Operator vs. Manufacturer Relationship and ISO 13485 / CE Requirements

shimonv

Trusted Information Resource
#1
Hi all
Here is my question:

Company A is medical device manufacture outside the US and EU with ISO certificate 13485.​
Company B is the owner/applicant/operator of the products produced by company A and is situated in the US.​
Company B does sells and after service support for the products produced by company A to both US and EU market.​
Does company B have to comply with ISO 9001 or 13485 assuming it does not modify the product or labeling?​
Does company B needs to be audited by notified body since the CE is registered to its name?


Thanks!
Shimon
 
Elsmar Forum Sponsor
M

MIREGMGR

#2
Yes, as the responsible manufacturer, Company B must have both an ISO 13485 quality system certificate and an appropriate CE certificate. Company B's QMS also must be US FDA QSR compliant.

Company A is a contract manufacturer.

Note that for the US, both Company A and Company B must be Registered and must List the devices...Company A as a contract manufacturer outside the US, Company B as the specification provider and as the initial importer.

Also note that for USA, both companies must have MDR reporting systems, and must coordinate so that both report in a timely manner in regard to any reportable incidents communicated to either.
 

shimonv

Trusted Information Resource
#3
Thanks MIREGMGR!​

And what if A and B are parent company and subsidiary, where B is the parent company situated in the US and does only sells and after-sell support.

A is the R&D and manufacturing subsidiary outside the US.

I gather both will need to register with FDA, and the CE certificate & FDA clearance letter will be issued to the owner, which is B.

What about ISO 13485? should it be issued to A or B or is there a need for two separate certification which is costly and does not make sense to me since for all purposes and intent - B is like a distributor.


Shimon
 
M

MIREGMGR

#4
If the two companies are legally related, you get into complicated territory on the FDA side in which different definitions are used in the various laws and rule sets. If you don't want to get complicated, the conservative approach is to operate as if the companies are independent.

If the two companies operate with full integration, i.e. they only have one Quality Manager and one management structure, then you can justify one quality system and one ISO 13485 certificate. Note that this would require a complicated NB relationship in which two widely separated locations would be supervised.

My assumption though is that because of physical separation, they each have a quality system and management. Given that, the conservative approach is for each to have an ISO 13485 certificate and its own oversight.

B is like a distributor
My view is, never let the fact that a location does distribution, sales and/or support drive your decision as to their regulatory responsibility. These three activities should be the last activities you consider, not the first. As you've described the situation, B is the specification developer and "owns the design" in the sense of being the topmost regulatorily responsible party. If only one location were going to have a regulatory system-stance, in my view it would be B, with A additionally included.
 

Rocke

Involved In Discussions
#5
Hi all
Here is my question:

Company A is medical device manufacture outside the US and EU with ISO certificate 13485.​
Company B is the owner/applicant/operator of the products produced by company A and is situated in the US.​
Company B does sells and after service support for the products produced by company A to both US and EU market.​
Does company B have to comply with ISO 9001 or 13485 assuming it does not modify the product or labeling?​
Does company B needs to be audited by notified body since the CE is registered to its name?


Thanks!
Shimon

Hi Shimon,
If company B is the company mentioned on the labeling, then by definition, they are the manufacturer in the EU. You do not mention what type of MDs are in question, but it is my understanding that a notified body needs to be involved in the CE-marking of all medical devices except for non-sterile class I devices, custom-made devices, and IVD devices not covered by Annex II.

A quality management system is always mandatory and as ISO 13485 is harmonized to the EU MD directives, compliance to the directives is presumed, if this standard is implemented. Certification is not always a requirement - it depends on the product classification.
Activities that are outsourced to e.g. company A can be covered by company A's quality system, if proporly controlled. That all depends on the contractual distribution of responsibilities. Ultimately, the product and all related processes must be under the control of the legal manufacturer.

You should note that if company B (legal manufacturer) is located outside the EU, they require a European Representative, which must be identified on the labeling.
 

somashekar

Staff member
Super Moderator
#6
And what if A and B are parent company and subsidiary, where B is the parent company situated in the US and does only sells and after-sell support.
Regulatory do not care about how your business is run and where your subsidiary are located.
Your B can be the lead site and A the site 1.
Look up GHTF guideline on multisite auditing in the post attachment list and formulate your QMS for ISO 13485 per the details in it.
You can even see THIS thread and find the GHTF guideline in it.
Your NB will audit the site B within the MDD assessment periodically and they may also trigger an audit at site A through site B. Your good quality agreement between Lead site (B) and site 1 (A) can be so established to clearly state responsibilities and cover all such regulatory audits.
 
Last edited:
Thread starter Similar threads Forum Replies Date
M Nonconforming parts accepted by operator with intention Human Factors and Ergonomics in Engineering 31
H EU MDR Economic Operator notifications? EU Medical Device Regulations 1
M Inter-operator Variability Testing - Requirements for EU Medical Device Regulations 5
B Operator protection - When to apply table 7, Dielectric strength test voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
P Economic operator responsibilities - when the Importer is also Distributor - Article 13 EU Medical Device Regulations 1
P How far an operator can reach into a machine before it becomes an ergonomic problem CE Marking (Conformité Européene) / CB Scheme 5
F Operator Manual and Type B Uncertainty Measurement Uncertainty (MU) 3
R Operator's manuals as prevention design control FMEA and Control Plans 2
M Informational EU – M2M Data Exchange available services for accessing MDR EUDAMED data available for Economic Operator (EO) organisations Medical Device and FDA Regulations and Standards News 0
G 3 Operator Std. Dev. and uncertainty calculation Measurement Uncertainty (MU) 2
T Root causes - operator's actions Problem Solving, Root Cause Fault and Failure Analysis 17
K Article 23 (Parts and components) - Economic Operator EU Medical Device Regulations 1
shrutisancheti EU User manual / operator manual / service manual guidance document(s) CE Marking (Conformité Européene) / CB Scheme 2
GoSpeedRacer How to handle operator involvement in the non conformance ISO 13485:2016 - Medical Device Quality Management Systems 5
Fender1 Operator process buy-in - how to get it? Manufacturing and Related Processes 6
K Operator Checks - How to show that they were completed on a checklist as acceptable IATF 16949 - Automotive Quality Systems Standard 2
C ISO 9001:2015 - Non Applicable Design and Development (Power Plant Operator) Design and Development of Products and Processes 2
A Failure Mode from Operator "Components Fallen but Loaded" FMEA and Control Plans 3
A 21CFR807.25 (d) Owner-Operator Contact question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
U USB Connector Operator Protection (MOPP) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
somashekar Adding an other process into the system - Authorised Economic Operator (AEO) Scheme ISO 13485:2016 - Medical Device Quality Management Systems 4
L Warning Letter adressed to owner/operator or facility? US Food and Drug Administration (FDA) 1
B How to change the FDA Device Registration and Listing Module - Owner/Operator Other US Medical Device Regulations 10
M Template for Quality Awareness Training to all Operator Document Control Systems, Procedures, Forms and Templates 6
P Countermeasure For CMM Operator Error Problem Solving, Root Cause Fault and Failure Analysis 4
I Operator Initialling on Paperwork - FDA Good Documentation Practices Other US Medical Device Regulations 11
M Can I continue using IEC60601-1 Ed2 PSU for MOOP (Means Of Operator Protection) IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M Operator Mistakes (Errors) and responding to CAR (Corrective Action Requests) Nonconformance and Corrective Action 20
K Poka Yoke for Leaf Spring placement by Operator Misc. Quality Assurance and Business Systems Related Topics 2
D Nonconformance or Misinterpretation - When an operator does a Nonconformance Report Nonconformance and Corrective Action 31
D If a control operator add a photo as a visual aid, can you improve Occurrence? FMEA and Control Plans 12
C Operator Forgot to Follow Procedure - Now What? Manufacturing and Related Processes 6
L GB15980-2003 Bioburden Count Trend on Operator Hands Other ISO and International Standards and European Regulations 1
L Decision Matrix to help Operator driven Continuous Improvement Lean in Manufacturing and Service Industries 3
I Corrective Action Reply to a Customer - Operator Error Nonconformance and Corrective Action 50
M Voluntary Action by Human Operator Influences Availability? Reliability Analysis - Predictions, Testing and Standards 2
J Standardized Operator Work Practices - Work Instructions - To what Level ? Manufacturing and Related Processes 8
optomist1 Nested Gage R & R One Operator One Sided Spec Using Minitab Software 5
C Corrective Action and Preventive Action for Operator Error (Cosmetic - Handling) Preventive Action and Continuous Improvement 15
optomist1 MSA Gage R&R - Destructive Test with One Sided Specification and One Operator Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
S SPC Gap Analysis for Operator Training Needs Training - Internal, External, Online and Distance Learning 6
C Authorised Economic Operator Status (EU) - HMRC (Her Majesty's Revenue and Customs) Other ISO and International Standards and European Regulations 6
M Operator/Patient Leakage Current IEC 60601 - Battery Charger IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
Q Adding Malicious Operator to FMEA? Quality Manager and Management Related Issues 12
S Control Plan Details vs. Operator Instruction Sheets - Document Contents IATF 16949 - Automotive Quality Systems Standard 9
P ERP (Emergency Response Plan) for Small Air Operator EASA and JAA Aviation Standards and Requirements 7
S Operator Error - How to do a Corrective Action Nonconformance and Corrective Action 14
H FAA Operator - Acceptance of EASA Form 1 for Repaired Parts EASA and JAA Aviation Standards and Requirements 5
M PFMEA (Process FMEA - Potential Causes and Failure) - Operator's Role FMEA and Control Plans 11
R Process Qualification, Equipment Qualification, Operator Qualification - Sequence? Manufacturing and Related Processes 10

Similar threads

Top Bottom