Packaging Re-Validation - Sealer machine broke down or was changed to a new one
Ajit Basrur
Staff member
Admin
#11
Hi!
You discussed the situation when a new sealer is installed. I performed the IQ, OQ and PQ for one sealer, but there are other two on the site. They are the same brand and model. Could you please advise: Should I perform the same tests as for the first sealer?
Is it possible to shorten the qualification process of the other two sealers? Do I have to perform all OQ tests again?
Thank you.
Welcome to the Cove :bigwave:
If you shorten the qualification procedure for the similar sealer, how can you demonstrate and assure that the second and third sealer behave identical as first one ?
Elsmar Forum Sponsor
Y
Yarocin
#12
The other sealers are the same brand and model. They are identical and they are used for sealing of the same type of materials. I assume that in this case I can give just a good rationale for shortened qualification.
Thread starter Similar threads Forum Replies Date
J Shelf Life Validation for new Heat Impulse Medical Packaging Sealer Qualification and Validation (including 21 CFR Part 11) 2
W Packaging / Heat Sealer Validation - IOQ (Installation/Operational Qualification) Qualification and Validation (including 21 CFR Part 11) 15
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
P Unrealistic Packaging Validation Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 39
Y Packaging validation for non-sterile Medical Equipment Other Medical Device Related Standards 1
V Tyvek packaging for Medical device process validation Qualification and Validation (including 21 CFR Part 11) 2
shimonv Packaging Validation for Non-Sterile Package EU Medical Device Regulations 6
R Medical Device Packaging Label Validation Other US Medical Device Regulations 8
P Sterile Medical Packaging and Shelf Life Validation Sample size ISO 13485:2016 - Medical Device Quality Management Systems 10
R Can a Distributor perform Packaging Validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
J Packaging Validation - IOQ Protocol template for a Sartorius Checkweigher Design and Development of Products and Processes 2
I Work Instruction for Validation of Packaging to EN ISO 11607-2 Requirements Document Control Systems, Procedures, Forms and Templates 3
P Sterile Packaging Validation Requirements - Accredited Laboratory or not? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
P Sterile Packaging Validation - FDA Requirements for Statistically Valid Tests 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
ScottK Validation of SAP - Secondary Packaging Manufacturer Qualification and Validation (including 21 CFR Part 11) 8
S Does Package Validation required for Service Parts Packaging Carton Box? ISO 13485:2016 - Medical Device Quality Management Systems 1
T Packaging Validation - New Machinery - Terminally Sterilized Medical Devices Qualification and Validation (including 21 CFR Part 11) 7
M Do I need to do Packaging Validation? Other Medical Device and Orthopedic Related Topics 3
B Packaging Validation is Suffocating me - We deliver our products non-sterilized Other Medical Device and Orthopedic Related Topics 14
B Packaging Validation - Storage Distribution Testing Qualification and Validation (including 21 CFR Part 11) 4
A Packaging Validation / Qualification Report for Medical Devices - Example ISO 13485:2016 - Medical Device Quality Management Systems 5
M Packaging standards - ASTM vs ISTA Other Medical Device Related Standards 0
M Packaging studies - ISTA vs ASTM CE Marking (Conformité Européene) / CB Scheme 0
F UDI-PI required on packaging (MDR) EU Medical Device Regulations 2
C Packaging Long Parts Manufacturing and Related Processes 3
G Special 510K possible? Only the packaging has changed Other US Medical Device Regulations 0
B Setting flexible packaging specification limits ISO 14971 - Medical Device Risk Management 4
V Packaging in Distributor Practices EU Medical Device Regulations 0
R Green dot required on packaging? Medical Device and FDA Regulations and Standards News 2
shimonv Single lot release for sterile packaging 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
R Applicable ISO standards - packaging and others Medical Device and FDA Regulations and Standards News 4
A FDA guidance on non-sterile Medical Device Packaging Medical Device and FDA Regulations and Standards News 7
C 21 CFR 803 malfunction detected during packaging - report? Other US Medical Device Regulations 1
E Latex in packaging Other Medical Device Related Standards 4
H Expired packaging material - Expired unused Tyvek pouches Other Medical Device and Orthopedic Related Topics 2
A GUDID - Packaging Levels 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
A Packaging device with accessories already on the market Other Medical Device Regulations World-Wide 0
T Root Cause Failure Analysis - Not following Customer packaging Specification Problem Solving, Root Cause Fault and Failure Analysis 9
Q EU Rep for Packaging waste, WEEE and Battery disposal CE Marking (Conformité Européene) / CB Scheme 0
L AS5553 Clause 3.1.7 e "Control packaging material ..................reused" AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
L Packaging unsterile medical devices ISO 13485:2016 - Medical Device Quality Management Systems 6
M MDR - Packaging - Secondary Medical Device Boxes EU Medical Device Regulations 2
M UDI size on packaging EU Medical Device Regulations 3
M UDI placement - Packaging levels EU Medical Device Regulations 2
E Unit Dose Packaging Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
K Packaging and Damaged Parts returned from a Service Provider (Anodize) - Who is Responsible? Manufacturing and Related Processes 20
B 510(k) clearance claim on medical device product packaging Other US Medical Device Regulations 6
E Packaging Specifications Mercedes-Benz U.S. International IATF 16949 - Automotive Quality Systems Standard 4
T Colour of symbols on Medical Device labelling and packaging Other Medical Device Related Standards 1
R Labels on sterile packaging EU Medical Device Regulations 5
Similar threads
Top Bottom