspection? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
K Relabeling or Not Relabeling - Adding our internal Part Number barcode label 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
J DoE - Number of runs in Plackett-Burman (Minitab) Using Minitab Software 2
G Heat Number on CofC - Stainless steel tube Manufacturing and Related Processes 2
M Determining number of employees within the "Scope" of the QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
qualprod What tier number fits in a company producing paper labels stuck on automotive parts? IATF 16949 - Automotive Quality Systems Standard 7
M Medical Device News TGA Consultation: Changes to a number of definitions and the scope of the medical device regulatory framework in Australia Medical Device and FDA Regulations and Standards News 0
J Question on very low NDC number with tolerable GRR ratio's Capability, Accuracy and Stability - Processes, Machines, etc. 7
S Is Lot Number required in addition to UDI serial number? Other Medical Device Regulations World-Wide 7
T Change control and configuration management - When to create a new model/part number? Other Medical Device and Orthopedic Related Topics 0
D IMDS for machining - Can my customer request a IMDS number for a part? APQP and PPAP 2
shimonv GTIN Number on the Packaging of a Serviced Medical Device EU Medical Device Regulations 0
W Is the RPN (risk priority number) in the PFMEA really a RPN without the detectability ISO 14971 - Medical Device Risk Management 4
M How to count the number of employees? Other Medical Device and Orthopedic Related Topics 4
bobdoering "nds" or Number of Discriminate Samples - the Necessary Tool to Work With "ndc"! Imported Legacy Blogs 0
H FDA product code and regulation number Other US Medical Device Regulations 3
P FDA's DRLM (Device Registration & Listing Module) Listing Number Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
Sidney Vianna High number of certificate suspensions in the IAQG OASIS database AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
J GLP Animal Study - Change number of treatment sites Other US Medical Device Regulations 2

Similar threads

Top Bottom