Out of Tolerance Gage - Justification for not Recalling Product

P

pangchiaboon

#1
Dear Sirs,

I face difficulty to justify the acceptance of out of tolerance (base on standard) gage and not to recall product. If happened that external calibrator reported that a gage with 1 micron off spec, normally what are the details should we consider and to use as justification not to recall the product and continue to use the gage?

Pang Chia Boon
 
Elsmar Forum Sponsor

harry

Super Moderator
#2
A gage that's out by 1 micron would most likely be still good enough for normal use. Investigate and if so, there's no necessity to recall your product. What are you measuring and what is the specification.
 

somashekar

Staff member
Super Moderator
#3
Dear Sirs,

I face difficulty to justify the acceptance of out of tolerance (base on standard) gage and not to recall product. If happened that external calibrator reported that a gage with 1 micron off spec, normally what are the details should we consider and to use as justification not to recall the product and continue to use the gage?

Pang Chia Boon
Not to recall a product and continue to use the gage are two completely different aspects.
Not to recall is based on the fitness for use assessment and a simple risk assessment to take a very practical and logical decision.
(Without knowing much about your product and situation, its pretty difficult to make a judgement here and now)
However ...
The gage when detected to be out of spec must be prevented from use rightaway.
 

Bev D

Heretical Statistician
Staff member
Super Moderator
#4
Is the gage reading too large or too small?
Where does the product it measures run vs the specification? If teh product runs well within spec (more than a micron away) then there is no need to recall product.

Other considerations are the R&R of the gage for the product it emasures. is the R&R more than a micron?
 

BradM

Staff member
Admin
#5
Hello there!:bigwave:

An out of tolerance gage is comparing the found error to some established requirement. So, you have two things to look at:

Harry and Bev hit upon the as-found error. Given the magnitude, could it have yielded bad product? Could the gage have suggested that product is good, when in actuality it is bad? If 1 micron would have no impact, then you're done.

To make the numbers easy, let's say your error is:

.01

But your process (and product) is not affected unless there are deviations of say.. .5

That .01 error is insignificant to the .5

Which leads to the second part which is your tolerance. There is a whole Art and discussion on establishing tolerances. But understand that depending on what you are using (like mfg. tolerance), you may find your instrument failing many times on tolerance.

So like above, if you process critical parameter is .5 and my current tolerance is .01; maybe I could increase my tolerance to .05 to avoid so many failures. Just a thought..

Moving forward, however you handle this I would write it up. Note the failure, document what investigation you conducted, and complete with whatever action is necessary. Even if the error had no impact, you would still want to document that. :):agree1:

For the future, consider developing a remedial action plan that can give you a plan up-front, of how to handle failures. It might even involve steps to take given what the magnitude of the failure is.
 
J
#6
Dear Sirs,

I face difficulty to justify the acceptance of out of tolerance (base on standard) gage and not to recall product. If happened that external calibrator reported that a gage with 1 micron off spec, normally what are the details should we consider and to use as justification not to recall the product and continue to use the gage?

Pang Chia Boon
All good answers already.
The bottom line is whether a 1 micron error on the gage is significant enough on your product to cause the product to be unfit for use. If not then don't recall. If so, recall. If not sure after investigation, contact customer. Odds are that, for such a small deviation, no recall will be needed.

Peace
James
 
A

Al Dyer

#7
All good responses, and maybe mine is too simple.

Have you called the customer to ask if everything is alright? The answer will give the direction needed.

Al...
 
Thread starter Similar threads Forum Replies Date
N % Tolerance - Type 1 study on the gages, then a gage R&R (ANOVA) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
B Gage R&R Acceptable (10-30%), deduct Total Variation from Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
F Position tolerance check gage manufacturing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
J Gage R&R without tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
N Straightness Gage Tolerance General Measurement Device and Calibration Topics 0
B Gage R&R MSA - Variation, Tolerance, Variance by different methods IATF 16949 - Automotive Quality Systems Standard 23
J Gage R&R Study (Crossed) - How to determine Process Tolerance Using Minitab Software 4
P Strange Gage R&R results for a Caliper (Tolerance Method) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
C P/T % of a gage R&R to 0 MMC Tolerance study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
W Gage R&R for gage pins used to inspect a hole ID called a Statistical Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
C Roundness Tester Gage R and R Frustration - Tolerance range of 20 microns Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 13
C Using a Pin Gage to check True Position of Inline Holes, Design, Tolerance. Etc Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
S Gage R&R - Percent of Tolerance or Percent of Variation? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
B Accepting/Rejecting Threads WITHIN Gage Tolerance General Measurement Device and Calibration Topics 2
M Honest GR&R (Gage R&R) Tolerance Analysis Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
A %Total Variation & %Tolerance - Differences in Gage R&R Forms and Templates Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
C Gage Resolution vs. Tolerance Manufacturing and Related Processes 5
B Addressing Lots that were measured with Out of Tolerance Gage Pins Nonconformance and Corrective Action 5
D Gage R&R %Tolerance NDC Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
R % Gage R&R Based on Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
C Gage R&R 10:1 Tolerance to Gage Discrimination Rule Explanation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
Jerry Eldred Downgrading Tolerance on Gage Block Set General Measurement Device and Calibration Topics 5
D Part Variation, Unilateral Tolerance, Destructive Testing - Gage R&R Experts pls help Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
N Gage R&R - Tolerance - APQP Forms Disk from the AIAG - Formulas IATF 16949 - Automotive Quality Systems Standard 3
smryan Gage R&R Tolerance - Manufacturer's stated tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
Q Part Tolerance and Equipment Accuracy of thickness gage General Measurement Device and Calibration Topics 1
I Gage Error vs. Dimensional Tolerance - Accuracy needed for measurement equipment Measurement Uncertainty (MU) 10
Q Tolerance Set Plugs to Calibrate Thread Ring Gage - Question General Measurement Device and Calibration Topics 4
C Gage R&R @ 10% - Does that mean you should cut your tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
0 Gage R&R Total Tolerance question - Use the full process tolerance? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
D Gage Out of Tolerance Procedure Document Control Systems, Procedures, Forms and Templates 3
B Calibration Accuracy vs. Gage/Part Tolerance General Measurement Device and Calibration Topics 4
P Calculating Gage R&R with Part Tolerance instead of Total Variation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
N Gage R&R - Using Tolerance method, or part to part variation? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
O Gage R&R when using Tolerance - Really bad Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 18
D One-Sided Tolerance Gage R&R Analysis Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
Y Gage R&R - Calculating by Specification Tolerance vs. Study Variation - Differences Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
R Setting Calibration Tolerance on Gauges - Oven Calibration and Stepped Plug Gage General Measurement Device and Calibration Topics 10
M Gauge (Gage) R&R study - Samples required from full tolerance spread? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
G Destructive test, gage R&R with one sided tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
G How to complete gage R&R when you have only a single sided tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 14
D Best Gage R&R (GR&R - GRR) Method for Product Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
U How do part tolerance and gage tolerance interact? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
R Gage Tolerance vs. Process Parameter Tolerance - Process qualification Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
S Gage pin clearance on close tolerance - Hole size is ?.8746 ? .0001 General Measurement Device and Calibration Topics 9
P Gage R&R - What does the % tolerance says about my Gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
R What does the P/t ratio (Precision/Tolerance) =RR/(USL-LSL) say about my gage study? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
G Tolerance vs Total Variation in Gage R&R report - MSA 3rd edition pg. 116 Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 14
B ANOVA Gage R&R % Tolerance and Preference Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
E Full Scale Tolerance - How to apply full-scale accuracy specification to a gage? General Measurement Device and Calibration Topics 4