Out of tolerance gauge blocks - How did these blocks go out of tolerance?

D

dbulak

#1
I would like a second opinion from others on this matter. I just got back a set of gauge blocks fom being calibrated. Two of the blocks were found to be out of tolerance. One block was over the maximun tolerance limit and the other was under the tolerance limit. These blocks are never used during the year. The just lay in the box. My question is how did these blocks go out of tolerance? What phenomenon is taking place to make them go out of tolerance? Or did the calibration company not get good readings? Thanks in advance!
 
Elsmar Forum Sponsor

Jerry Eldred

Forum Moderator
Super Moderator
#2
Re: Out of tolerance gauge blocks

I don't know that this is "THE" answer. But what I would do is plot the history on the impacted blocks to convince myself whether the readings made statistical sense. If. for example, I had five years at one year intervals, and I saw a general trend, then a "JUMP" in the readings, that would personally be a red flag to me.

If you just switched vendors (i.e.: the blocks have previously been cal'd elsewhere), that may be a potential red flag.

If you have any instruments accurate enough and adequate resolution to reasonably resolve the shift, try some experimental readings to see if they correlate.

These are just some checkpoints I would do.

If you discern that your experimental readings don't correlate to the illedged shift, that is another red flag.

If these are critical or important, I might suggest sending them blindly to a different TRUSTED vendor, and have them re-calibrated. Compare the readings (to within uncertainties).

I'm not at all an expert with gage blocks. But there are a number of reasons their readings could be skewed.

There numerous "if this then that" options too lengthy for this post. Hopefully the above will give you some options and ideas.
 
J

jfgunn

#3
Did you receive an accredited calibration for the blocks?

Did the vendor provide the measurment uncertainty for the calibration?

Did the vendor take this measurement uncertainty into accoun t when making the decision to say that the block was out of tolerance? For example, I had a gage block recently where the tolerance was 4 uin. Starrett got a reading of +4.1uin with an uncertainty of +/-2uin. They told me the block failed. Depending on how I use the block and its associated uncertainty, the block may or may not have "failed".

Also, what were the sizes of the blocks? Were they relatively long or short. Gage blocks can grow up to 1uin/in/yr depending on how they were manufactured, stabilized, etc...

If you have a little more info about the measurement data from this calibration and previous calibrations, I could try to provide a better answer.

I hope this helps a little.
 
D

dbulak

#4
I guess my question is how can blocks grow and how can they shrink? What is this called?
 
W

wmarhel

#6
Physical growth (not measurement error): coefficient of thermal expansion?:cool:
Depending on the environment, you can see changes in relatively short periods of time. The more precision the item, the more likely it is to experience some form of fluctuation caused by environmental factors. Hence the reason for a controlled environment and allowing items to stabilize in temperature before calibrating or using them for calibration.

Wayne
 

rmf180

Involved In Discussions
#7
We have established our own internal tolerance limits for assets. First, consider the use of your blocks. What do you do with them? If you use them to calibrate hand tools with .001" resolution and accuracy, then do you really need to worry about a tolerance of +.000004/-.000001?

As stated in one of the early replies, check out the uncertainty statement as well. Very often you will see that the uncertainty is as large as the reported error. Our calibration providers have removed the In/Out of tolerance statement from their cert and left the decision to the customer.
 
K

Kevin H

#8
Were they hardened steel blocks? Starret, I believe used to make their hardened steel blocks from 52100 grade steel. Hardly a corrosion resistant steel, so you may have picked up minor corrosion products.
Also, they're quenched and tempered, so they're microstructure is tempered martensite. There is enough alloy content that you could have some retained austenite that over time would transform to martensite creating stresses and changing dimensions - the crystal structure for austenite is face centerd cubic, while martensite is body centered tetragonal. Martenisite is considered to be a metastable phase, not a totally stable one.
 

BradM

Leader
Admin
#9
I guess my question is how can blocks grow and how can they shrink? What is this called?
Gremlinology.:lol: They sneak in and cut off some on some gauges, and glue some on other gauges!!:lol::lmao:

OK... enough of the horseplay. I guess there is some element of truth to the calibration gremlin. Kevin and some others hit upon my thoughts, which were the type of material the gauge blocks are made of. Too, are we talking about 1/4 inch gauge blocks, or 10 inch gauge blocks? I always had a terrible time with the long gauge blocks.

Also, did the same organization perform the verification both times? The others have covered it so well about competence and uncertainties. If there are differences among the two sources, you might be seeing a difference between them.

Let us know how this progresses; I'd be interested to know what you find.
 
B

bwilk

#10
To get a decent answer to your question we are going to need a little more information on the gauge blocks, and their calibration.

What temperature are they stored at?
What grade are they?
What spec did they fail?
How much did they fail by?
Whats the Measurement uncertainty stated by the calibration provider?
Do you use gloves when you handle them?
What presevative do you use when you store then fro the whole year?

There are tons of factors that can explain changes in the value of a gauge block. Just need some more info to start narrowing it down.
 
Thread starter Similar threads Forum Replies Date
R Feeler Gauge Calibration - What tolerance is used for calibrating feeler gauges? General Measurement Device and Calibration Topics 4
A Calibration of Pressure Gauge, How to Determine Calibration Tolerance General Measurement Device and Calibration Topics 1
C Gauge R&R %Tolerance Template for One-sided Tolerance - Wire Pull Tester Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
R Circularity tolerance of 0.25mm - The correct Go/NoGo gauge size Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
B Out Of Tolerance Gauge And Holding Back Quarantined Product Misc. Quality Assurance and Business Systems Related Topics 7
M Gauge (Gage) R&R study - Samples required from full tolerance spread? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
A Nested ANOVA MSA - Excel sheet with bilateral tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
dhuot1990 Out of Tolerance for Weights Manufacturing and Related Processes 3
F Tolerance Acceptance Criteria General Measurement Device and Calibration Topics 3
M Definition of Characteristics Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
R Quantity Discrepancy Tolerance Standard(s)? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
T Simulate Tolerance Range with Real Parts Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
D Tolerance and MPEs for Re-calibration Manufacturing and Related Processes 0
S GD&T Bonus Tolerance OOS Capability, Accuracy and Stability - Processes, Machines, etc. 2
WEAVER Can you determine exact date an instrument went out of tolerance? General Measurement Device and Calibration Topics 10
K COPLANARITY: Composite profile tolerance on multiple surfaces- what does" lower dimensional reference frame tolerance" control? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
I Profile tolerance with a positional tolerance to a hole ? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
A % STUDY VARIATION VS. % TOLERANCE Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 31
A +0/-.001 Tolerance question - Feature size is 1.249 +0/-.001 - Actually measures 1.2493 (.0003 OHL Inspection, Prints (Drawings), Testing, Sampling and Related Topics 12
L Validation without Tolerance Qualification and Validation (including 21 CFR Part 11) 0
A Need to calculate tolerance Intervals with a set of non-normal data and 3-Parameter Weibull distribution Using Minitab Software 0
Q Acceptable Tolerance - how to derive? Measurement Uncertainty (MU) 16
D Tolerance symbol Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
D Calibration tolerance question using Pipettes Medical Device and FDA Regulations and Standards News 3
M Determining a tolerance value for Measuring devices in-house inspection General Measurement Device and Calibration Topics 12
G 0.00005" tolerance on micrometer with 0.0005" resolution? General Measurement Device and Calibration Topics 4
N Drawing tolerance vs. Measurement device General Measurement Device and Calibration Topics 7
I Control Plan (Product/Process specification/ Tolerance) acceptance FMEA and Control Plans 27
J Mechanical inspection techniques of close tolerance parts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
Proud Liberal Small radius with tight tolerance on a torus General Measurement Device and Calibration Topics 4
L GRR for a tolerance that has changed Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
T How to justify this High %Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
D Tolerance definition based on expected Cp/cpk Reliability Analysis - Predictions, Testing and Standards 14
N % Tolerance - Type 1 study on the gages, then a gage R&R (ANOVA) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
S Tolerance limits for micro balance General Measurement Device and Calibration Topics 1
J Determining SPC tolerance Statistical Analysis Tools, Techniques and SPC 21
Q % Study variation low, % tolerance high - GR&R Interpretation help Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
R Uncertainty in measurement larger than tolerance Measurement Uncertainty (MU) 2
WEAVER Electronic Weighing Scale Calibration Tolerance Manufacturing and Related Processes 1
I Determining Calibration Tolerance of a Measurement Device General Measurement Device and Calibration Topics 2
Ronen E Tolerance intervals (?) question - Flow Rate Probability Range Statistical Analysis Tools, Techniques and SPC 6
D Tolerance when verifying an analytical balance General Measurement Device and Calibration Topics 3
E Discussion between co-worker on tolerance and uncertainty and how to apply it. Thoughts? 17025 ISO 17025 related Discussions 1
C Thread Tolerance - Can I use a 3A Ring on a part that the drawing calls out for a 2A? Manufacturing and Related Processes 5
T GRR based on part tolerance or process variation. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
A Excel OOT (Out of Tolerance) Formula Excel .xls Spreadsheet Templates and Tools 3
N Tolerance Interval plots in Minitab Statistical Analysis Tools, Techniques and SPC 0
S Question about a basic additive variance/tolerance example Quality Tools, Improvement and Analysis 2
G Assigning a calibration tolerance - An x-y coordinate machine - Uncertainty as my verification tolerance General Measurement Device and Calibration Topics 4
S SPC (Statistical Process Control) for Unilateral Tolerance - Questions Statistical Analysis Tools, Techniques and SPC 6

Similar threads

Top Bottom