Overhaul of a 20 mm threshold -3rd Ed of IEC60601-1
N
nikolaos
Our medical equipment is certified IEC/EN 60601-1 2nd Ed. and is already in the market. According Clause 4.2.4.3 of the 3rd ED. (overhaul of a 20mm threshold) this is a trouble, having 100mm wheels and small footprint. Equipment does not overturn launched against the step, but does not overhaul it. Previous products has no complains about this. The overcoming of the step was usually done by users by pulling the equipment rear transport handle and pointing a foot at the bottom of the cart, as usual for children stroller. Do you think that this is acceptable by a CB if considered in the Risk Analysis?
Thank you.
Peter Selvey
I think this is excellent case to show how scientific, objective analysis in RM can give a reasonable clear result.
First, a couple of preliminary points: having "no complaints" is not necessarily evidence that there is no unacceptable risk. The issue at hand may be back injury or strain, and seems unlikely that you would get a formal complaint for this point, even though such injuries may have already occurred.
Second: Clause 9.4.2.4.3 is applicable only to devices above 45kg. I'm not aware of too many cases where a stroller and child would be anywhere near 45kg, so this might not be a reasonable comparison to make.
Now to the core risk based decision:
Although the flow in ISO 14971 is based on measuring risk and comparing to set limits, the "risk equations" can be turned around to focus on the relative position of acceptable and actual probabilities for the specific case. This is much easier than trying to measure and set limits for risk as a function of both probability and severity.
Using this method, the basic questions in this case are:
Q1: what is an acceptable probability of medium level back injury?
Let's consider each logarithmic step one by one:
Injury 1 in every 10 times of moving the device (p = 0.1) --> clearly not acceptable
0.01 times / moving --> most would consider not acceptable
0.001 times / moving --> some might accept, some might not
0.0001 times / moving --> most would accept
0.00001 times / moving --> all would accept
So, even a relaxed view would set a probability limit of 0.001, or 1 injury per 1000 times of moving the device.
Q2: is your current design above or below this limit?
It can be very difficult to estimate probability. But we can often make a clear statement relative to the limit. Most people would agree that the probability of injury from man-handling a 45kg device over a 20mm step will be higher than 0.001.
So, we can say with confidence that the risk is unacceptable, and the design should be modified.
There may be technical error somewhere in the above analysis, but the main intention is to show the basic concept of focusing on the relative position of acceptable and actual probability for a certain situation. This is more simple and accurate than trying to create broadly applicable schemes for numerical evaluation of risk based on both severity and probability.
A QMS Overhaul - Reorganizing and Renaming Documents - Internal References Document Control Systems, Procedures, Forms and Templates 24
M Informational From the Implant Files – French commission calls for massive medtech regulation overhaul Medical Device and FDA Regulations and Standards News 1
B IATF 16949 Cl. 8.5.1.5 - Requirement for Periodic Overhaul IATF 16949 - Automotive Quality Systems Standard 10
L Where to Start to Completely overhaul a QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
E How to write a Project Plan to Overhaul our QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
D Aerospace Manufacturer's Equipment Overhaul Manuals Federal Aviation Administration (FAA) Standards and Requirements 7
B Overhaul vs. Repair vs. Rework - FAA Definitions Federal Aviation Administration (FAA) Standards and Requirements 7
P Parts Salvage, Overhaul and Resale - AS9120 Registration question AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
J ROHS compliance for MRO (Maintenance, Repair & Overhaul) Items RoHS, REACH, ELV, IMDS and Restricted Substances 11
M Overhaul QMS or start over? How do you decide? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 40
S Pre-Submission for Breakthrough Device Designation - What's the threshold? US Food and Drug Administration (FDA) 8
A The requirement of the PMS plan--suitable indicators and threshold values EU Medical Device Regulations 0
F ISO 17025:2017 Clause 7.7 Ensuring the validity of results - Threshold ISO 17025 related Discussions 9
V Protocol, report or SOP for threshold analysis? Human Factors and Ergonomics in Engineering 0
V ICH Q3 - To "Include OR Not to Include" impurities below reporting threshold General Measurement Device and Calibration Topics 2
R 7.4.1 and Vendors...Can there be a $$ Threshold? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
E Moving Window Test - Is this test about determining the Threshold Voltage? Quality Tools, Improvement and Analysis 2
M Statistical Method to decide Threshold Value from a Data Set Statistical Analysis Tools, Techniques and SPC 1
S Which Normality Test more acceptable to FDA; Also, Non-Normal Threshold? Qualification and Validation (including 21 CFR Part 11) 5
R How to conduct the Threshold Test of Mobile Equipment specified in 3rd edition? IEC 60601 - Medical Electrical Equipment Safety Standards Series 35
E FMEA Action Plan Threshold (RPN) - Auditor says Action Plan for an RPN > 84 IATF 16949 - Automotive Quality Systems Standard 21
G Robust Threshold Range Levels APQP and PPAP 1
S Gage R&R - Applied threshold for the %EV and %AV values? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
G What is robust threshold range level? I.2.2.15 Bulk Material (Ap - F) APQP and PPAP 1
A Becoming an ISO27001 3rd Party Auditor Career and Occupation Discussions 4
A Non-Conformances Found After 3rd Party Sorting Supplier Quality Assurance and other Supplier Issues 12
G Do HIPAA Rules Apply to a 3rd Party Logistics Shipper? Other US Medical Device Regulations 2
J Extent of 2017/745: providing 3rd party mass produced devices to NHS patients EU Medical Device Regulations 0
S AIAG CQI Auditor Qualification and 3rd Party Certification Requirements General Auditing Discussions 2
J Where I can get a copy of either the 3rd or 4th edition of UL544? Other Medical Device Regulations World-Wide 7
W Internal Auditing carried out by a 3rd party - Review of previous audits AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
M 3rd Party Sorting and Inspection Company Certification Service Industry Specific Topics 4
D Auditing Our Outsourced 2nd-3rd Party Internal Audit Company ISO 13485:2016 - Medical Device Quality Management Systems 6
M CE self-certify, or needs testing by 3rd party? CE Marking (Conformité Européene) / CB Scheme 12
Douglas E. Purdy ISO/IEC 17025:2017 3rd Ed. Changes from 2nd Ed. ISO 17025 related Discussions 6
A CAR from 3rd party AS9100D auditor - Root cause dilemma AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 45
S ISO 9001:2015 Gap Analysis - In-House or 3rd Party? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
T List of Tests - IEC 61010-1, 3rd edition and IEC 60601-1, 3.1 edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
O IATF 16949 3rd party auditor training sources IATF 16949 - Automotive Quality Systems Standard 1
O Difference in 60601-1-2 3rd and 4th ed separation RF portable equipment CE Marking (Conformité Européene) / CB Scheme 1
O Standards used for IEC 60601-1-2 3rd ed. Vs. 4th ed. CE Marking (Conformité Européene) / CB Scheme 1
M IATF 16949 - Multiple Locations - 3rd party audit scope IATF 16949 - Automotive Quality Systems Standard 1
J 3rd Party Certification and QMS Revisions Registrars and Notified Bodies 4
Sidney Vianna Proposed Change to 3rd Party Audit Process - Limiting Scope of Audit Registrars and Notified Bodies 19
A How to 3rd Party Audit ISO 9001 Clause 7.1.6 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
A How to approach ISO 9001:2015 Clause 7.1.6 when 3rd Party Auditing ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Stijloor Calculating (3rd Party) Audit Days for Company with Seasonal Employees General Auditing Discussions 3
C IEC EN 61010 3rd Edition Safety Design Checklist Other ISO and International Standards and European Regulations 3
L Differences in the requirements between IEC 60601-1 3rd ed. and EN 60601-1:2006 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E GR&R ANOVA 3rd Factor for Fixture Nests Using Minitab Software 3