Oxygen Generator - Classification
sriramsl
Dear all
We have a device that Generates Oxygen from Chemicals, through chemical reaction. When a mechanical trigger is twisted by hand, two chemicals come in to contact and then it generates Oxygen, the Oxygen generated builds pressure in the chamber and this pressure is controlled through pressure regulators and using a flow meter, the Oxygen is delivered to the patient. There are no electrical parts. This is purely a mechanical device. The energy for delivery of oxygen is generated by generation of pressure through production of oxygen from chemical reaction. The operation is not through energy generated by the body or Gravity. Hence I have concluded that the device is an Active Medical Device .
Question 1 - Is my conclusion that device is Active device correct?
Question 2 - Should Oxygen has to be considered as a "substance", will Rule 12 of Annex VIII of MDR apply?
Thank you for your support
Raisin picker
Reading MDCG 2021-24, it seems that your device is an active device. See the example with heating pads on page 11. And according to the examples on page 44/45, rule 12 should apply (IIb).
The whole classification, however, relys on your intended use. MDCG 2021-24 has the example "Oxygen concentrator used to deliver oxygen enriched air directly to the patient", which might be close to what you describe. If the oxygen is used for some other purpose (can't think of an example right now that fits medical device definition), everything might be different.
sriramsl
Thank you Raisin Picker for your clarification - Yes MDCG 2021-24 clearly points out both -
1- the heating pad example of converting chemical energy in to heat energy - in our case the chemical energy in to gas production which in turn is turned into penumatic pressure....
2. Yes the example of Oxygne Concentrator gives Rule 12 applicability, but am surprised it is under Class IIb, not Class IIa - may be because the pressure that is built up is considered as potentially hazardous
Thanks again Raisin - Really helpful
Were you by any chance with Picker International, which got dismantled and sold to Philips?
Regards
Sriram
Raisin picker
I think the oxygene concentrator is providing a substance that some living beings consider as essential ;-) So too high or too low a pressure (or O2 concentration) can both be harmful, even deadly. That's why I mentioned the intended use. If you don't supply the oxygen to the lungs for respiration, you might get away with IIa.
And no, my name is just a variation of nitpicker. I have no direct or indirect connection to Philips (or any other major company out there).
R OXYGEN RICH ENVIRONMENT analysis IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N Medical Devices in oxygen rich environment IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P IEC 60601 11.2.2 - Oxygen Rich Environment sources of ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Oxygen enriched environment applicability - Operating table used in general surgeries in hospital IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
C Requirements for dissolved oxygen and/or head space oxygen in Pharmaceuticals Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
S Oxygen mask and nebilzer mask Applied standards EU Medical Device Regulations 2
M Regulations for oxygen/air-tubing and connectors inside of ME equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
F Heat and Moisture Oxygen Port Changes Classification 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
T CFDA issues notice for Small Oxygen Respirators (9/16/2013) China Medical Device Regulations 0
B Defining Expected Oxygen Leakage for Safety Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Ronen E Nidek Medical Class I recall: Medical Mark5 Nuvo Lite Oxygen Concentrators Other US Medical Device Regulations 0
N Chemical Oxygen Demand (COD) Determination in Industrial Waste Water General Measurement Device and Calibration Topics 5
R Testing Medical Equipment used in conjunction with Oxygen Rich Environments IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M Oxygen Data Logger that we can place with each shipment via aircraft ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
T Is Oxygen a Drug per FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
A An Example of a Risk Analysis Report for a Class II Medical Device (Oxygen Mask) ISO 14971 - Medical Device Risk Management 29
A Recreational Oxygen Concentrator - FDA or other requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
H Oxygen Purity Requirements - Oxygen to use in electrical arc furnace ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
F Risk of fire in an Oxygen rich environment - Fire Prevention - IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
sonflowerinwales Oxygen gauge on Nitrogen supply ISO 13485:2016 - Medical Device Quality Management Systems 16
F Oxygen Concentrator Hazards ISO 13485:2016 - Medical Device Quality Management Systems 9
ScottK Oxygen Cleaning and Surfactants used for leak detection Various Other Specifications, Standards, and related Requirements 1
S Do we need to submit our memory foam oxygen mask gasket for FDA approval? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
C How about the classification under 93/42/EEC? Sterile oxygen mask ISO 13485:2016 - Medical Device Quality Management Systems 7
G Process monitoring of portable & stationary oxygen units: Can we use Cpk, Ppk, SPC ? Capability, Accuracy and Stability - Processes, Machines, etc. 3
G Process monitoring - Stationary and portable oxygen units ISO 13485:2016 - Medical Device Quality Management Systems 1
Tagin Using a meme generator to poke fun at quality Funny Stuff - Jokes and Humour 2
N Electrosurgical generator with switchable sockets - creepage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
O Selection of a calibrator (pressure generator, temperature) for my field devices General Measurement Device and Calibration Topics 2
M HF/RF Generator for dielectric strength IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S Defibrillator Requirement for ME Equipment connected to HF Electrosurgical Generator IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Using a Rhode FSH6 Tracking Generator General Measurement Device and Calibration Topics 2
R Does an Outside Steam Generator need to be Registered in the EU? EU Medical Device Regulations 2
M Diesel Generator Stack - Sample port guidelines required Miscellaneous Environmental Standards and EMS Related Discussions 2
G Is cardiac waveform generator a medical device? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Jerry Eldred Philips PM5645 Teletext Generator Procedure General Measurement Device and Calibration Topics 3
T Reference Spectrum Generator for In-House Calibration General Measurement Device and Calibration Topics 2
Scott Catron Warning Sign Generator - An essential part of your safety program... Coffee Break and Water Cooler Discussions 3
Ajit Basrur Random Insult Generator (RIG) Coffee Break and Water Cooler Discussions 2
T HP3325A Function Generator - Troubleshooting Modulation on Output General Measurement Device and Calibration Topics 8
B The Idea Generator - Quick and Easy Kaizen Book, Video, Blog and Web Site Reviews and Recommendations 1
G Masterbatch Pellet Classification Manufacturing and Related Processes 2
J Question regarding classification of custom made device EU Medical Device Regulations 5
S Art 117 product classification under rule 5 EU Medical Device Regulations 0
M Teeth Whitening Product Classification Canada Medical Device Regulations 0
T IVD Device Software - Risk Classification IEC 62304 - Medical Device Software Life Cycle Processes 16
C Applicability of classification rule 9 under MDR EU Medical Device Regulations 6
P Classification of Active non-implantable diagnostic device EU Medical Device Regulations 7