Dear fellows,

In the Fluke calibrator's operator manual it is stated as follows:

Absolute Uncertainty Definition
The 5500A specifications include stability, temperature coefficient, linearity, line and load regulation, and the traceability of the external standards used for calibration. You do not need to add anything to determine the total specification of the 5520A for the temperature range indicated.

So, will it have any effects while estimating the Type B uncertainty?

Thanks and looking forward.
 
Elsmar Forum Sponsor

dwperron

Trusted Information Resource
#2
Dear fellows,

In the Fluke calibrator's operator manual it is stated as follows:

Absolute Uncertainty Definition
The 5500A specifications include stability, temperature coefficient, linearity, line and load regulation, and the traceability of the external standards used for calibration. You do not need to add anything to determine the total specification of the 5520A for the temperature range indicated.

So, will it have any effects while estimating the Type B uncertainty?

Thanks and looking forward.
The specifications give you most everything you need as uncertainty contributors for the calibrator. A couple of fine points:
As they point out these numbers only apply for the quoted temperature range. If you operate outside of that range (unlikely) you will need to add temperature compensation as a contributor.
It also assumes the you had Fluke calibrate your 5520A. If another lab calibrated it you will have different calibration uncertainty contributors that you will need to account for.

Another thing that Fluke has done is this:
" Specification Confidence Interval >99% "
That means their uncertainties are quoted with a k=3 multiplier, not the traditional k=2 that you normally report. So when you are combining uncertainties you need to divide these by 3, not 2, to get from expanded uncertainty to uncertainty.
 

fahimk

Involved In Discussions
#3
The specifications give you most everything you need as uncertainty contributors for the calibrator. A couple of fine points:
As they point out these numbers only apply for the quoted temperature range. If you operate outside of that range (unlikely) you will need to add temperature compensation as a contributor.
It also assumes the you had Fluke calibrate your 5520A. If another lab calibrated it you will have different calibration uncertainty contributors that you will need to account for.

Another thing that Fluke has done is this:
" Specification Confidence Interval >99% "
That means their uncertainties are quoted with a k=3 multiplier, not the traditional k=2 that you normally report. So when you are combining uncertainties you need to divide these by 3, not 2, to get from expanded uncertainty to uncertainty.
Thanks for the elaborate response and for highlighting the confidence interval thing.

Yes, it's not Fluke but another lab that calibrated. The prescribed temperature and environmental factors have been taken into account by the very lab. So, in that case, what factors, in general, need to be taken into account?

In one of Fluke's guide about uncertainty for DMM (http://download.flukecal.com/pub/literature/webinar-uncertainty-presentation-Dec%202011.pdf) the Uncertainty Type B is limited to accuracy and resolution.
 

dwperron

Trusted Information Resource
#4
This isn't really an easy answer.

In order to correctly do this, you will need to know the difference between the uncertainty of the standards that Fluke used in the calibrations and the uncertainty of your calibration source. You can do that by comparing the uncertainties provided by Fluke for their calibration results to those provided by your provider. You would have to add the differences in these uncertainties to the published "uncertainties" in the Fluke manual.

This can be a significant difference. Even in good calibration labs almost all of the time their measurement uncertainties will be higher than Fluke's because they also go through Fluke for calibrations and then add their own level of uncertainty to those results. In really bad cases you will find the occasional lab that will calibrate with the best instruments that they have available, even if they are less accurate than your calibrator. In a case like that your uncertainties will drastically increase.

Or you can do what nearly every other lab does. You don't use the Fluke numbers as "uncertainties", but instead use them as "specifications". Then you will normalize them assuming a rectangular distribution when you use them in your uncertainty budget as a Type B contributor.

Speaking of Type B contributors, looking over the Powerpoint you attached they are saying that in the calibration of a DMM that the significant Type B contributors will be your calibrator accuracy and the resolution of the DMM. That is true in the vast majority of cases, but not a "rule". For instance, if the DMM measures temperature and has an internal cold junction calibration (CJC) the temperature range where the DMM is used might need to be addressed.
 
Thread starter Similar threads Forum Replies Date
shrutisancheti EU User manual / operator manual / service manual guidance document(s) CE Marking (Conformité Européene) / CB Scheme 2
B Operator*part interaction - AIAG MSA manual indicates ANOVA Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
S Statistics to define the capability of operator controlled manual machines Statistical Analysis Tools, Techniques and SPC 9
H EU MDR Economic Operator notifications? EU Medical Device Regulations 1
M Inter-operator Variability Testing - Requirements for EU Medical Device Regulations 5
B Operator protection - When to apply table 7, Dielectric strength test voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
P Economic operator responsibilities - when the Importer is also Distributor - Article 13 EU Medical Device Regulations 1
P How far an operator can reach into a machine before it becomes an ergonomic problem CE Marking (Conformité Européene) / CB Scheme 5
R Operator's manuals as prevention design control FMEA and Control Plans 2
M Informational EU – M2M Data Exchange available services for accessing MDR EUDAMED data available for Economic Operator (EO) organisations Medical Device and FDA Regulations and Standards News 0
G 3 Operator Std. Dev. and uncertainty calculation Measurement Uncertainty (MU) 2
T Root causes - operator's actions Problem Solving, Root Cause Fault and Failure Analysis 17
K Article 23 (Parts and components) - Economic Operator EU Medical Device Regulations 1
GoSpeedRacer How to handle operator involvement in the non conformance ISO 13485:2016 - Medical Device Quality Management Systems 5
Fender1 Operator process buy-in - how to get it? Manufacturing and Related Processes 6
K Operator Checks - How to show that they were completed on a checklist as acceptable IATF 16949 - Automotive Quality Systems Standard 2
C ISO 9001:2015 - Non Applicable Design and Development (Power Plant Operator) Design and Development of Products and Processes 2
A Failure Mode from Operator "Components Fallen but Loaded" FMEA and Control Plans 3
A 21CFR807.25 (d) Owner-Operator Contact question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
U USB Connector Operator Protection (MOPP) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
somashekar Adding an other process into the system - Authorised Economic Operator (AEO) Scheme ISO 13485:2016 - Medical Device Quality Management Systems 4
L Warning Letter adressed to owner/operator or facility? US Food and Drug Administration (FDA) 1
B How to change the FDA Device Registration and Listing Module - Owner/Operator Other US Medical Device Regulations 10
M Template for Quality Awareness Training to all Operator Document Control Systems, Procedures, Forms and Templates 6
P Countermeasure For CMM Operator Error Problem Solving, Root Cause Fault and Failure Analysis 4
I Operator Initialling on Paperwork - FDA Good Documentation Practices Other US Medical Device Regulations 11
M Can I continue using IEC60601-1 Ed2 PSU for MOOP (Means Of Operator Protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
Mr.Happy Operator Mistakes (Errors) and responding to CAR (Corrective Action Requests) Nonconformance and Corrective Action 20
K Poka Yoke for Leaf Spring placement by Operator Misc. Quality Assurance and Business Systems Related Topics 2
D Nonconformance or Misinterpretation - When an operator does a Nonconformance Report Nonconformance and Corrective Action 31
shimonv Operator vs. Manufacturer Relationship and ISO 13485 / CE Requirements ISO 13485:2016 - Medical Device Quality Management Systems 5
D If a control operator add a photo as a visual aid, can you improve Occurrence? FMEA and Control Plans 12
C Operator Forgot to Follow Procedure - Now What? Manufacturing and Related Processes 6
L GB15980-2003 Bioburden Count Trend on Operator Hands Other ISO and International Standards and European Regulations 1
L Decision Matrix to help Operator driven Continuous Improvement Lean in Manufacturing and Service Industries 3
I Corrective Action Reply to a Customer - Operator Error Nonconformance and Corrective Action 50
M Voluntary Action by Human Operator Influences Availability? Reliability Analysis - Predictions, Testing and Standards 2
J Standardized Operator Work Practices - Work Instructions - To what Level ? Manufacturing and Related Processes 8
optomist1 Nested Gage R & R One Operator One Sided Spec Using Minitab Software 5
C Corrective Action and Preventive Action for Operator Error (Cosmetic - Handling) Preventive Action and Continuous Improvement 15
optomist1 MSA Gage R&R - Destructive Test with One Sided Specification and One Operator Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
S SPC Gap Analysis for Operator Training Needs Training - Internal, External, Online and Distance Learning 6
C Authorised Economic Operator Status (EU) - HMRC (Her Majesty's Revenue and Customs) Other ISO and International Standards and European Regulations 6
M Operator/Patient Leakage Current IEC 60601 - Battery Charger IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
Q Adding Malicious Operator to FMEA? Quality Manager and Management Related Issues 12
S Control Plan Details vs. Operator Instruction Sheets - Document Contents IATF 16949 - Automotive Quality Systems Standard 9
P ERP (Emergency Response Plan) for Small Air Operator EASA and JAA Aviation Standards and Requirements 7
S Operator Error - How to do a Corrective Action Nonconformance and Corrective Action 14
H FAA Operator - Acceptance of EASA Form 1 for Repaired Parts EASA and JAA Aviation Standards and Requirements 5
M PFMEA (Process FMEA - Potential Causes and Failure) - Operator's Role FMEA and Control Plans 11
Similar threads


















































Top Bottom