We have changed the appreance and PCB of a originally 510(k) cleared device. Should we submit a new 510(k), for example, a special 510(k), or we just need to document our change in our design control documentation.
I have noticed that there are a guidance document"Deciding when to submit a 510(k) for a change to an exsiting device",and in the flow chart B, especially the B4-B8, i find that if we didn't change our IFU, and if cilinical data is not needed, and deisgn verification didn't rise new problem of safety and effectiveness, then we should not submit a new 510(k)?
And i think our device fit the criteria mentioned above.So a documentation is enough, is that true? If it is true, do we have to notice FDA before our new device coming into market?
Thanks a lot.
I have noticed that there are a guidance document"Deciding when to submit a 510(k) for a change to an exsiting device",and in the flow chart B, especially the B4-B8, i find that if we didn't change our IFU, and if cilinical data is not needed, and deisgn verification didn't rise new problem of safety and effectiveness, then we should not submit a new 510(k)?