PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor

#1
Clause 14.6 helpfully lists some hazards to consider related to PEMS including: undesired feedback, unavailable data, lack of integrity of data, incorrect data, etc.
Are these listed hazards relevant to devices that don't provide data to the user?

The device I'm working on does not provide any data to the user but uses data internally (e.g. reading a pressure sensor and controlling a motor based off the reading to reach an ideal pressure). I think for devices that use data only, the hazards would be the resulting issue, such as loss of function. Incorrect data from the pressure sensor would be a foreseeable event potentially leading to the device providing too much or too little performance. Do you agree?
 
Elsmar Forum Sponsor

Tidge

Quite Involved in Discussions
#2
Clause 14.6 helpfully lists some hazards to consider related to PEMS including: undesired feedback, unavailable data, lack of integrity of data, incorrect data, etc.
Are these listed hazards relevant to devices that don't provide data to the user?
If data is 'flowing' between units internal to a system (and not 'presented' to a user/patient), it seems to me that it could be possible (based on the design) for a hazardous situation could arise that is worth analyzing. It will of course depend on the nature of your device's design. It may be possible to expose the need for appropriate risk controls without having to keep digging deeper.

In the case you describe (a pressure sensor participating in the control of a motor): I can imagine that this hypothetical device requires a pressure sensor calibration (a risk control) because of ambient differences in air pressure (sea level vs. mountainous elevations). Presumably such a risk control would derive from "incorrect data" from the pressure sensor.
 
#3
If data is 'flowing' between units internal to a system (and not 'presented' to a user/patient), it seems to me that it could be possible (based on the design) for a hazardous situation could arise that is worth analyzing. It will of course depend on the nature of your device's design. It may be possible to expose the need for appropriate risk controls without having to keep digging deeper.
Yes I agree that those are hazardous situations that should be analysed. However, I would not call them Hazards. I would say the hazards relevant to those hazardous situations are the device delivering too much or too little pressure. My issue is with the text of 60601-1 clause 14.6 saying that they are hazards. I would only consider them to be hazards if the device was providing the information to the user so the data is a function of the device.
 

Benjamin Weber

Trusted Information Resource
#4
You should apply your risk management process: Analyse and evaluate the probability of the line of events that could lead to a hazardous situation, e.g. overpressesure due to wrong sensor output. Then evaluate the severity and probability of the corresponding harm, e.g. serious injury of the patient.

Finally evaluate the risk: Is it acceptable? OK. If not, implement risk control/mitigation measures. This could be regular sensor calibration or maybe redundancy. If depends on your device.

To make a long story short: Your risk evaluation decides, if there is a hazard or not.
 
Thread starter Similar threads Forum Replies Date
R Please share your method used for PEMS Validation IEC 62304 - Medical Device Software Life Cycle Processes 5
J Medical Device which is NOT a PEMS? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S PEMS Documentation for Medical Device Software - Requirements IEC 62304 - Medical Device Software Life Cycle Processes 3
R Does dental unit need to meet the requirements of clause 14 (PEMS)? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
P Medical Devices - PEMS, PESS, PE/E/E and the like... once again! IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
R Can manufacturer separately evaluate the clause 14 (PEMS) of 60601-1: 2005? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
E Normal Condition Hazards in Risk Analysis ISO 14971 - Medical Device Risk Management 3
T How do you define your Hazards? <a Risk Management discussion> ISO 14971 - Medical Device Risk Management 16
K Identification of hazards and Risk file IEC 62366 - Medical Device Usability Engineering 7
M Identifying Hazards - Risk management process ISO 14971 - Medical Device Risk Management 6
M Medical Device News ECRI – 2019 Top 10 Health Technology Hazards Medical Device and FDA Regulations and Standards News 0
D Hazard analysis for our medical device - Hazards seem to overlap 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D Product Safety in AS9100D Clause 8.1.3 - Hazards Analysis AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 18
M ECRI Top 10 Health Technology Hazards for 2017 Other Medical Device and Orthopedic Related Topics 3
Jen Kirley How to evaluate hazards in using bar code scanners in chemical manufacturing Occupational Health & Safety Management Standards 1
D Different kinds of Risk Analysis for various Hazards ISO 14971 - Medical Device Risk Management 3
M Control Measures for Hazards already deemed Low Risk ISO 14971 - Medical Device Risk Management 6
C Hazards of Old Electronics Quality Manager and Management Related Issues 9
O Hazards vs. Hazardous Situation Confusion ISO 14971 - Medical Device Risk Management 11
E Harms Severity Scoring - Combination of Hazards ISO 14971 - Medical Device Risk Management 4
C List of Potential Hazards Definitions in Annex E of BS EN ISO 14971:2009 ISO 14971 - Medical Device Risk Management 4
P ESD and Bio-Hazards in a Medical Equipment Repair/Maintenance Facility Manufacturing and Related Processes 2
S Refinery Cold Weather Hazards! Manufacturing and Related Processes 0
S Should Identified Hazards and Risks necessarily have Linkage with Legal Requirements? Occupational Health & Safety Management Standards 8
K Do you have to use RPN in Medical Device Risk Analysis? Identification of Hazards ISO 14971 - Medical Device Risk Management 6
S Do Environmental 'Aspects' and 'Hazards' express the same idea? Occupational Health & Safety Management Standards 5
S Should the hazards with negligible risks be documented and/or reviewed? Occupational Health & Safety Management Standards 11
Sidney Vianna ISO/TS 22002-1:2009 gives food manufacturers new tool to control hazards Food Safety - ISO 22000, HACCP (21 CFR 120) 0
C HACCP in Egg Production - Controlling hazards at the farm level Food Safety - ISO 22000, HACCP (21 CFR 120) 6
E ISO14971 - Determination of possible hazards (SU, UU, PC) ISO 14971 - Medical Device Risk Management 3
K Should we rate according to the prob. a risk will occur or that risk causes hazards? ISO 14971 - Medical Device Risk Management 5
F Oxygen Concentrator Hazards ISO 13485:2016 - Medical Device Quality Management Systems 9
M PFMEA Detection Rating for Operator Hazards FMEA and Control Plans 5
H Safety Hazards Finding and Recommendation Report - Seeking Format Examples Various Other Specifications, Standards, and related Requirements 0
S Painting on Floor of Machine Shop - What Aspects/Impacts and Hazards to consider? Miscellaneous Environmental Standards and EMS Related Discussions 10
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C ISO/IEC 17021-1 clause 7.1.2 - Determination of competence criteria Document Control Systems, Procedures, Forms and Templates 2
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C ISO/ IEC 17021 Resource requirement (need help) Document Control Systems, Procedures, Forms and Templates 5
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 0
S IEC 62304 software costs and time Medical Device and FDA Regulations and Standards News 2
S IEC 62304 - Software verification cost IEC 62304 - Medical Device Software Life Cycle Processes 3
Sravan Manchikanti Software Risk Management & probability of occurrence as per IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
M IEC 62304 Software changes - Minor labeling changes on the GUI IEC 62304 - Medical Device Software Life Cycle Processes 3
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K IEC 62304 - Testing Independance IEC 62304 - Medical Device Software Life Cycle Processes 5
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Exclusions or justification for non-applicability of IEC standards Reliability Analysis - Predictions, Testing and Standards 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7

Similar threads

Top Bottom