Packaging Validation for Non-Sterile Package
shimonv
Trusted Information Resource
Hi Guys,
Does anyone know to which standard you need to test a non-sterile package, like a PC box?
I could not find an harmonized standard for non-sterile packaging.
Thanks,
Shimon
shimonv
Thanks Ronen and Pads38,
I reached out to a friend outside the cove and summarized what I've learned:
1. It is required in the EU; See MDD, Annex I, section 8.6:
"Packaging systems for non-sterile devices must keep the product without deterioration at the level of cleanliness stipulated and, if the devices are to be sterilized prior to use, minimize the risk of microbial contamination; the packaging system must be suitable taking account of the method of sterilization indicated by the manufacturer."
2. It is also required by QSR. See 21 CFR 820.130:
"Each manufacturer shall ensure that device packaging and shipping containers are designed and constructed to protect the device from alteration or damage during the customary conditions of processing, storage, handling, and distribution."
3. In the US there is a consensus standard for it - ASTM D4169.
4. It gets interesting when you look at the MDD list of harmonized standards. There is no harmonized standard for non-sterile packaging. But ISO 11607-1: Packaging for terminally sterilized medical devices, which is a harmonized standard, references the ASTM D4169.
5. My colleague tells me that the notified body will accept packaging validation based ASTM D4169.
6. Another interesting side story - if the packaging is for telecom equipment, there is a free shipping standard for it from ETSI (European Telecommunications Standard Institute); I believe It's ETSI EN 300 019.
7. There are a bunch of other environmental standards that can be used to show compliance to the regulation. Armed with this knowledge you just need to go to a testing lab and see what type of test they can offer.
Learn and be wise!
Shimon
v9991
https://www.aatb.org/sites/default/files/sites/default/files/private/AATB Guidance Doucment No. 9 final 10.23.17[1].pdf
Qualification of Packaging and Validation of Shipping/Transport Procedures
(American Association of Tissue Banks (AATB))
Environmental Engineering (EE); Environmental conditions and environmental tests for telecommunications equipment; Part 2-4: Specification of environmental tests; Stationary use at non-weatherprotected locations
https://www.etsi.org/deliver/etsi_en/300001_300099/3000190204/02.05.01_60/en_3000190204v020501p.pdf
Packaging Process Validation
Recognized Consensus Standards
Gert Andersen
Registered
Hi,
I'm working on a project which is a Class 1 Custom Made Device.
Thanks for guiding me in the right direction, which lead me to this alternative:
ASTM D7386 - 16
Standard Practice for Performance Testing of Packages for Single Parcel Delivery Systems
Gary07
Registered
Now with implementation of MDR, I guess there is also no change compared to MDD -> No additional requirements for non-sterile packaging compared what we know. Was by someone this ASTM D7386-16 Standard sucesfully appled (as an alternative)?
Kind Regards
Gary
Thread starter Similar threads Forum Replies Date
Y Packaging validation for non-sterile Medical Equipment Other Medical Device Related Standards 1
B Packaging Validation is Suffocating me - We deliver our products non-sterilized Other Medical Device and Orthopedic Related Topics 14
R Packaging Validation Manufacturing and Related Processes 1
adztesla Sterile packaging validation for medical device - ASTMD4332 and ISTA2A ISO 13485:2016 - Medical Device Quality Management Systems 3
L ISO 11607-1 Packaging system validation Design and Development of Products and Processes 9
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
P Unrealistic Packaging Validation Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 41
V Tyvek packaging for Medical device process validation Qualification and Validation (including 21 CFR Part 11) 2
R Medical Device Packaging Label Validation Other US Medical Device Regulations 8
P Sterile Medical Packaging and Shelf Life Validation Sample size ISO 13485:2016 - Medical Device Quality Management Systems 10
R Can a Distributor perform Packaging Validation? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
J Packaging Validation - IOQ Protocol template for a Sartorius Checkweigher Design and Development of Products and Processes 2
I Work Instruction for Validation of Packaging to EN ISO 11607-2 Requirements Document Control Systems, Procedures, Forms and Templates 3
P Sterile Packaging Validation Requirements - Accredited Laboratory or not? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
P Sterile Packaging Validation - FDA Requirements for Statistically Valid Tests 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
G Packaging Re-Validation - Sealer machine broke down or was changed to a new one Other ISO and International Standards and European Regulations 11
ScottK Validation of SAP - Secondary Packaging Manufacturer Qualification and Validation (including 21 CFR Part 11) 8
J Shelf Life Validation for new Heat Impulse Medical Packaging Sealer Qualification and Validation (including 21 CFR Part 11) 2
W Packaging / Heat Sealer Validation - IOQ (Installation/Operational Qualification) Qualification and Validation (including 21 CFR Part 11) 15
S Does Package Validation required for Service Parts Packaging Carton Box? ISO 13485:2016 - Medical Device Quality Management Systems 1
T Packaging Validation - New Machinery - Terminally Sterilized Medical Devices Qualification and Validation (including 21 CFR Part 11) 7
M Do I need to do Packaging Validation? Other Medical Device and Orthopedic Related Topics 3
B Packaging Validation - Storage Distribution Testing Qualification and Validation (including 21 CFR Part 11) 4
A Packaging Validation / Qualification Report for Medical Devices - Example ISO 13485:2016 - Medical Device Quality Management Systems 5
C Sterile label on secondary packaging when mixed contents EU Medical Device Regulations 4
I Higher level of packaging UDI-PI EU Medical Device Regulations 8
P Looking for Korean FDA Drug Master File requirements for packaging systems. Other Medical Device and Orthopedic Related Topics 1
rob3027 ISO 13485 justification for design and development exclusion for medical packaging ISO 13485:2016 - Medical Device Quality Management Systems 18
D Packaging requirement ? EU Medical Device Regulations 0
E UDI on product/packaging levels EU Medical Device Regulations 12
N Microbial testing on food packaging materials Food Safety - ISO 22000, HACCP (21 CFR 120) 3
M Sterile barrier system with outer protective packaging Symbol EU Medical Device Regulations 11
S Combination Product - Packaging US Food and Drug Administration (FDA) 6
A Sterile Packaging Quick Query EU Medical Device Regulations 5
S Reintroduce packaging component/med device from white room back to class 7 clean room Manufacturing and Related Processes 1
M Packaging level EU Medical Device Regulations 2
M Customer Specific Requirements - Packaging Spec IATF 16949 - Automotive Quality Systems Standard 10
S Design & Development records for Medical Devices Packaging and Labelling ISO 13485:2016 - Medical Device Quality Management Systems 9
T CE-IVD mark on packaging EU Medical Device Regulations 1
A MDR requirement where unit of use packaging is too small for UDI carrier EU Medical Device Regulations 1
K Re-packaging under MDR EU Medical Device Regulations 3
M Packaging standards - ASTM vs ISTA Other Medical Device Related Standards 1
M Packaging studies - ISTA vs ASTM CE Marking (Conformité Européene) / CB Scheme 2
F UDI-PI required on packaging (MDR) EU Medical Device Regulations 4
C Packaging Long Parts Manufacturing and Related Processes 3
G Special 510K possible? Only the packaging has changed Other US Medical Device Regulations 0
B Setting flexible packaging specification limits ISO 14971 - Medical Device Risk Management 4
V Packaging in Distributor Practices EU Medical Device Regulations 0
R Green dot required on packaging? Medical Device and FDA Regulations and Standards News 2
shimonv Single lot release for sterile packaging 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
Similar threads
Top Bottom