One function, many devices? IEC 62304 and Medical Device Imaging Software

S

SayDo

#1
Hi! I'm a first time "poster" and I'd like to first of all thank everyone in this forum - there is some really great stuff posted here!

I've just been hired into a software development company that develops common filter and image enhancement software libraries for use in a variety of medical imaging equipment.

We want this software to be IEC62304 compliant to reduce any obstacles our clients may have in wanting to adopt our technology.

But when I read IEC62304, I'm confused because it seems to apply to software that is being developed as part of a development program for a single medical device.

And since out software is intended for medical devices, it doesn't fit the definition of SOUP.

Does IEC62304 apply to this kind of software?:confused:
 
Elsmar Forum Sponsor

Marc

Hunkered Down for the Duration with a Mask on...
Staff member
Admin
#2
Does anyone here have any ideas or help for this one? My Thanks in advance!
 

glork98

Involved In Discussions
#3
One issue is how your customers choose to view your software. They can consider it SOUP in which case they need to test it and place appropriate hazard mitigation strategies around it. If you develop to the standard then your customers will find it more valuable. ($$$)

You can view your product as the library. Something to determine is the class of your software and the class of the final products it will be put in. No safety hazard, or "Class A," under 62304 is roughly nothing in term of process requirements. OK, I'm exaggerating but only a bit. If you want to be "B" (patient and operator safety critical) you'll be valuable to customers making most medical devices.

I don't think you're library will be involved in life supporting products (Class "C") but it's a fairly small step to that from "B."
 

Marcelo

Inactive Registered Visitor
#4
Hello Saydo and welcome to the Cove!

IEC 62304 is for embedded software and also for software which is a medical device in itself (which seem to be the case of yours).

MEDICAL DEVICE SOFTWARE
SOFTWARE SYSTEM that has been developed for the purpose of being incorporated into the MEDICAL DEVICE being developed or that is intended for use as a MEDICAL DEVICE in its own right

Even if this wasn´t the case, IEC 62304 does provide a general framework for software that medical device/healthcare software developers could use if needed. People sometimes are too worried about the bureocratic nature of standards (for example, the scope or being applicable or not) that they sometimes forget that standards are nothing more than collections of information from the literature and best practices :)
 
Last edited:

Marcelo

Inactive Registered Visitor
#6
Can you explain how those standards apply? One is usability and the other is nuclear-specific.
Woops, my mistake.

It should be

"IEC 62304 is for embedded software and also for software which is a medical device in itself (which seem to be the case of yours)."

Thanks!

Er..didn´t understand what you said about nuclear-specific.
 
S

SayDo

#8
Thanks! THese are good things to think about.

When we started, we didn't sweat the scope (are we medical device or SOUP) of IEC62304 so much. We did our development as close to IEC practices as we could. But, there were obvious limitations - we can't do a full risk analysis because our "libraries" must be managed by the software that the manufacturer develops, we don't know all the intended uses of their products, etc. We figured Class B was appropriate for us.

Its just that now, we want to claim compliance but... the "lifecycle" causes some issues. Our technology is produced far in advance of its use in a medical device. We perceive that the manufacturer may consider us "SOUP", but we want to lessen that burden - this is where we add value!

SOFTWARE SYSTEM that has been developed for the purpose of being incorporated into the MEDICAL DEVICE...
And - yes, our software is intended to be embedded in a medical device, but we don't know which one!

Sooo... Can we claim compliance? Will the Medical Device manufacturer be able to claim compliance if they use our software and don't consider it SOUP?
 

glork98

Involved In Discussions
#9
Actually you can do a risk analysis. In it you would identify mitigation actions that the user needs to take.

If you have artifacts that demonstrate that you developed the code in compliance with 62034 then your customers won't need to call your code SOUP and do other analysis and risk mitigation around your code. This is actually harder than it sounds as, I believe, your company should be audited for compliance.

Perhaps a path is to supply all your design docs, source code and test code to the customers and let them include it in their development process. They don't write anything but they review, compile, test and control the code themselves. High buck RTOSes take this route.
 

sagai

Quite Involved in Discussions
#10
I would argue a bit, sorry, from the Customer prospective, I believe it will be a SOUP/OTS to validate prior to build in the Customer's finished medical device as long as it had researched and developed in a separate QMS.
br
Sz.
 
Thread starter Similar threads Forum Replies Date
W Pfmea function requirement and failure mode FMEA and Control Plans 6
P Search function minimum length Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 3
K Multiple function device products EU Medical Device Regulations 0
D Question regarding ECO process, specifically for Life Science products and defining form fit and function ISO 13485:2016 - Medical Device Quality Management Systems 1
shimonv FDA News FDA guidance on Multiple Function Device Products (8/2020) Other US Medical Device Regulations 1
D DQ documents of new equipment - Function specification Qualification and Validation (including 21 CFR Part 11) 0
J Auditing of Support Function IATF 16949 - Automotive Quality Systems Standard 9
K Measuring Function MEDDEV 2.1/5 relation with MDD 93/42/EEC CE Marking (Conformité Européene) / CB Scheme 1
T Spirometer - Pulmonary Function Testing - ATS/ERS:2005 EU Medical Device Regulations 5
M How to complete structure Analysis and Function analysis sections' columns of AIAG-VDA DFMEA form FMEA and Control Plans 0
8 Need Help - Runout - Function Gage Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
C Function definition for DFMEA FMEA and Control Plans 1
T AS9100D - Scope of QMS for New Company - Only Choosing a Function Subset Due to Management AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
C FMEA - Multiple function failures considerations FMEA and Control Plans 3
S Status products when/if the QMS ceases to function? EU Medical Device Regulations 1
M New Medical Device Accessory with New Function - Is this a major change? Other Medical Device Regulations World-Wide 3
W Is sharing resources treated same way as remote function? IATF 16949 - Automotive Quality Systems Standard 1
H Conveying Form Fit and Function Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
G Function in french called "numbre de posage(s)" - CMK Question Quality Tools, Improvement and Analysis 3
M Remote Office only has Invoice Function - Audit Needed? IATF 16949 - Automotive Quality Systems Standard 0
J FAI per AS9102 affecting Fit, Form, Function AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
MDD_QNA What is the function of the Critical Component List? ISO 14971 - Medical Device Risk Management 2
M Cross Function Process Map for Document Control Process Maps, Process Mapping and Turtle Diagrams 3
F Which Business Function should Approve Suppliers and maintain the ASL Supplier Quality Assurance and other Supplier Issues 8
L How to Introduce the Quality Function to the production floor in a bad environment Misc. Quality Assurance and Business Systems Related Topics 6
A ISO 27001 function wise or department wise audit questionnaire with control & clauses IEC 27001 - Information Security Management Systems (ISMS) 3
M Taguchi Loss Function Problem Misc. Quality Assurance and Business Systems Related Topics 2
C IVDD 98/79/EC Measuring Function Requirements EU Medical Device Regulations 2
N Regulatory and Safety Standards for Medical Device with Measuring Function IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
M Where to start Maintenance Function Kaizen Lean in Manufacturing and Service Industries 4
L Remote Control Inspection Digital Indicator with a TIR (range) function Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
J CE Mark Classification for a a pulmonary function test filter CE Marking (Conformité Européene) / CB Scheme 2
S Outsourcing PPAP Function APQP and PPAP 8
F What is an FAA PAI Maintenance Function List? Federal Aviation Administration (FAA) Standards and Requirements 3
Z Scope of Employee and Function Levels of Internal Audits Internal Auditing 3
T Difference between the terms Function & Process when used in Construction Management Business Continuity & Resiliency Planning (BCRP) 12
N Role of Quality Function in Approving Work Instructions Document Control Systems, Procedures, Forms and Templates 4
C What can you do if Quality is treated as a separate function in an organization? Quality Manager and Management Related Issues 24
V Annual Performance Appraisal - Person vs. Function/Business Career and Occupation Discussions 5
B Which type of DOE was likely used for this transfer function? Six Sigma 4
A Function for Effects - Beginner in usage of the IQRM FMEA and Control Plans 3
P Nonlinear Fit for a Complicated Curve - Fitting a Function to the Data Statistical Analysis Tools, Techniques and SPC 6
V SOPs and Internal Audit Criteria for Regulatory Affairs (Submissions Function) US Food and Drug Administration (FDA) 4
V What kind of Data should support Fit/Function Characteristics? APQP and PPAP 1
S Probability Density Function of a Function of a Random Variable Statistical Analysis Tools, Techniques and SPC 13
J Definition Form Fit or Function - Materials Composition and Manufacturing Processes? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 2
J Function of a Change Number in our ECN / document vault program as a tie-in Problem Solving, Root Cause Fault and Failure Analysis 3
M Quality System Procedure for HR Function needed Document Control Systems, Procedures, Forms and Templates 4
S Critical to manufacturing but not product function - Non-conformance or Not? Nonconformance and Corrective Action 12
M TS 16949 and Form, Fit or Function? What about Calibration of Oven Controllers? IATF 16949 - Automotive Quality Systems Standard 5

Similar threads

Top Bottom