In my device I have BF applied part that uses USB to connect to PC. I will use USB isolator between applied part and PC that have 2MOPP (for signal and power). Applied part hasn't direct contact with patient skin, only trough clothes. Device can be used by active patient only. PC is used as biofeedback and patient hasn't interest to touch it. Unfortunately PC is in patient environment and as IEC60601-1 Subclause 16.1 says
Clause 16.6.1 says that between devices in ME System (within patient environment) touch current shall not exceed 100uA. In our system probability of connection accessible parts of PC to a patient directly or trough the body of the operator is negligible like clause 8.4.2 c) says. Subclause 3.8 says that
Can I state from this that clause 16.6.1 is not required? PC of course bears CE mark and comply 60950.
It is my first post and I would like to say hello to every one.
In my device I have BF applied part that uses USB to connect to PC. I will use USB isolator between applied part and PC that have 2MOPP (for signal and power). Applied part hasn't direct contact with patient skin, only trough clothes. Device can be used by active patient only. PC is used as biofeedback and patient hasn't interest to touch it. Unfortunately PC is in patient environment and as IEC60601-1 Subclause 16.1 says
within the PATIENT ENVIRONMENT, the level of safety equivalent to ME EQUIPMENT complying with this standard;
On the other hand, a part that an active PATIENT can reach out and touch might present no more RISK to that PATIENT than it presents to an OPERATOR.