Packaging Requirements for Non-Sterile Class II Medical Devices

M

maple68

#1
Hi,

We are about to putting our Non-sterile, class II hand- held medical device on the market. I was asked to look into the packaging requirements. I understand the labelling requirements. But I have no idea what packaging requirements may be applicable to us. I appreciate your help and advice on this.
Thanks a lot,
Maple
 
Elsmar Forum Sponsor

pkost

Trusted Information Resource
#2
There are no specific requirements for the packaging. However, it must not be detrimental to the function and efficacy of the device.

i.e. If drops and knocks will damage your device, your packaging must be sufficiently padded to prevent this damage occurring

Or, a more interesting one that I came across recently - if your device is available in both sterile and non sterile, do not package your non-sterile device in something that could be construed as sterile packaging...it avoids confusion and reduces risk!
 
M

maple68

#3
Thanks a lot for your reply. It's helpful.

May I please ask a further question?

It look there are no specfic requirements for our non-sterile device; however, if we need to decide the packaging, such as, what's the material for the box? what's the filling to protect the device, any drop or vibration test? and etc. what should we do. Do we need to work closely with the packaging supplier to decide?or we need to have somebody design for us.

Thanks again for any advice,

Maple
 

pkost

Trusted Information Resource
#4
ER 3 states:
The devices must achieve the performances intended by the manufacturer
and be designed, manufactured and packaged in such a way
that they are suitable for one or more of the functions referred to in
Article 1 (2) (a), as specified by the manufacturer.
Your packaging must therefore be sufficient to protect the device and you must be able to demonstrate that you meet this requirement. How you do that is up to you.

If you can demonstrate that your device is ok being wrapped in tissue paper then that is fine. If you decide that it needs to be packaged in a metal crate with integral suspension then that is fine too - it is really up to you!

With regards to how to demonstrate that it is capable, there are 3 methods I can think of off the top of my head:
1. the easiest (but least thorough) is taking using the packaging solution for another product that has similar concerns during transportation/storage.
2. If that isn't possible you could do a variety of tests, depending on what your risk assessment has identified as a possible cause of failure - drop test, vibration test, thermal/humidity test, pressure test etc.
3. Package up your device and send it out with your standard couriers to your most obscure location, then get it sent back


With regards to whether you should work closely with your packaging supplier or get someone to design it for you...I can't answer that, all I can say is that it is important that it is done right!
 
M

maple68

#5
Thanks a lot, Pkost.

Those three sugestions are very practical. We will go from there.

Thanks again,
 
#6
Maple,

However it gets mandated to have the package qualified by scientifically getting it tested by means of standards compliance through accredited labs as the equipment's used for testing should even be calibrated (when you come across UL / CB or FDA regulations).

Here are few standards that most medical device companies follow:
International Safe Transit Association
ISTA 1A - ISTA 2A - ISTA 3A

and / or you can also refer to the following ASTM standards as applicable to your device.
ASTM-D3078
Bubble Emission Leak Test
ASTM-D4169
Distribution Simulation Testing
ASTM-D4728
Random Vibration Testing
ASTM-D5276
Drop Test
ASTM-D642
Compression Test
ASTM-D999
Vibration Testing
 
M

maple68

#7
Thanks a lot, mhkeerthi. Greatly appreicated your comments. This is really helpful. I'm working on the test protocol...:)
 
#8
Hello

Just to pick up on this theme of packaging, I am classifying the packaging of a non-sterile Class II device for the EU market.

Our device (1x3x3 cm) is wrapped in medical grade bubble wrap and placed in a medical grade box. I have 3 questions:

1. The primary packaging is the bubble wrap. Is there a requirement for labelling?

2. The box is required to protect the primary packaging and stop it unfolding (i.e. the bubble wrap is not stand alone in protecting the device as a blister pack would be). Is the box considered as primary or secondary packaging?

3. The device is currently labelled with a super-small copy of the box label. Would it be sufficient to label with the LOT number?

Thank you in advance for your help!
 
T

TempusVernumProcella

#9
Help? I have a similar question regarding device packaging but don't know how to start a new thread...
 

Mark Meer

Trusted Information Resource
#10
...Our device (1x3x3 cm) is wrapped in medical grade bubble wrap and placed in a medical grade box. I have 3 questions:
1. The primary packaging is the bubble wrap. Is there a requirement for labelling?

2. The box is required to protect the primary packaging and stop it unfolding (i.e. the bubble wrap is not stand alone in protecting the device as a blister pack would be). Is the box considered as primary or secondary packaging?
In my opinion, the whole thing (bubble-wrap + box) would be considered the "primary" packaging. Other levels of packaging generally refer to different configurations (e.g. a kit), or multiple units (e.g. cartons, shipping containers).

For the most part, the regulations don't specify different requirements for different levels (the UDI requirements being a noticable exception), and much of the information requirements are stated like "on device and/or packaging", so you can put on either device or package.

That being said, where/how your device is intended to be sold should be considered. Is it intended for over-the-counter sale to the general public? Will is be sold in the US? The US FDA (for example) has some very specific requirements in this case.

3. The device is currently labelled with a super-small copy of the box label. Would it be sufficient to label with the LOT number?
The system of identification and traceability is up to you, as long as requirements are met. For a Class II, there isn't any requirement for serial numbers, so LOT numbers should be fine assuming you have no reason to increase the level of traceability.
 
Thread starter Similar threads Forum Replies Date
A Medical Device Class 2A, non sterile bodyworn packaging requirements CE Marking (Conformité Européene) / CB Scheme 7
C ISO 7010 m002 mandatory action - Color requirements on Packaging IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
TheMightyWife Address Change - Medical Device Packaging Labelling Requirements EU Medical Device Regulations 7
I IATF 8.5.1.5 Requirements c) provision of resources and d) packaging and preservation IATF 16949 - Automotive Quality Systems Standard 1
J Change to Packaging Testing Requirements EU Medical Device Regulations 1
S RoHS2 Packaging and Repair Requirements questions RoHS, REACH, ELV, IMDS and Restricted Substances 3
F FDA Device Labelling/Packaging Information Requirements Other US Medical Device Regulations 3
J What is the gowning requirements for the Pharmaceutical Packaging area ? Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 7
J Canada Bulk Packaging and IFU requirements for Class IIa and IIb Medical Devices Canada Medical Device Regulations 2
Q FDA Medical Device Accessory Packaging Requirements US Food and Drug Administration (FDA) 12
S Packaging and Label Change Requirements - ISO 13485/FDA 21CFR820 ISO 13485:2016 - Medical Device Quality Management Systems 3
I Work Instruction for Validation of Packaging to EN ISO 11607-2 Requirements Document Control Systems, Procedures, Forms and Templates 3
P Sterile Packaging Validation Requirements - Accredited Laboratory or not? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
I Drug Products - Packaging and Labeling Operations Requirements - 211.130 Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
A CE Marking and Packaging Requirements for Class I Sterilized Medical Devices EU Medical Device Regulations 8
P Sterile Packaging Validation - FDA Requirements for Statistically Valid Tests 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
K Customer Notification Requirements - New Packaging Machine for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 1
J Translation Requirements on Medical Device Packaging Labels ISO 13485:2016 - Medical Device Quality Management Systems 5
J Labeling Requirements all levels of Packaging EU Medical Device Regulations 5
N Medical Device Secondary Packaging Facility - Regulatory Compliance Requirements? EU Medical Device Regulations 2
J FDA Packaging Test Requirements - Product is PE Bag 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
C Minimum label content requirements for each level of packaging (inner, outer, etc) ISO 13485:2016 - Medical Device Quality Management Systems 5
N CE Marking MDD & EN 980 Labeling Requirements, IFU, Packaging, etc. EU Medical Device Regulations 6
H CE Language Requirements on labeling (both device and packaging) ISO 13485:2016 - Medical Device Quality Management Systems 27
M Packaging standards - ASTM vs ISTA Other Medical Device Related Standards 0
M Packaging studies - ISTA vs ASTM CE Marking (Conformité Européene) / CB Scheme 0
F UDI-PI required on packaging (MDR) EU Medical Device Regulations 2
C Packaging Long Parts Manufacturing and Related Processes 3
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
G Special 510K possible? Only the packaging has changed Other US Medical Device Regulations 0
B Setting flexible packaging specification limits ISO 14971 - Medical Device Risk Management 4
V Packaging in Distributor Practices EU Medical Device Regulations 0
R Green dot required on packaging? Medical Device and FDA Regulations and Standards News 2
shimonv Single lot release for sterile packaging 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
R Applicable ISO standards - packaging and others Medical Device and FDA Regulations and Standards News 4
A FDA guidance on non-sterile Medical Device Packaging Medical Device and FDA Regulations and Standards News 7
C 21 CFR 803 malfunction detected during packaging - report? Other US Medical Device Regulations 1
E Latex in packaging Other Medical Device Related Standards 4
H Expired packaging material - Expired unused Tyvek pouches Other Medical Device and Orthopedic Related Topics 2
A GUDID - Packaging Levels 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
A Packaging device with accessories already on the market Other Medical Device Regulations World-Wide 0
P Unrealistic Packaging Validation Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 39
T Root Cause Failure Analysis - Not following Customer packaging Specification Problem Solving, Root Cause Fault and Failure Analysis 9
Q EU Rep for Packaging waste, WEEE and Battery disposal CE Marking (Conformité Européene) / CB Scheme 0
L AS5553 Clause 3.1.7 e "Control packaging material ..................reused" AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
L Packaging unsterile medical devices ISO 13485:2016 - Medical Device Quality Management Systems 6
M MDR - Packaging - Secondary Medical Device Boxes EU Medical Device Regulations 2
Y Packaging validation for non-sterile Medical Equipment Other Medical Device Related Standards 1
M UDI size on packaging EU Medical Device Regulations 3
M UDI placement - Packaging levels EU Medical Device Regulations 2