Search the Elsmar Cove!
**Search ALL of Elsmar.com** with DuckDuckGo Especially for content not in the forum
Such as files in the Cove "Members" Directory
Social Distancing - It's not just YOUR life - It's ALL of OUR lives!
Me <——————— 6 Feet ———————-> You
On going Stability for IVDs
ernieto
Starting to get Involved
#1
Don't know if this is the right forum.
Is there any requirement for ongoing or post-approval stability for IVDs? Looking for what is the requirement through FDA and EU? Thanks
ernieto
#4
Thanks for the reply. I have that document and I also use ISO 23640:2011 In vitro diagnostic medical devices: Evaluation of stability. Do you now where FDA or EU mentions that is required?
invitro_spain
Involved In Discussions
#5
Hi Ernieto,
I think the best standard for the US is the CLSI EP25a as QAengineer said. In any case, the relationship between the CLSI and the European Harmonized standards is acceptable.
QAengineer13
Quite Involved in Discussions
#6
Its an implied requirement from the FDA /EU or any regulatory agency because the Requirement for PERFORMANCE OF SAFE & EFFECTIVE medical device/IVD is part of the essential requirements, so I would say the stability is one of the key requirements that could be used to strengthen your performance claims of the marketed device.