Hi,
I'm trying to understand the requirements for MOPP. The device I am considering is a battery (~3V) powered device with multiple patient connections(ecg electrodes) considered as one Type BF applied part.
We clearly need to meet the leakage current requirements of 8.4.2 a) and the expectation is that we would use a series impedance to limit the current that could flow to the patient connection. No current is intentionally supplied to the patient so this would only be to protect against an internal fault that applied a voltage to the patient connection circuitry.
So we would be single fault safe and this resistor would be one MOPP. What I cannot think is how to get the second MOPP?
This must be a very common scenario, what am I missing?
Thanks
Stephen
8.4.2 ACCESSIBLE PARTS and APPLIED PARTS
a) The currents from, to or between PATIENT CONNECTIONS shall not exceed the limits for PATIENT LEAKAGE CURRENT and PATIENT AUXILIARY CURRENT specified in Table 3 and Table 4 when measured as specified in 8.7.4.
8.5.1.1 General
ME EQUIPMENT shall have two MEANS OF PROTECTION to prevent APPLIED PARTS and other ACCESSIBLE PARTS from exceeding the limits specified in 8.4.
I'm trying to understand the requirements for MOPP. The device I am considering is a battery (~3V) powered device with multiple patient connections(ecg electrodes) considered as one Type BF applied part.
We clearly need to meet the leakage current requirements of 8.4.2 a) and the expectation is that we would use a series impedance to limit the current that could flow to the patient connection. No current is intentionally supplied to the patient so this would only be to protect against an internal fault that applied a voltage to the patient connection circuitry.
So we would be single fault safe and this resistor would be one MOPP. What I cannot think is how to get the second MOPP?
This must be a very common scenario, what am I missing?
Thanks
Stephen
8.4.2 ACCESSIBLE PARTS and APPLIED PARTS
a) The currents from, to or between PATIENT CONNECTIONS shall not exceed the limits for PATIENT LEAKAGE CURRENT and PATIENT AUXILIARY CURRENT specified in Table 3 and Table 4 when measured as specified in 8.7.4.
8.5.1.1 General
ME EQUIPMENT shall have two MEANS OF PROTECTION to prevent APPLIED PARTS and other ACCESSIBLE PARTS from exceeding the limits specified in 8.4.