Packaging Addition - Non-sterile barrier to protect the whole package

D

deuce64

#1
Bit of an odd one here. Our Class II (USA) product is currently provided in a single sterile barrier that we have proven in shelf life/distribution out to 3 years. This product contains latex and as such, has seen some degradation issues over the years....though real time studies have proven 3 year shelf life. To further protect this device and lessen complaints we're considering the addition of a foil pouch like condoms use as a non-sterile barrier to protect the whole package. While we plan on performing ship/shake and baseline tests per iso11607, would anyone consider that the benefit of adding this pouch now, with limited shelf life data, would outweigh the risks of NOT adding it. i.e. don't wait for acc. aging data, but implement now to make a speedy improvement. Would love some thoughts. thanks
Note: foil added after eto degassing period (post sterilization)
 
Last edited by a moderator:
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#2
Bit of an odd one here. Our Class II (USA) product is currently provided in a single sterile barrier that we have proven in shelf life/distribution out to 3 years. This product contains latex and as such, has seen some degradation issues over the years....though real time studies have proven 3 year shelf life. To further protect this device and lessen complaints we're considering the addition of a foil pouch like condoms use as a non-sterile barrier to protect the whole package. While we plan on performing ship/shake and baseline tests per iso11607, would anyone consider that the benefit of adding this pouch now, with limited shelf life data, would outweigh the risks of NOT adding it. i.e. don't wait for acc. aging data, but implement now to make a speedy improvement. Would love some thoughts. thanks
You'd have to repeat your sterilization validation to prove this outer package doesn't interfere.
 

somashekar

Staff member
Super Moderator
#4
To further protect this device and lessen complaints ......
What are the complaints like and what has been the root cause and how did you deduce that such a sterile barrier will indeed lessen the complaint...
You could be still overdoing the packaging in your new way with no real benefit.
 
D

deuce64

#5
We have done testing to prove that the foil pouch has protective benefits. Truthfully, as with any latex package, true benefits are only realized in the long haul with realtime aging as accelerated aging is fraut with false negatives. Nonetheless, trying to accertain whther people would be comfortable with the added package that has yet to undergo full term acc aging
 

v9991

Trusted Information Resource
#6
While it is welcome to upgrade the product (package); its little different set of expectation(diligence) for regulated industries;

1) How do you intend to prove that there are no side effects because of additional barrier. (because of additional barrier to moisture, latex could not retain some of desired characteristics; viz., brittle/fragile or smell etc., ) ; essentially how do you avoid from triggering new/different set of complaints.

2) Do you intend to update the labeling details, so as to indicate the set circumstances which customer to avoid(or take care of special precautions

accelerated and real time testing are all to address all such questions;

and current concept, is to assess the relevance of all such above questions through risk-assessment and take necessary call. (remember the emphasis on all such questions and assessment of impact & reasons and ur preparedness to address them; idea is to have list down all relevant-posibilities and take action or be prepared for each one of them.)

:2cents:
 
Last edited:

pkost

Trusted Information Resource
#7
I would be extremely concerned that users may view the foil pouch as a sterile barrier. The products that I work with are commonly double pouched with the everything being sterile. The outer pouch is opened and the inner pouch dropped into the sterile field. In your case this would contaminate the sterile area.

My experience would suggest that you would also be unable to label the fact that the outer pouch is non sterile adequately.....People tend to assume from appearances, and avoid reading the label on something "so obvious"

[edit] to clarify this is biased towards the devices we manufacture, your product may be something completely different where something like what I describe is not an issue
 
D

deuce64

#8

Continued thanks to those who are offering input.

I do see pkost's point about double sterile vs. single sterile confusion.Our product was initially sold in a single barrier, and the proposed pouch andIFU are pretty explicit. Additionally, our sales force intends to inservice thechange...if nothing else, its a reason to reach out to the customer. As you mayor may not know, latex is acted on by many environmental factors: Ozone, heat,humidity, oxygen. As such, our years of working with these products has provento us that accelerated aging cannot reliably replicate the affects of real timeaging - producing false negatives. This leads back to my original query to thisforum: ...if we put aside the factors related to actually proving that thepouch will work, and for the sake of this discussion say that 'it does work',i'm still wondering if those reading would look at the non-sterile package andsay, "ok, i know this is better for my product, but i don't have all theaging data on the pouch to say that it will withstand the test of time.However, even if the pouch were to fail at some point before shelf life ends,(and no longer be a environmental barrier), the product is no worse-off than itwas without it....and likely better than it was ...therefore, i will implement at risk until such aging data (on the pouch)becomes available." thanks again for thoughts.
 
M

MIREGMGR

#9
The sterility issues (technical and communications) have been covered, and I agree with the comments.

In addition, per 11607 and good industry practice you should have done a distribution-cycle validation for your original sterile barrier packaging system (SBPS). Adding a second packaging material in contact with the SBPS raises obvious questions about vibration rubthrough of your SBPS during your distribution validation. If your project were happening here, we definitely would regard it as mandatory to revalidate the SBPS.

Given that requirement, for a product such as yours that is to be protected from UV, ozone and other factors, we would take the different approach of starting over with a hybrid-foil-materials SBPS.

(...) the product is no worse-off than itwas without it....and likely better than it was ...therefore, i will implement at risk (...)
I don't understand how such an approach would be regulatorily justified.
 
Last edited by a moderator:
D

deuce64

#10
MIREGMGR - good points re: 11607. Indeed, we are doing/have done that testing at baseline and have shown no detrimental effects. We also have done the performance testing to show the added benefit in a side-by-side ozone exposure test. As you are aware, there is an aging component of 11607-01 which essentially looks at the seal integrity out at the accelerated (and realtime) age interval. Since we are trying to impart fixes to an increased complaint trend - one that is not visible for months after corrective actions are put into affect - we are approaching the deficiency to the 11607 aging component by justifying that the risks of NOT implementing (i.e. higher user inconvenience rates) is outweighed by the potential benefits that would be acheived through having this pouch. Worst case is that the pouch doesn't provide a benefit, and we have just added labor and materials that are not useful. So that is the angle we're taking on 'regulatorily justifying' the change. To this I ask your's and other's thoughts.
 
Thread starter Similar threads Forum Replies Date
C Packaging Long Parts Manufacturing and Related Processes 1
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
G Special 510K possible? Only the packaging has changed Other US Medical Device Regulations 0
B Setting flexible packaging specification limits ISO 14971 - Medical Device Risk Management 4
V Packaging in Distributor Practices EU Medical Device Regulations 0
R Green dot required on packaging? Medical Device and FDA Regulations and Standards News 2
shimonv Single lot release for sterile packaging 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
R Applicable ISO standards - packaging and others Medical Device and FDA Regulations and Standards News 4
A FDA guidance on non-sterile Medical Device Packaging Medical Device and FDA Regulations and Standards News 7
C 21 CFR 803 malfunction detected during packaging - report? Other US Medical Device Regulations 1
E Latex in packaging Other Medical Device Related Standards 4
H Expired packaging material - Expired unused Tyvek pouches Other Medical Device and Orthopedic Related Topics 2
A GUDID - Packaging Levels 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
A Packaging device with accessories already on the market Other Medical Device Regulations World-Wide 0
P Unrealistic Packaging Validation Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 13
T Root Cause Failure Analysis - Not following Customer packaging Specification Problem Solving, Root Cause Fault and Failure Analysis 9
Q EU Rep for Packaging waste, WEEE and Battery disposal CE Marking (Conformité Européene) / CB Scheme 0
L AS5553 Clause 3.1.7 e "Control packaging material ..................reused" AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
L Packaging unsterile medical devices ISO 13485:2016 - Medical Device Quality Management Systems 6
M MDR - Packaging - Secondary Medical Device Boxes EU Medical Device Regulations 2
Y Packaging validation for non-sterile Medical Equipment Other Medical Device Related Standards 1
M UDI size on packaging EU Medical Device Regulations 3
M UDI placement - Packaging levels EU Medical Device Regulations 2
E Unit Dose Packaging Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
K Packaging and Damaged Parts returned from a Service Provider (Anodize) - Who is Responsible? Manufacturing and Related Processes 20
B 510(k) clearance claim on medical device product packaging Other US Medical Device Regulations 6
E Packaging Specifications Mercedes-Benz U.S. International IATF 16949 - Automotive Quality Systems Standard 4
T Colour of symbols on Medical Device labelling and packaging Other Medical Device Related Standards 1
R Labels on sterile packaging EU Medical Device Regulations 5
A Interpretation of Article 16 (2b) - Packaging, samples and Certificate EU Medical Device Regulations 10
I Medical Device Packaging Multilingual Labeling EU Medical Device Regulations 3
N Process Flow diagram steps for inspection and packaging APQP and PPAP 4
C ISO 7010 m002 mandatory action - Color requirements on Packaging IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
shimonv GTIN Number on the Packaging of a Serviced Medical Device EU Medical Device Regulations 0
shimonv Is UDI required on the Packaging of a Serviced/Repaired Medical Device? Other US Medical Device Regulations 3
C QR Code on Medical Device Packaging and Notification to Notified Body EU Medical Device Regulations 1
TheMightyWife Address Change - Medical Device Packaging Labelling Requirements EU Medical Device Regulations 7
E Use of FDA Cleared logo on packaging US Food and Drug Administration (FDA) 8
V Tyvek packaging for Medical device process validation Qualification and Validation (including 21 CFR Part 11) 2
S Symbols - Device label versus Packaging label ISO 13485:2016 - Medical Device Quality Management Systems 1
Y Sterile device packaging under MDR EU Medical Device Regulations 3
A Packaging seal inspection Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
A Capability Analysis for Packaging Seal Strength with spec. >0.1 Kgf using Minitab Using Minitab Software 6
A Medical Device Class 2A, non sterile bodyworn packaging requirements CE Marking (Conformité Européene) / CB Scheme 7
N Medical device non sterile packaging requirement CE Marking (Conformité Européene) / CB Scheme 1
P Drug packaging / filling operations Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 8
J Animal-Derived products in Packaging Manufacturing and Related Processes 0
P China drug packaging DMF China Medical Device Regulations 0
C ISO 11607-1 Packaging System - Temperature and shipping conditioning Other Medical Device Related Standards 4
M Regulatory intricacies - tertiary packaging of class III MD EU Medical Device Regulations 6

Similar threads

Top Bottom