Hello,

Has anyone any experience working with Oxygen concentrators?

What are the standards that regulate Oxygen concentrators for home use other than ISO 8359:1996? Because I'm looking at this standard (ISO 8359) and it does not write about the type of flow control device that can be used. but I know from valve manufacturers that oil/grease should not be present when oxygen is used.

Thank you!
 
Elsmar Forum Sponsor
J

John Nabors - 2009

#2
Freewind-

Two disclaimers- my last experience with oxygen concentrators was over 20 years ago as a repair technician and I am not familiar with the standard you are citing, but here's my :2cents:..

The solenoid valves used for flow control in the concentrators I worked on were designed to isolate the lubricated, metal on metal contact areas from the airflow to the molecular sieve beds. Any surfaces exposed to the air flow did not require lubrication. However, I was told in a training class given by Mountain Medical Equipment in 1984 that any minor (as in less than a gram) inadvertant escapes of lubricant would be adsorbed onto the surface of the molecular sieve material (in their case Union Carbide molecular sieve) and would not be exposed to the patient.

It could well be that over the last 2+ decades that concentrators no longer operate anything like the ones I worked on (actually including Bendix concentrators built in the 70's) but if it is still the same principle of two molecular sieve beds swapping flow, I wouldn't be too terribly concerned with very minor, accidental escapes of graphite into the sieve beds, the sieves will catch and hold it long before it gets to the top of the beds. The only caveat I could think of is this: I was working in Florida where the humidity necessitated frequent replacement (annually at the very least) of the sieve beds, if you are in a more arid climate where the sieve beds last longer, perhaps it could then be more of an issue.

Regards -John

(I have no affiliation with any of the organizations I mentioned.)
 
J

John Nabors - 2009

#3
Freewind-

I just reread your original post and realized that a certain dimwit (who shall remain nameless except to say that it was ME!!) may have missed a critical point in your post. Assuming that oxygen concentrators work the same way they did when I worked on them....

it does not write about the type of flow control device that can be used. but I know from valve manufacturers that oil/grease should not be present when oxygen is used.
...the valves requiring lubrication that direct the flow to the sieve beds will only be exposed to ambient (21%) oxygen. The 90+% oxygen concentration flow coming out of the top of the molecular sieve beds (as I remember) is only regulated by check valves that should not require any lubrication.

Someone please correct me if I am wrong.

Regards -John
 
R

rbhagat

#4
Hi all,

i am working on oxygen concentrator and facing a problem in using molecular sieve beds. can anyone provide me some technical help in using the sieve beds
 
Last edited by a moderator:
J

John Nabors - 2009

#5
Hi all,

i am working on oxygen concentrator and facing a problem in using molecular sieve beds. can anyone provide me some technical help in using the sieve beds
If you can explain the problem and assuming my tired old brain can recall a solution, hopefully I or someone else here can help. I will certainly try.
 
R

rbhagat

#6
Hi
I am working on 5L oxygen concentrator. I am using the sieve beds with dimensions 420 x 75mm. the problem is that when I use readymade sieve beds, machine gives the purity above 90% for all the flow rates. But when I use self made cylinders filled with 5A & 13X sieve the purity dropped down. It gives only 40% on all flow rates. I am using the same dimension of the cylinders and also the same pressure in the sieve. I also tested the sieve beds at different pressure, with different sieves and I also checked the leakage,there is not any leakage but the purity is still low.



Is there any special procedure or equipment required to fill the sieve beds with molecular sieve?
 
J

John Nabors - 2009

#7
Are you using a vibratory table of some sort to pack the sieve material? If the sieve material is not vibrated down to the the highest possible density you will not acheive the oxygen purity you are seeking. Remember that the adsorption of nitrogen by the sieve is a function of the total surface area of the sieve material in the beds; that adsorption capacity is what enables the sieve beds to purge each other and the diversion of oxygen to the patient is really just a by-product of the purge process.

When I was in this line of work (over 20 years ago!!), we did both original manufacture of molecular sieve beds and also 'recycling' by refilling sieve beds that were designed with re-use in mind. In both cases, the beds were refilled on a vibratory table to settle the sieve material in order to pack the beds with the maximum possible ammount of sieve.

Also, be sure to protect your sieve material from moisture/humidity. If you want a really graphic example of what humidity will do to sieve material, take some sieve from an old sieve bed and pour a cup of water into it. The friction of adsorption will actually heat the water to boiling temperature within a couple of seconds.

Good Luck!!

Kind Regards -John
 
R

rbhagat

#8
Thanks John for the information.

There are some more points that I want to know. Sieve is very sensitive to moisture so is there any special room used for the sieve bed packing?


I packed the sieve beds in normal working room with temp. 25 oC and RH. 50 % (approximately)

Currently I am using 2.6kg molecular sieve for 5L in both cylinders of volume 49784.7 mm3/cylinder.

Is there any standard formula or relation between the dimensions of the cylinder, quantity of the sieve, packing density of the sieve in cylinder, and concentration /flow rate of oxygen or in any other parameters?
 
O

oscianwind

#9
Dear bhagat
did the problem of refilling of oxygen sieve bed solved? I'm also fessing same problem can you please tell me the complete procedure of refilling. I'm working on Respironics millennium and airsep. thanks
 

somashekar

Staff member
Super Moderator
#10
Dear bhagat
did the problem of refilling of oxygen sieve bed solved? I'm also fessing same problem can you please tell me the complete procedure of refilling. I'm working on Respironics millennium and airsep. thanks
Welcome to the cove oscianwind ..
You are picking up and responding to a Sept 2008 post and bhagat has last login recorded in Dec 2008. Most likely he will not see your post.
Please anyone here can you comment on this post >>>
 
Thread starter Similar threads Forum Replies Date
A Recreational Oxygen Concentrator - FDA or other requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
N Medical Devices in oxygen rich environment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
P IEC 60601 11.2.2 - Oxygen Rich Environment sources of ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Oxygen enriched environment applicability - Operating table used in general surgeries in hospital IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
C Requirements for dissolved oxygen and/or head space oxygen in Pharmaceuticals Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
S Oxygen mask and nebilzer mask Applied standards EU Medical Device Regulations 2
M Regulations for oxygen/air-tubing and connectors inside of ME equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
F Heat and Moisture Oxygen Port Changes Classification 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
T CFDA issues notice for Small Oxygen Respirators (9/16/2013) China Medical Device Regulations 0
B Defining Expected Oxygen Leakage for Safety Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Ronen E Nidek Medical Class I recall: Medical Mark5 Nuvo Lite Oxygen Concentrators Other US Medical Device Regulations 0
N Chemical Oxygen Demand (COD) Determination in Industrial Waste Water General Measurement Device and Calibration Topics 5
R Testing Medical Equipment used in conjunction with Oxygen Rich Environments IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M Oxygen Data Logger that we can place with each shipment via aircraft ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
T Is Oxygen a Drug per FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
A An Example of a Risk Analysis Report for a Class II Medical Device (Oxygen Mask) ISO 14971 - Medical Device Risk Management 29
H Oxygen Purity Requirements - Oxygen to use in electrical arc furnace ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
F Risk of fire in an Oxygen rich environment - Fire Prevention - IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
sonflowerinwales Oxygen gauge on Nitrogen supply ISO 13485:2016 - Medical Device Quality Management Systems 16
ScottK Oxygen Cleaning and Surfactants used for leak detection Various Other Specifications, Standards, and related Requirements 1
S Do we need to submit our memory foam oxygen mask gasket for FDA approval? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
C How about the classification under 93/42/EEC? Sterile oxygen mask ISO 13485:2016 - Medical Device Quality Management Systems 7
G Process monitoring of portable & stationary oxygen units: Can we use Cpk, Ppk, SPC ? Capability, Accuracy and Stability - Processes, Machines, etc. 3
G Process monitoring - Stationary and portable oxygen units ISO 13485:2016 - Medical Device Quality Management Systems 1
I Сorrespondence between hazards and risks ISO 14971 - Medical Device Risk Management 2
E Normal Condition Hazards in Risk Analysis ISO 14971 - Medical Device Risk Management 3
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T How do you define your Hazards? <a Risk Management discussion> ISO 14971 - Medical Device Risk Management 16
K Identification of hazards and Risk file IEC 62366 - Medical Device Usability Engineering 7
M Identifying Hazards - Risk management process ISO 14971 - Medical Device Risk Management 6
M Medical Device News ECRI – 2019 Top 10 Health Technology Hazards Medical Device and FDA Regulations and Standards News 0
D Hazard analysis for our medical device - Hazards seem to overlap 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D Product Safety in AS9100D Clause 8.1.3 - Hazards Analysis AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 18
M ECRI Top 10 Health Technology Hazards for 2017 Other Medical Device and Orthopedic Related Topics 3
Jen Kirley How to evaluate hazards in using bar code scanners in chemical manufacturing Occupational Health & Safety Management Standards 1
D Different kinds of Risk Analysis for various Hazards ISO 14971 - Medical Device Risk Management 3
M Control Measures for Hazards already deemed Low Risk ISO 14971 - Medical Device Risk Management 6
C Hazards of Old Electronics Quality Manager and Management Related Issues 9
O Hazards vs. Hazardous Situation Confusion ISO 14971 - Medical Device Risk Management 11
E Harms Severity Scoring - Combination of Hazards ISO 14971 - Medical Device Risk Management 4
C List of Potential Hazards Definitions in Annex E of BS EN ISO 14971:2009 ISO 14971 - Medical Device Risk Management 4
P ESD and Bio-Hazards in a Medical Equipment Repair/Maintenance Facility Manufacturing and Related Processes 2
S Refinery Cold Weather Hazards! Manufacturing and Related Processes 0
S Should Identified Hazards and Risks necessarily have Linkage with Legal Requirements? Occupational Health & Safety Management Standards 8
K Do you have to use RPN in Medical Device Risk Analysis? Identification of Hazards ISO 14971 - Medical Device Risk Management 6
S Do Environmental 'Aspects' and 'Hazards' express the same idea? Occupational Health & Safety Management Standards 5
S Should the hazards with negligible risks be documented and/or reviewed? Occupational Health & Safety Management Standards 11
Sidney Vianna ISO/TS 22002-1:2009 gives food manufacturers new tool to control hazards Food Safety - ISO 22000, HACCP (21 CFR 120) 0

Similar threads

Top Bottom