Particular standards, essential performance in EMC Immunity Test

Eamon

Involved In Discussions
#1
I have a simple point to check about the interaction between particular standards and 60601-1-2 immunity testing.

When a particular standard specifies a minimum level for essential performance, is it simply the case that this essential performance requirement transfers into the requirements for immunity testing?

For example 60601-2-26:2012, 3ed (EEG) specifies (201.12.1.101.3) "The signal noise caused by the EEG amplifier and PATIENT CABLE shall not exceed 6µV peak-to-valley referenced to input (RTI)".

On one hand, these EP requirements specify test setups that are probably not compatible with, and certainly cannot all be simultaneously deployed within e.g. the specified RF EMF immunity test setup (-2-26 Figure 202.102).

On the other hand, IEC 60601-1-2 requires a test criterion of maintaining essential performance, and the particular standard sets minimum EP requirements.

Does this 6µV noise level specified as essential performance for EEG translate into a maximum RF EMF artefact level of 6µV in the EMC test? In this context, is the EMF artefact "caused by the EEG amplifier and patient cable"? It's certainly picked up by the cable, for the most part...

There might be other questions pertaining to other EMC immunity test setups, but perhaps someone can help me explore the principle of the matter, by starting with this one.
 
Elsmar Forum Sponsor

Peter Selvey

Staff member
Moderator
#3
Two aspects here:

In the general case, essential performance should be maintained in normal condition, and EMC immunity tests are in principle intended to represent the normal environment. So for example an NIBP should have ±3mmHg static pressure accuracy, it should maintain this through all the tests.

But there are always special cases. For example EEG might be performed in a location that has low EM noise for various reasons (screened room, and /or located away from transmission tower etc). These special cases can be documented in the RMF.

Second aspect is the 6µV limit specifically refers to the "signal noise caused by the EEG amplifier and PATIENT CABLE". So it could be argued that this is a requirement which is not affected by EM noise, except in the sense that the some EM fields might damage the device (e.g. ESD, high magnetic field?), in which case the requirement would be rechecked after the ESD test.

There are likely to be many items of essential performance which are not affected by environmental tests like EMC. The general standard (Clause 4.3) states that for essential performance, compliance is checked by inspection and if necessary functional tests. The "if necessary" parts allows for engineering judgement on a case by case basis.
 
Thread starter Similar threads Forum Replies Date
O What are the IEC 60601-1-2 Collateral and "Particular" standards CE Marking (Conformité Européene) / CB Scheme 1
M Understanding Versions of Collateral and Particular Standards IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
E Are IEC 60601-2-x particular standards mandatory as part of IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M 60601-1 3rd Edition Compliance with Particular Standards & Equivalency IEC/EN stds EU Medical Device Regulations 6
A List of Countries Using ISO standards (medical devices in particular) ISO 13485:2016 - Medical Device Quality Management Systems 5
rezayatmand IEC 60601-2-18 Medical electrical equipment - Part 2-18: Particular requirements for the basic safety and essential performance of endoscopic equipmen IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
V Deleting Risks if a particular Risk has been Eliminated ISO 14971 - Medical Device Risk Management 3
M IEC 60601-2-24 Ed 2.0 (Infusion Pumps Particular Std.) - Redlines IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Are P-charts for a Particular Defect or can Different Defects be counted in one? Statistical Analysis Tools, Techniques and SPC 2
D Quantifying Cost Benefits associated with a Particular Task Lean in Manufacturing and Service Industries 3
A IEC 60601-2-47 - For Europe, What particular standard is to be followed now? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T Access to a particular Medical Device Electronic Communication Test Port IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Classification of Medical Device with Partial Compliance with a particular standard IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S Typical Tools for Particular Tolerances Table? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 15
D IEC 60601-2-18 (IEC 601-2-18) Particular Requirements for Endoscopic Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
A Implementing a change for a particular Process - Document & Change Control ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Any experience with a GRC application (EMC's in particular)? Quality Assurance and Compliance Software Tools and Solutions 1
L Particular, Normative and Collateral - IEC EN 60601 Definitions of IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
E Particular standard IEC 60601-2-40 references obsolete EMC standard IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S AQL .65 Level III - Take a particular component off 100% Inspection AQL - Acceptable Quality Level 4
S EASA - ETSO authorisation (Subpart O of EASA Part 21) for a particular configuration EASA and JAA Aviation Standards and Requirements 9
D Attribute Gage R&R study - Automatic machine that tests for a particular feature Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
J Thread Wire Chart to use when Calibrating a particular size Thread Gage General Measurement Device and Calibration Topics 16
C What Particular Control Chart is Appropriate for Flour (food) Production? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
S How to get calibration vendors (manufacturers in particular) approved without 17025 ISO 17025 related Discussions 8
M Responding to external communciation - In particular 4.4.3 b) - Dinged Miscellaneous Environmental Standards and EMS Related Discussions 6
Marc ISO 13485:1996 - EN 46001 - Medical devices - Particular requirements - ISO 9001 ISO 13485:2016 - Medical Device Quality Management Systems 14
B Exclusions or justification for non-applicability of IEC standards Reliability Analysis - Predictions, Testing and Standards 1
P Citing EN standards Other Medical Device Related Standards 6
Ed Panek Compliance with Standards? When a standard is updated/revised CE Marking (Conformité Européene) / CB Scheme 3
N Harmonized Standards for MDR Certification EU Medical Device Regulations 6
A Some questions about voluntary consensus standards in traditional 510(k) submission US Food and Drug Administration (FDA) 6
D Imaging/Image Quality Standards Other Medical Device Related Standards 1
R ECG Risk Analysis Standards ISO 14971 - Medical Device Risk Management 2
I Management review in conformity assessment standards - Certification Bodies Management Review Meetings and related Processes 6
Govind Informational ISO COVID-19 response: freely available ISO standards Other ISO and International Standards and European Regulations 1
C Does a CE mark infer meeting all applicable standards? CE Marking (Conformité Européene) / CB Scheme 4
T Assessing Hazard-Related Use Scenarios where control measures exist through standards IEC 62366 - Medical Device Usability Engineering 25
Richard Regalado Informational ISO makes available relevant BCMS standards free for online access Business Continuity & Resiliency Planning (BCRP) 5
A Technical File - Updating Applied Standards EU Medical Device Regulations 4
A NB (Notified Body) Audit of Standards ISO 13485:2016 - Medical Device Quality Management Systems 3
R Applicability of new non-harmonized standards (MDD/MDR) EU Medical Device Regulations 8
dgrainger Informational EU published new Harmonised Standards List for the Directives (25/3/2020) Medical Device and FDA Regulations and Standards News 0
P Static Electricity/Pacemaker - Standards for static electricity generation CE Marking (Conformité Européene) / CB Scheme 3
M Referencing Harmonised Standards in Declarations of Conformity EU Medical Device Regulations 4
Pmarszal External Standards and Regulations Management Process Document Control Systems, Procedures, Forms and Templates 10
S IVF medical devices standards and test methods Medical Device and FDA Regulations and Standards News 4
M What are the standards for malls management? ISO 14001:2015 Specific Discussions 1
A Complaint alert levels - Industry standards pertaining to alert levels for complaints Customer Complaints 5
R MDR standards - which standards to comply with ? EU Medical Device Regulations 3

Similar threads

Top Bottom