PCBA Hardware Component different failure types - How to rate detection?

Cephissus

Involved In Discussions
#1
Hello all,

Let's say that a PCBA uses a specific Resistor in it.

We identified in the FMEA two different failure modes:

- Resistance too low: This is a design issue in which we wrongly calculate the Resistance value of the component
- Short circuit due to silver migration: This is a common failure of a component. Note that in this case the Resistance value was perfectly designed, but the component failed regardless of the defined initial value.

In the first case, Resistance too low, I can rate a low Occurrence value on the fact that we have used this PCBA design multiple times, that we have successfully dimensioned this specific Resistor several times, that we have good calculation sheet to determine its value.
I can also rate a low Detection value based on the fact that we have standard and effective testing method to verify whether the right Resistance value was selected.

In the second case, however, I can rate the occurrence value based on the standard FIT value given for this component based on its environment.

The question is: how do I rate detection for this second form of failure mode?
 
Elsmar Forum Sponsor

yodon

Leader
Super Moderator
#2
Apologies for the delay (work, vacation...)

I don't use detection in a design FMEA. Where a component can fail such that a hazardous situation is exposed, the design needs to incorporate updates to mitigate. This could be through redundancy (inherent safety by design) or taking action to prevent harm (protective measures like alarms).
 

GRP

Involved In Discussions
#3
Hello all,

Let's say that a PCBA uses a specific Resistor in it.

We identified in the FMEA two different failure modes:

- Resistance too low: This is a design issue in which we wrongly calculate the Resistance value of the component
- Short circuit due to silver migration: This is a common failure of a component. Note that in this case the Resistance value was perfectly designed, but the component failed regardless of the defined initial value.

In the first case, Resistance too low, I can rate a low Occurrence value on the fact that we have used this PCBA design multiple times, that we have successfully dimensioned this specific Resistor several times, that we have good calculation sheet to determine its value.
I can also rate a low Detection value based on the fact that we have standard and effective testing method to verify whether the right Resistance value was selected.

In the second case, however, I can rate the occurrence value based on the standard FIT value given for this component based on its environment.

The question is: how do I rate detection for this second form of failure mode?
I suggest you do use detection and to start by defining the requirement, i.e. PCBA to last for n hours under normal operating conditions ("normal" being defined somewhere).

The failure mode is identified: short circuit due to silver migration
The cause: depending on how you frame the analysis, you can remove silver migration from the FM and put it here. Otherwise decide if you will put the cause of silver migration and/or other causes going deeper.

The prevention, state what feature of your design prevents the PCBA failing before the defined requirement

The detection, state how are you going to detect if your PCBA will meet or not your requirement. If you look at J1739 you can basically perform pass/fail test, test to failure or degradation with measurements before and after. These tests rank 8-6 if this happens after release and 5-3 if it is before release. Bear always in mind this comes from a table which is a guideline.

I think that if you start by properly defining the requirement then you will see how to apply the detection controls.

Let me know if you are still following the subject and we can continue sharing views
 

Bev D

Heretical Statistician
Leader
Super Moderator
#4
Let’s think about the detection of silver migration in a slightly different way. When are you going to try to detect it? During design validation or during use? Do you know what causes Silver migration? Are you intending to try to detect the causes of silver migration?
 

Cephissus

Involved In Discussions
#5
Thanks Yodon. If you don't use Detection, how do you rank it in D? How is the RPN calculated?

Thanks GRP. You make me realize that Silver Migration should be described at lower cause levels. Maybe in the material properties of the Resistor or something. But still you talk about a detection of the PCBA failure, not about the Silver Migration failure itself.

Thanks Bev D. So since I know already that the part is going to fail sometimes because of Silver Migration, it's almost like I don't need to detect it. Yes that's it I guess. Detection should be ranked to 1 because everybody knows it's going to happen and you don't even have to test it. Therefore Occurence is a 10 because you also know its a system with problems. Since there is nothing you can do to decrease this occurrence ranking, you must modify the Design to decrease the Severity. !!

Thank you all for your thought provoking inputs ;)
 

yodon

Leader
Super Moderator
#6
Just to be clear, I don't use detection on product or design-related analyses (e.g., SHA, DFMEA); I do use it on process related analyses (e.g., PFMEA).

On the design side, detection could either be by the system or the user (or operator or some other individual). If the system can detect it, the design should incorporate something to deal with it (controls). If the user (or whomever) is expected to detect it (which is a big assumption), you can't necessarily expect they'll do the right thing to deal with it (so saying the user would detect it and act appropriately is, to me, not a control).

RPN is based strictly on severity and likelihood of occurrence.
 

GRP

Involved In Discussions
#7
Yodon, could you kindly share your views on not using detection in a DFMEA?

I mean, imagine you design a product and you conduct a DFMEA. So you have a list of all the functions the product should fulfill if manufactured as per print, with a list of potential failure modes and different causes for each failure mode. Although prevention is the way to go, sure, but at no point do you think how a cause and/or failure mode is detected (analytical or physical methods) before release?

Does your DFMEA provide no output to a design validation plan? In spite of redundancy or alarms, the design improvement does not call for a test to see if it is effective?
 

yodon

Leader
Super Moderator
#8
Does your DFMEA provide no output to a design validation plan? In spite of redundancy or alarms, the design improvement does not call for a test to see if it is effective?
That's a bit of a leap! :) For our medical device work, we follow ANSI/AAMI/IEC 62366 for usability engineering. Alarms would be part of the user interface and would be assessed (for effectiveness) through some usability study.
 

Bev D

Heretical Statistician
Leader
Super Moderator
#9
I'll take this a bit further... I don't use detection in calculating an RPN (I don't use occurrence either but that's a different topic). Detection is a mitigation and is far more complicated than a simple rating scale could ever quantify.

Severe failure modes will drive validation testing. (and the occurrence requirement will drive sample size and type of study design).
Severe failure modes will also drive the development, validation and implementation of manufacturing (release) testing. In particular manufacturing test and inspection methods must also pass an MSA.
 
Thread starter Similar threads Forum Replies Date
optomist1 PCBA Inspection Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
D PCBA was swollen after SMD Manufacturing and Related Processes 12
M REACH assessment of PCBA/finished product REACH and RoHS Conversations 2
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R PCBA process validation Qualification and Validation (including 21 CFR Part 11) 2
D Software Validation - Contract manufacturer of Components (PCBA's) Qualification and Validation (including 21 CFR Part 11) 7
M PCBA Drawing Best Practices IPC Class 3 ISO 13485:2016 - Medical Device Quality Management Systems 1
alonFAI Determining PCBA Xray Test Sampling Size Statistical Analysis Tools, Techniques and SPC 1
S Who is doing BGA cross sectional analysis in assembled PCBA in INDIA? Reliability Analysis - Predictions, Testing and Standards 1
T Faulty PCBA - Our Testing Fails but Manufacturer's Testing Passes Supplier Quality Assurance and other Supplier Issues 2
D PCBA Lab Temperature & Humidity Requirements - Advice needed Manufacturing and Related Processes 7
N Off Shore Supplier PCBA REACH Reporting RoHS, REACH, ELV, IMDS and Restricted Substances 4
P SMT (Surface Mount Technology) and PCBA Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
O PPAP Validation for Electronic Parts (PCBA) APQP and PPAP 6
T PCBA Verification and Testing Manufacturing and Related Processes 3
U Validation of Wave Solder Process - PCBA for Medical Products Qualification and Validation (including 21 CFR Part 11) 12
P Brightness Specification for PCBA Inspection? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
Q Reliability of a Ferrite Inductor mounted on a PCBA Reliability Analysis - Predictions, Testing and Standards 1
Q PCBA Qualification Plan and Protocol example wanted Qualification and Validation (including 21 CFR Part 11) 1
H Basic questions about Installation, Operational & Performance Qualification - PCBA Qualification and Validation (including 21 CFR Part 11) 11
Q PCBA Visual Inspection Tools & Methods - discounting Optical Tools Manufacturing and Related Processes 11
N PCBA (Printed Circuit Board Assembly) Conformal Coating Material and Thickness? Reliability Analysis - Predictions, Testing and Standards 15
D Would Design & Process FMEA (PCBA) be sufficient to fulfill requirement of Cl 7.1? ISO 13485:2016 - Medical Device Quality Management Systems 2
J Using Weibull to Calculate an Optimal PCBA Burn-In Test? Reliability Analysis - Predictions, Testing and Standards 4
P Can anyone help with ESD Protection (PCBA)? How is ESD generated? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
Ron Rompen Document archiving software & hardware Document Control Systems, Procedures, Forms and Templates 0
Q MDDS Hardware upgrades impact on SaMD US Medical Device Regulations 3
D How to properly measure serrations on hardware Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
J Hardware Validation Qualification and Validation (including 21 CFR Part 11) 1
T Ford Q1 MSA - Virtual Site Visit Hardware Manufacturing and Related Processes 14
A Is computing Hardware part of the IVD CE Marking (Conformité Européene) / CB Scheme 3
F Biocompatibility evaluation for Hardware/Interface Components? Other Medical Device Related Standards 13
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
W Air Quality Measurement Hardware and Software General Measurement Device and Calibration Topics 11
A Components of a medical device - Hardware robotic hand EU Medical Device Regulations 3
B Documenting hardware on Form 2 of AS9102 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
R Validation of Medical Device Hardware containing Software - How many to Validate ISO 13485:2016 - Medical Device Quality Management Systems 1
S Mammography standards or regulations for radiology software (no hardware)? Other Medical Device and Orthopedic Related Topics 0
P Software requirement and design specifications - Medical device contains both hardware and software IEC 62304 - Medical Device Software Life Cycle Processes 2
N Configuration Management - Physical Audit and Functional Audit for Software and Hardware General Auditing Discussions 4
C Supplier Assessment for Hardware supporting SaMD ISO 13485:2016 - Medical Device Quality Management Systems 1
M Missing Hardware in Accessory Kits Manufacturing and Related Processes 5
J Hardware Receiving Inspection AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 12
S Testing software-only device on hardware - What are the hardware parameters? Software Quality Assurance 6
C Aerospace Hardware Revision Control AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
M Finding a hardware supplier/distributor compliant with RoHS RoHS, REACH, ELV, IMDS and Restricted Substances 1
N Specifying a MINITAB PC Hardware - What matters? Using Minitab Software 2
K Hardware and OS Responsibility in FDA 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
E Hardware (Such as Rivets and Bolts) - Chemical and Physical Test Reports AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
T Necessary test documents for Class II Medical Device Hardware and Firmware ISO 13485:2016 - Medical Device Quality Management Systems 9