PCBA process validation

#1
Hi,
I work for a Medical Device Startup, and we are in the process of identifying our Process Validation Requirements.
I know that the FDA identified soldering as a "special process", but I wanted to see if anyone knows if this applies to PCBs assembly? If so, do we need to validate the assembly process for each of our PCBAs, or is it sufficient that the process at the Contract Manufacturer (who uses standard equipment for PCB assembly)?
Thanks
 
Elsmar Forum Sponsor

yodon

Staff member
Super Moderator
#2
Well, I was hoping someone would jump in but since nobody has yet, I'll give it a go.

Processes whose outputs are not 100% verified are expected to be validated. Is anything done to give assurance that each PCBA meets spec? Any scanning systems? Electrical tests? What about even later in the process where you might test the installed PCBA?

Stepping back a bit, how do you know the assembly process is set up correctly (e.g., maybe oven temperatures)? Just because you use standard equipment doesn't necessarily imply you're using it correctly!
 

outdoorsNW

Quite Involved in Discussions
#3
I work for a PCB assembly CM. Your PCB design and layout makes a big difference in how easily your finished PCB can be tested. A dedicated tester is best. A general purpose electrical tester is limited by the ability of its probes to touch the test points. The test probes we have are a bit smaller in diameter than a pencil and taper to a rounded point. Tall components can block access a lot of general purpose test equipment can't probe both sides at the same time. Of our three general purpose testers, only one has double side capability.

PCB design also impacts how difficult your board is to build. See the IPC-22xx Series for design guidelines. In a few cases, the IPC acceptance standards 600 and 610 apply different requirements to boards designed to 22xx series vs boards that are not.
 
Thread starter Similar threads Forum Replies Date
P SMT (Surface Mount Technology) and PCBA Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
U Validation of Wave Solder Process - PCBA for Medical Products Qualification and Validation (including 21 CFR Part 11) 7
D Would Design & Process FMEA (PCBA) be sufficient to fulfill requirement of Cl 7.1? ISO 13485:2016 - Medical Device Quality Management Systems 2
D Software Validation - Contract manufacturer of Components (PCBA's) Qualification and Validation (including 21 CFR Part 11) 7
C PCBA Hardware Component different failure types - How to rate detection? FMEA and Control Plans 8
M PCBA Drawing Best Practices IPC Class 3 ISO 13485:2016 - Medical Device Quality Management Systems 1
alonFAI Determining PCBA Xray Test Sampling Size Statistical Analysis Tools, Techniques and SPC 1
S Who is doing BGA cross sectional analysis in assembled PCBA in INDIA? Reliability Analysis - Predictions, Testing and Standards 1
T Faulty PCBA - Our Testing Fails but Manufacturer's Testing Passes Supplier Quality Assurance and other Supplier Issues 2
D PCBA Lab Temperature & Humidity Requirements - Advice needed Manufacturing and Related Processes 7
N Off Shore Supplier PCBA REACH Reporting RoHS, REACH, ELV, IMDS and Restricted Substances 4
O PPAP Validation for Electronic Parts (PCBA) APQP and PPAP 6
T PCBA Verification and Testing Manufacturing and Related Processes 3
P Brightness Specification for PCBA Inspection? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
Q Reliability of a Ferrite Inductor mounted on a PCBA Reliability Analysis - Predictions, Testing and Standards 1
Q PCBA Qualification Plan and Protocol example wanted Qualification and Validation (including 21 CFR Part 11) 1
H Basic questions about Installation, Operational & Performance Qualification - PCBA Qualification and Validation (including 21 CFR Part 11) 11
Q PCBA Visual Inspection Tools & Methods - discounting Optical Tools Manufacturing and Related Processes 11
N PCBA (Printed Circuit Board Assembly) Conformal Coating Material and Thickness? Reliability Analysis - Predictions, Testing and Standards 15
J Using Weibull to Calculate an Optimal PCBA Burn-In Test? Reliability Analysis - Predictions, Testing and Standards 4
P Can anyone help with ESD Protection (PCBA)? How is ESD generated? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
N Riveting - special process Manufacturing and Related Processes 11
M Material incoming to the production process reflected in PFMEA FMEA and Control Plans 5
A API Spec Q1 Purchasing Process - Supplier Reevaluation based on Supplier Risks 5.6.1.4 Oil and Gas Industry Standards and Regulations 7
B Handling lower detection limits for SPC and process performance Statistical Analysis Tools, Techniques and SPC 1
S Approval Process for Disinfectant Sanitizers in Dubai Elsmar Xenforo Forum Software Instructions and Help 0
D Measurables for Plastic Injection molding process Manufacturing and Related Processes 1
S Cleaning process center change ISO 13485:2016 - Medical Device Quality Management Systems 4
Z Rapid audit template for plastic parts manufacturing process Manufacturing and Related Processes 12
R Inspection and Work order process Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
T ISO 13485:2016 Clauses related to process matrix ISO 13485:2016 - Medical Device Quality Management Systems 3
A How to reduce the process SPC monitoring Capability, Accuracy and Stability - Processes, Machines, etc. 3
John Predmore Configuration Management as a process instead of a procedure AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
U Internal Auditor not trained but done Audit for some process Nonconformance and Corrective Action 5
B Two excellent examples of process capability analysis from Quality Magazine Capability, Accuracy and Stability - Processes, Machines, etc. 5
D ECO (Engineering Change Order) process questions ISO 13485:2016 - Medical Device Quality Management Systems 7
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
Pau Calvo Quality Management process is mandatory in ISO9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
Tagin Template or Checklist for Process Change Document Control Systems, Procedures, Forms and Templates 5
LostLouie Manufacturer divorced from Design process, is he justified in design process deficiencies? ISO 13485:2016 - Medical Device Quality Management Systems 9
J 1.11 Preliminary Identification of Special Product and Process Characteristics APQP and PPAP 4
D Question regarding customer feedback process ISO 13485:2016 - Medical Device Quality Management Systems 3
T What does AS9100 mean when it says you must establish a process to do X? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 24
J Process FMEA Template with examples - Cold and Hot Forged components FMEA and Control Plans 4
D Questions regarding process validation ISO 13485:2016 - Medical Device Quality Management Systems 6
D Question regarding ECO process, specifically for Life Science products and defining form fit and function ISO 13485:2016 - Medical Device Quality Management Systems 1
P Managing How PPAP Requirements are Communicated in the Manufacturing Process and to Suppliers APQP and PPAP 5
I Control Plan (Product/Process specification/ Tolerance) acceptance FMEA and Control Plans 27
L Process changes and biocompatibility (ISO 10993-1) Other Medical Device Related Standards 1

Similar threads

Top Bottom