Dear community,
I have read a lot of posts in the past and they have been extremely helpful. Now it is time for my first post here
We are building a desktop medical device. The current hardware has passed 60601 safety testing, as a device for a controlled medical environment (non-home use), since it was only used as a prototype in our lab and in clinical studies. It uses a 60601-certified AC/DC PSU (24V, 2x MOPP, two-prong mains plug), runs on a 24V secondary and has a type B applied part. The casing consists of a mix of sheet metal and insulated plastic, and the AP is electrically connected to the metal casing/chassis.
We have had some issues regarding electrical noise in some of our circuits (affecting our measurement performance), and noticed some HF leakage current on the AP and casing; within limits, but enough to be able to make an LED flicker by placing it between the metal casing and some grounded metal. It can also be felt when lightly brushing the device. It's a similar feeling to operating an aluminium body MacBook while using a two-prong (non-PE) power brick.
There is now a push to certify the next iteration, including for a home healthcare environment (60601-1-11), and at the same time, the desire to mitigate the above issues by grounding the casing/chassis, ideally via a three-prong variant of the same certified AC/DC PSU. The PE aspect is not relevant, since we operate on 24V, but it could act as FE.
However, 60601-1-11 seems to put two roadblocks in our way:
- Clause 6 states that home devices shall be Class I (no PE), and shall not have a functional earth terminal.
- Clause 6 also states that home devices shall only have applied parts of type BF or CF
It seems we cannot rely on PE/FE to provide a stable ground for our sensitive measurements. And even if we redesign the casing to be fully nonconductive, fully isolating the AP from the grounded chassis will be very difficult (think: optical sensor in contact with the patient, mounted on a conductive aluminum structure for stability, and a laser whose metal housing is mounted/electrically connected onto the same structure and simultaneously connected to the GND of its power supply).
Is there a way to argue for exemptions to clause 6 via risk assessment (e.g. PE not necessary for safety, only for performance improvement; no increased risk due to AP type B instead of BF*)? Does anybody have experience with a similar case?
Thank you very much!
* In the past, we were able to argue against the need for testing according to Figure 18 (external voltage on no-PE metal accessible part) by arguing that the likelyhood of exposed external voltages in the vicinity of the device is negligible.
I have read a lot of posts in the past and they have been extremely helpful. Now it is time for my first post here
We are building a desktop medical device. The current hardware has passed 60601 safety testing, as a device for a controlled medical environment (non-home use), since it was only used as a prototype in our lab and in clinical studies. It uses a 60601-certified AC/DC PSU (24V, 2x MOPP, two-prong mains plug), runs on a 24V secondary and has a type B applied part. The casing consists of a mix of sheet metal and insulated plastic, and the AP is electrically connected to the metal casing/chassis.
We have had some issues regarding electrical noise in some of our circuits (affecting our measurement performance), and noticed some HF leakage current on the AP and casing; within limits, but enough to be able to make an LED flicker by placing it between the metal casing and some grounded metal. It can also be felt when lightly brushing the device. It's a similar feeling to operating an aluminium body MacBook while using a two-prong (non-PE) power brick.
There is now a push to certify the next iteration, including for a home healthcare environment (60601-1-11), and at the same time, the desire to mitigate the above issues by grounding the casing/chassis, ideally via a three-prong variant of the same certified AC/DC PSU. The PE aspect is not relevant, since we operate on 24V, but it could act as FE.
However, 60601-1-11 seems to put two roadblocks in our way:
- Clause 6 states that home devices shall be Class I (no PE), and shall not have a functional earth terminal.
- Clause 6 also states that home devices shall only have applied parts of type BF or CF
It seems we cannot rely on PE/FE to provide a stable ground for our sensitive measurements. And even if we redesign the casing to be fully nonconductive, fully isolating the AP from the grounded chassis will be very difficult (think: optical sensor in contact with the patient, mounted on a conductive aluminum structure for stability, and a laser whose metal housing is mounted/electrically connected onto the same structure and simultaneously connected to the GND of its power supply).
Is there a way to argue for exemptions to clause 6 via risk assessment (e.g. PE not necessary for safety, only for performance improvement; no increased risk due to AP type B instead of BF*)? Does anybody have experience with a similar case?
Thank you very much!
* In the past, we were able to argue against the need for testing according to Figure 18 (external voltage on no-PE metal accessible part) by arguing that the likelyhood of exposed external voltages in the vicinity of the device is negligible.