OQ, PQ Requirement for Aerospace? Recently purchased CNC

D

Document Dame

#1
Greetings all,

My company recently purchased a CNC and a friend told me we need to have Operational Qualification and Performance Qualification or it would be a problem for us during our AS9100D transition audit.


I know OQ, PQ is a requirement in the medical field, but is it required in the aerospace field too?

Thanks!
 
Elsmar Forum Sponsor

Coury Ferguson

Moderator here to help
Staff member
Super Moderator
#2
Re: OQ, PQ requirement for Aerospace?

If I understand your question:

Yes I believe...you would be required to perform a First Article Inspection using AS9102, to validate and verify the results of the CNC.

Just my opinion.
 

howste

Thaumaturge
Super Moderator
#3
Greetings all,

My company recently purchased a CNC and a friend told me we need to have Operational Qualification and Performance Qualification or it would be a problem for us during our AS9100D transition audit.


I know OQ, PQ is a requirement in the medical field, but is it required in the aerospace field too?

Thanks!
The relevant AS9100 requirements are in 8.5.1.1 and 8.5.1.3:

AS9100 Rev D said:
8.5.1.1 Control of Equipment, Tools, and Software Programs

Equipment, tools, and software programs used to automate, control, monitor, or measure production processes shall be validated prior to final release for production and shall be maintained.
AS9100 Rev D said:
8.5.1.3 Production Process Verification

The organization shall implement production process verification activities to ensure the production process is able to produce products that meet requirements.

NOTE: These activities can include risk assessments, capacity studies, capability studies, and control plans.

The organization shall use a representative item from the first production run of a new part or assembly to verify that the production processes, production documentation, and tooling are able to produce parts and assemblies that meet requirements. This activity shall be repeated when changes occur that invalidate the original results (e.g., engineering changes, production process changes, tooling changes).

NOTE: This activity can be referred to as First Article Inspection (FAI).
The organization shall retain documented information on the results of production process verification.
AS9100 doesn't specifically require First Articles to AS9102. You should verify if your customers require FAI per the AS9102 requirements.
 
D

Document Dame

#4
Thank you folks for the information.

I apologize for my delayed reply, but a broken hip slowed me down a little.
 

Eredhel

Quality Manager
#6
We're also a CNC shop with AS9100. The AS9102 first article is not an AS9100 requirement, it's a guidance in its Annex. But it can often be customer driven. I think you'll find most AS9102 requirements will be part driven, not machine driven. But until you have customers requiring it you don't need to. If an auditor requires it have them "show you the shall". We verify with First Part Inspections and our own internal First Articles, that meets the standard but is not an AS9102 FAIR.
 
Thread starter Similar threads Forum Replies Date
C Requirement to link Quality Manual to ISO 9001 clause numbers? ISO 13485:2016 - Medical Device Quality Management Systems 13
B FDA requirement for CAPA Signoff ISO 13485:2016 - Medical Device Quality Management Systems 6
W Pfmea function requirement and failure mode FMEA and Control Plans 6
Y 510k requirement for bringing in gloves Medical Device and FDA Regulations and Standards News 0
S PSUR - MDR First Submission Requirement? EU Medical Device Regulations 2
J When PPAPs are not a contractual requirement APQP and PPAP 9
I Overwhelmed with attribute MSA requirement for visual inspection IATF 16949 - Automotive Quality Systems Standard 8
Brizilla Employee Data Privacy Policy - ISO 9001:2015 requirement(s)? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
P Clinical Trial Requirement - No equivalent devices in the market CE Marking (Conformité Européene) / CB Scheme 3
J Requirement for Retention of Records of Withdrawn Documents of External Origin Document Control Systems, Procedures, Forms and Templates 3
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 1
A The requirement of the PMS plan--suitable indicators and threshold values EU Medical Device Regulations 0
S Battery powered device - electrical protection requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 18
K Regulatory requirement of SaMD with machine learning component IEC 62304 - Medical Device Software Life Cycle Processes 3
A ASL requirement when the supplier is certified for ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
P Regulatory or registration requirement for disinfectant in ASEAN Registrars and Notified Bodies 1
P IATF 16949 requirement - error-proofing in control plan IATF 16949 - Automotive Quality Systems Standard 3
C ISO/ IEC 17021 Resource requirement (need help) Document Control Systems, Procedures, Forms and Templates 5
D ISO 14971:2019 vs MDR Annex 1, Requirement #4 - "Manufacturers shall inform users of any residual risks" ISO 14971 - Medical Device Risk Management 5
B IATF16949 audit requirement - Auditor request UCL and LCL must be show Xbar-R, IATF 16949 - Automotive Quality Systems Standard 7
J 21 CFR 821 Medical Device Tracking Requirement 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
J Is Device Tracking Card an actual requirement under FDA regulation? Other US Medical Device Regulations 0
W Misinterpretation of requirement acceptable as root cause? Problem Solving, Root Cause Fault and Failure Analysis 19
S ECG Cable Banana to Snap or Tab electrode Adapters -- ISO 10993 Requirement Other Medical Device Related Standards 0
D IATF 16949 Requirement for CMMI in a Global Company Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 0
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
L Stage 2 audit - Requirement for 3 months of records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
M Requirement to manufacture under a quality management system EU Medical Device Regulations 4
A Risk Number for each software requirement IEC 62304 - Medical Device Software Life Cycle Processes 7
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
I Laboratory Fridge / Freezer Inventory - Requirement(s) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
A Several sections of traditional 510(K) have the same or similar requirement of content. US Food and Drug Administration (FDA) 4
S Requirement to Conduct New Shelf-life Testing? (re-do testing for design change) EU Medical Device Regulations 3
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
Q User Requirement Specification for HR (Human Resource Management System) Manufacturing and Related Processes 1
P Obstruction alarm requirement ISO 80601-2-12 Other Medical Device Related Standards 1
P Electrosurgical Device User Need: Cord Flexibility -> Requirement Other Medical Device and Orthopedic Related Topics 4
C Industrial scales and MSA (IATF 16949 requirement 7.1.5.1.1) IATF 16949 - Automotive Quality Systems Standard 30
C ISO 13485 Requirement in Australia/NZ for class 1? ISO 13485:2016 - Medical Device Quality Management Systems 1
N Label Placement Requirement in MDD/MDR EU Medical Device Regulations 1
adir88 MDR requirement: Risk Management Plan for "each device" ISO 14971 - Medical Device Risk Management 5
S Article 61 - MDR Clinical Investigation Requirement EU Medical Device Regulations 5
V 3 PQ lots mentioned as requirement in PPAP. Is this necessary? APQP and PPAP 2
I MSA requirement for 5 Micrometers + CP changes need customer approval? IATF 16949 - Automotive Quality Systems Standard 2
F AS9100D, raw material standards are a requirement for manufacturing? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
S Clinical Evaluation - Is this an ISO 13485:2016 requirement? ISO 13485:2016 - Medical Device Quality Management Systems 4
S ISO17020 - Equipment calibration requirement General Measurement Device and Calibration Topics 3
A ISO 17025 Requirement 6.2 - Department Manager Testing Qualifications ISO 17025 related Discussions 1
F Process Review - What is the ISO requirement for reviewing SOPs and quality documentation? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4

Similar threads

Top Bottom