Oven Validation Protocol (procedure) needed
Jeff Frost
Trucle
Our Calibration Engineer recommends AMS 2750 as a starting point for calibration and survey methods for ovens.
BradM
First, a belated welcome to the Cove!:bigwave:We are sure glad you dropped by, and hope you tell your friends.
When you ask for an oven validation protocol, is this the FDA? Or, are you in the aerospace industry, and need to perform uniformity studies?
For calibration/ uniformity surveys, Jeff provided the "Handbook" so to speak, for the aerospace industry. That will let you know for your particular equipment how many T/C's, location, tolerance, etc.
For a validation, that is a little different story (for pharmaceutical industry); hence Harry's link. Your protocol should generally follow a pre-existing plan established for protocols. Per Marc's question, is this an IQ validation plan? or are you interested in temperature uniformity/ tolerance?
For temperature, you should place T/C probes uniformly throughout the realistic work zone, and verify deviation from the control probe, overshoot, recovery time, correlation to load probes, etc.
TrucLe
Thank You for your help. I would like to know validation plan oven for temperature and humility (FDA). I need to validate oven before i use it to test shelf life of medical device. What the performance test i should do? (temperature test, humility ......)Please give advice.
BradM
Do you have a current validation plan in place at your facility? Is there any up-front functional specifications written for what the oven is supposed to do/is expected to do?
You should have a general layout on how you will proceed. You will need to perform an Installation Qualification, and then some level of Operational/ Performance Qualification.
As to actually validating the oven, it will depend on what you want to use it for. If you are dry heat sterilizing, then you will probably have to verify sterility issues along with the functionality of the oven.
It's important that your validation documentation follow any standard approach you have within your organization for validation, and to address the requirements established (hopefully) when the equipment was purchased.
dataglo
Do a search on GHTF.org. Its medical device specific but it identifies requiements for IQ, OQ PQ. If you still need support after that detail on the forum the paremeters your trying to validate.
rogerheffington
I would like to clarify terms: You validate a process and qualify equipment, You must qualify equipment that you wish to use in a process. Then you validate the process. When I perform an IOQ on an oven, I always conduct uniformity and distribution studies. Additionally, if it is a depyrogenation oven, there would be biological challenges. May I ask: did you recieve the sample protocol you are seeking?
Roger
W Is an Oven I/O/Q Validation Required or is Calibration Sufficient? Qualification and Validation (including 21 CFR Part 11) 7
M Validation of Processes - Is Cl 7.5.2 applicable to bricktile firing in an oven? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
A Oven Controller Specifications - BAC5621 and AMS2750F AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
J CQI-9 - Can OEM do Oven Studies? Reliability Analysis - Predictions, Testing and Standards 1
W Equipment for Oven Calibration, Uniformity and System Accuracy Tests AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
S Performing PQ on a Small (tabletop) Air Heat Treating Oven ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
T Auditor states my Oven Thermocouples must be Calibrated General Measurement Device and Calibration Topics 18
I Is it common practice for a bake oven to require a CQI-9? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
M Traceable Calibration Of An Oven Thermometer General Measurement Device and Calibration Topics 4
D How to Calibrate and Verify a Heating Oven General Measurement Device and Calibration Topics 3
J Moisture Detection and Measurement in Vacuum Oven (Plastic Dryness) General Measurement Device and Calibration Topics 2
WEAVER Thermocouple Extension Wire for Temperature Measurement in an Oven General Measurement Device and Calibration Topics 9
N Is Air Burst test for oven treated condoms still required under ISO 4074? Other Medical Device Related Standards 1
P NADCAP Certified - Heat Treat Continuous Belt Driven Oven question AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
M TS 16949 and Form, Fit or Function? What about Calibration of Oven Controllers? IATF 16949 - Automotive Quality Systems Standard 5
K Pyrometry Question - Furnace and Oven Instrument Calibration IAW AMS 2750 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
N Can a Calibrated Digital Thermometer be used to monitor my oven? General Measurement Device and Calibration Topics 12
B Calculating Measurement Uncertainty in Oven Calibration (using thermocouples) Measurement Uncertainty (MU) 4
W Oven Leak Test Method - Integrity of the chamber to hold gas Qualification and Validation (including 21 CFR Part 11) 3
B Temperature Gage on an Oven was Out of Calibration Period General Measurement Device and Calibration Topics 15
N Bench Top Oven Calibration Standards General Measurement Device and Calibration Topics 8
S Oven Qualification - Thermal profile during the OQ of a vacuum oven General Measurement Device and Calibration Topics 6
P VMP template for a Dry Heat Depyrogenation Oven FAT thru PQ Document Control Systems, Procedures, Forms and Templates 1
R Paperless Chart Recorder to Monitor Oven Cycle Temperatures Manufacturing and Related Processes 5
R Setting Calibration Tolerance on Gauges - Oven Calibration and Stepped Plug Gage General Measurement Device and Calibration Topics 10
Adriane Need Help to Verify an Infrared Cure Oven General Measurement Device and Calibration Topics 9
Marc Convection Oven Cooking - If you have one, do you like it? Cooked a turkey yet? Coffee Break and Water Cooler Discussions 17
C Oven Calibration Range: How many points? General Measurement Device and Calibration Topics 10
W Leak Detection - Oven Seals General Measurement Device and Calibration Topics 1
G MSA Studies on heat shock tests - Should we just be calibrating the oven? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
Y SaMD Verification and Validation SOP and Plan IEC 62304 - Medical Device Software Life Cycle Processes 2
C Spreadsheet Validation and Release Qualification and Validation (including 21 CFR Part 11) 2
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
M Sterile packing validation tests to be performed and protocol Other Medical Device and Orthopedic Related Topics 1
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 0
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
M ERP / QMS related software standards for Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
C ISO 19227 Validation Cost Other Medical Device Related Standards 2
A CNC Mill and Lathe - Validation Manufacturing and Related Processes 2
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
B Oracle Cloud ERP Validation during Quarterly Patch ISO 13485:2016 - Medical Device Quality Management Systems 1
D Software validation team Misc. Quality Assurance and Business Systems Related Topics 3
J Validation Sample Size for Tray Seal Qualification and Validation (including 21 CFR Part 11) 1
F AS9100 - Validation, FAIR's, ITAR and Sub-Contracting AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
R PCBA process validation Qualification and Validation (including 21 CFR Part 11) 2
A ISO 17025 - Methods validation and clients ISO 17025 related Discussions 3
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3