Number of MOPP for Type BF Applied Part (Medical Device Electronics)

S

s_g_robertson

Hi,

I'm trying to understand the requirements for MOPP. The device I am considering is a battery (~3V) powered device with multiple patient connections(ecg electrodes) considered as one Type BF applied part.

We clearly need to meet the leakage current requirements of 8.4.2 a) and the expectation is that we would use a series impedance to limit the current that could flow to the patient connection. No current is intentionally supplied to the patient so this would only be to protect against an internal fault that applied a voltage to the patient connection circuitry.

So we would be single fault safe and this resistor would be one MOPP. What I cannot think is how to get the second MOPP?

This must be a very common scenario, what am I missing?
Thanks
Stephen

8.4.2 ACCESSIBLE PARTS and APPLIED PARTS
a) The currents from, to or between PATIENT CONNECTIONS shall not exceed the limits for PATIENT LEAKAGE CURRENT and PATIENT AUXILIARY CURRENT specified in Table 3 and Table 4 when measured as specified in 8.7.4.

8.5.1.1 General
ME EQUIPMENT shall have two MEANS OF PROTECTION to prevent APPLIED PARTS and other ACCESSIBLE PARTS from exceeding the limits specified in 8.4.
 
H

hcrook

You should be able to meet 2MOPP with Creepages/Clearances, see Subclause 8.9.
 
Last edited by a moderator:

Peter Selvey

Leader
Super Moderator
The MOP (MOOP and MOPP) and mainly written around electric shock from mains parts, and can be somewhat hard to read or not appropriate when considering shock from internal circuits.

In the case of an ECG, there is usually a concern for the right leg drive (RLD or any noise cancellation circuit). Usually, all measurement leads (LA, LL, RA etc) are high impedance op-amps so there is negligible leakage in normal use. But for example if one of the input protection diodes shorts to rail, the RLD will go opposite to rail. To protect this the RLD normally has a large resistor like 1M (e.g. 3V/1M = 3uAdc which << 50uA limit in SFC).

The question then is whether this is a 2 MOP system (1 MOP = diode; 1 MOP = resistor). But diode is just one of many possible faults, and difficult to characterise and control as an MOP. So it is better to think of the 1M resistor as a high integrity component (never fails, equivalent to 2 MOP). It's OK because the 1M resistor is rated to 100V but used at 3V, clearly never going to fail. Or to be extra safe you could break the 1M into 470K + 470k to make a real 2 MOP system.

And if you wanted to go crazy, measure spacing (creepage/clearance) around the resistor. But this is a point where 601-1 is fundamentally wrong, the 3.4mm for 2MOPP is way over the top. The value in the standard assumes overvoltage transients which can still occur between secondary circuits, but they can't exist at the board level inside the circuit.

Test labs don't normally go that deep and may not even consider the risk from a RLD circuit. Or maybe short the "RA input diode D4" but just report the result without figuring out what provides the protection.

If you have respiration (using higher frequency impedance measurement) again you need to look at the circuit, figure out where the shock could come from, what would be worst case fault, and what component keeps things under the limits. Often it is a low value capacitor in series with high frequency oscillator (e.g. 60kHz). Again you might treat the cap as a high integrity component or make two in series to create a 2 MOP system.

If there is any other applied part e.g. temp probe, they usually have plastic barrier which can be tested for 500V to make 2 MOPP.
 

kevinkre

Registered
Hi Peter,

Thanks so much for your help! We are going to go with your advice and do two 470k on RLD to be extra safe as you recommend. However, we want to stay with our 0402 resistor size which is 50V not 100V. I think we can say it is clearly not going to fail in a battery powered device with 3.3V.

Please reply if you think this is not good...either the small footprint resistor or the 50V rating.

All the best,

Kevin
 
Thread starter Similar threads Forum Replies Date
M Number of samples for MDSW validation EU Medical Device Regulations 3
A Lot number on label vs lot number on device - Traceability ISO 13485:2016 - Medical Device Quality Management Systems 0
B Sample Size vs Number of Measurements Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
2tall Multiple Frequencies within 1 Gauge Number file?? Using GAGEpack Software 0
M Number of decimal places for measuring General Measurement Device and Calibration Topics 3
M Material Heat Number / Lot Number ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
S Adding new product code/Catalogue number to existing 510K US Medical Device Regulations 0
J Average number of Nonconformances during internal quality audit for Medical Device Manufacturers Internal Auditing 3
I Restricting scope to one part number for one customer IATF 16949 - Automotive Quality Systems Standard 6
T EN 17141 - number of monitoring locations Other Medical Device Related Standards 2
renenatasha Single Registration Number (SRN) EU Medical Device Regulations 2
Ed Panek ANNUAL FDA Registration - What 510(k) number to use? US Food and Drug Administration (FDA) 1
B How to change hour format to text but not changing the number ? Excel .xls Spreadsheet Templates and Tools 1
Y CE marking with NB identification number EU Medical Device Regulations 0
H Notified Body Number with CE Mark CE Marking (Conformité Européene) / CB Scheme 4
L How to deal with the model number symbol for X-RAY PRODUCT Other Medical Device Related Standards 1
R UDI Number Identification EU Medical Device Regulations 5
G Number of Destructively Tested Devices Needed for Ethylene Oxide Validation Other Medical Device Related Standards 4
D Number of people to be interviewed during an internal audit? Internal Auditing 10
S How to calculate Effective Number of People for ISO 13485 Certification? General Auditing Discussions 2
Ch00Ch00 Evaluate a process - Issue number Lean in Manufacturing and Service Industries 7
N How to find column number for the last non-empty column? Using Minitab Software 0
S Business Number versus Company ID? Canada Medical Device Regulations 3
B Document Issued with incorrect Document number code Document Control Systems, Procedures, Forms and Templates 3
M Are you allowed to reuse the same asset number if the device was replaced? General Measurement Device and Calibration Topics 10
D Control Number for Class III and IV devices Canada Medical Device Regulations 1
K How to handle GTINs for different configurations of one device with one part number? Other Medical Device Related Standards 0
K How can Acceptance Number and Reject Number be larger than Sample Size on Z1.4 Table? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
D Determining the the maximum number of reprocessing cycles of attachments CE Marking (Conformité Européene) / CB Scheme 2
K Why 'FD&C act section number' and 'section number' in the title of the act are different? US Food and Drug Administration (FDA) 7
A Risk Number for each software requirement IEC 62304 - Medical Device Software Life Cycle Processes 7
K Adding the Notified Body Number CE Marking (Conformité Européene) / CB Scheme 2
C Quantifying risk in choosing the number of parts, operators and replicates in a GR&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Jane's Like-for-like critical raw material change qualification - type of testing/ number of lots required ISO 13485:2016 - Medical Device Quality Management Systems 4
Q LOT or Serial Number Symbol not used when the information is contained in the UDI? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Proud Liberal Power Query to combine multiple tables of unequal number of columns Statistical Analysis Tools, Techniques and SPC 1
Q How is Medical Device Number (MDL) assigned to companies? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
D How to get EUDAMED Single Registration Number (SRN) EU Medical Device Regulations 19
M Informational US FDA – Modifications to the List of Recognized Standards, Recognition List Number: 052 Medical Device and FDA Regulations and Standards News 0
M Nonconformity for missing form number on the job description document Nonconformance and Corrective Action 1
I Number of decimals in equation calculated by MiniTab Using Minitab Software 0
F 2017/745 Article 31 Single Registration Number Medical Device and FDA Regulations and Standards News 5
Z Do we still have to put UDI if we have DI and lot number seperately on our product ? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
Z Two Payment Identification Number (PIN) for the same order in DFUF website 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Ed Panek GTIN (Global Trade Item Number) for non CE Mark Other Medical Device Related Standards 4
A The latest revision number and changes in automotive core tools IATF 16949 - Automotive Quality Systems Standard 1
G Belgium - How do I get an enterprise CBE Number Other Medical Device Regulations World-Wide 0
M Document Control - Revision Number in Document Names Document Control Systems, Procedures, Forms and Templates 4
M Informational USFDA – FDA Modernization Act of 1997: Modifications to the List of Recognized Standards, Recognition List Number: 051 Medical Device and FDA Regulations and Standards News 0
S How to number and control translated SOPs - Three languages Document Control Systems, Procedures, Forms and Templates 4

Similar threads

Top Bottom