Performing a TUS on a Furnace used at One Temperature - AMS2750-E

H

homersam

Hello, I am just new here and am starting to work with AMS2750 rev E

I have a furnace to perform a TUS on it, oven is used at one temp. (121oC +/- 5oC)
Furnace I believe is class D type 3

question would I have to do any temperatures below this or above this temp.?
also there is an excess t/c in oven but no display-how is this checked?

I will probably be back here a few times making enquiries
any help would be much appreciated,
 

dwperron

Trusted Information Resource
If you are only having the oven qualified at a single point then you only need to perform the TUS at that point.

As for the "excess" thermocouple - what is it there for? If you are using it then it needs to be calibrated like any other thermocouple. If you are not using it, you might consider removing it to eliminate questions about it.
 
H

homersam

Thanks for reply dwperron & bigqman

just a few more points

is it not necessary to perform TUS at upper & lower limits
ie 121oC nominal +/- 5oC, checked at 116oC & 126oC?

(would id plate on furnace then require calibrated and for use at 121oC only)
I thought id plate on furnace had to have temperature range marked on id plate, is 121oC only fine?

excess controlling t/c has no readout on control panel of furnace,
inside electrical panel excess controlling t/c is connected to block with three separate settings with three arrows pointing to numbers, looks like it is set to 155oC

I was going to check excess controller is working by inputting a temperature over 155oC from my calibrator and see if it shuts off the furnace, just for completeness

I checked another furnace over a year ago with colleague who was the expert,
that furnace was for use at 515oC one set point, but my colleague insisted on a TUS at 450oC, 515oC & 550oC with excess set at 560oC it all worked out ok as furnace was brand new and calibrated fine
I now don't see the point in doing extra checks if they are not necessary

I will probably have more questions later, thanks for your help :bigwave:
 

dwperron

Trusted Information Resource
is it not necessary to perform TUS at upper & lower limits
ie 121oC nominal +/- 5oC, checked at 116oC & 126oC?


Upper and lower limits are checked for temperature bands, such as an oven qualified for use from 175 to 350°F would be checked at the high and low limits (at least).

I thought id plate on furnace had to have temperature range marked on id plate, is 121oC only fine?

That would be the right thing to do if it is only qualified at 121°C

I was going to check excess controller is working by inputting a temperature over 155oC from my calibrator and see if it shuts off the furnace, just for completeness

If you are only using the probe for over temperature control it does not need to be calibrated. The shut off check is the right thing to do.
 
H

homersam

thanks again dwperron for reply

just looking again at the other furnace my colleague done a survey on a while back, I do remember something in AMS2750 about a temperature set point over 350oC that it needs a TUS carried out below the 350oC and one at set point in this case 515oC, does this look right or have I read it incorrectly?

or is one TUS at 515oC permissible to AMS2750 Rev E
 

dwperron

Trusted Information Resource
The oven your colleague surveyed sounds like it was done over a qualified temperature range, not a single point. The rules for that are as follows:

2750E section 3.5.6:
Periodic TUS temperatures shall be any temperature within each qualified operating range.
3.5.6.1 - For single operating ranges greater than 600°F (335°C), during each periodic test TUS temperatures shall be selected so that one temperature is within 300°F (170°C) of the maximum and another temperature is within 300°F of the minimum of the qualified operating range and there are no more than 600°F (335°C) increments in between.

If this were the initial TUS:
3.5.5.1 - Initial survey temperatures shall be the minimum and maximum of the qualified operating temperature range(s).
3.5.5.2 - Additional temperatures shall be added as required to ensure that no two adjacent survey temperatures are greater than 600°F (335°C) apart.
 
H

homersam

thanks again dwperron



this furnace is only used at one set point , ie 515oC +/- 8oC

so I am not sure why he chose to do a TUS at 450oC, 515oC & 550oC?
would one set point of 515oC be acceptable?

another question if I may
when I am doing the TUS I use paper forms that were set up years ago and manually write every temp. during survey (data logger DPR3000 records t/cs and controlling t/c )
I then go through data logger to find out min. /max temp at each set point and record this on our paper forms and highlight them on data logger print out, the "mid soak" values are also recorded manually on form, this takes a lot of time and one mistake and that's the one an auditor will pick up on
(Our data logger will print every 5 mins. but doesn't print at say 1.05, 1.15,1.25, any thing ending in a 5 so I have to watch time and hit "print" button at each time ending in a 5
Colleague before me has been used to filling in paper forms for years and he used to take them home to fill them in, he wasn't keen in upgrading


Are there newer data loggers/recorders available that would allow you to produce an electronic cal. report that would have min/max temp,ramp up, run time (30 mins.) etc. to satisfy AMS2750 E?

Can the newer data loggers/recorders be set up to perform SAT test reports
(our furnaces are done every fortnight)

My boss is interested in upgrading our old DPR3000 and some training for me and a new colleague could you recommend one or two newer data loggers/recorders so I may look them up,
 
W

Will White

this furnace is only used at one set point , ie 515oC +/- 8oC

so I am not sure why he chose to do a TUS at 450oC, 515oC & 550oC?
would one set point of 515oC be acceptable?

The only reason to do more than one set point is if your furnace is operated at multiple temperatures. Since you run your production at only one temperature you only need to perform your TUS at that one test temp of 515C.


another question if I may
when I am doing the TUS I use paper forms that were set up years ago and manually write every temp. during survey (data logger DPR3000 records t/cs and controlling t/c )
I then go through data logger to find out min. /max temp at each set point and record this on our paper forms and highlight them on data logger print out, the "mid soak" values are also recorded manually on form, this takes a lot of time and one mistake and that's the one an auditor will pick up on
(Our data logger will print every 5 mins. but doesn't print at say 1.05, 1.15,1.25, any thing ending in a 5 so I have to watch time and hit "print" button at each time ending in a 5
Colleague before me has been used to filling in paper forms for years and he used to take them home to fill them in, he wasn't keen in upgrading

Your current system sounds like it is not capable of meeting the current requirements of 2750E. The shortest interval permitted for data capture is once every 2 minutes (see paragraph 3.5.13.3.2 of 2750E).

Are there newer data loggers/recorders available that would allow you to produce an electronic cal. report that would have min/max temp,ramp up, run time (30 mins.) etc. to satisfy AMS2750 E?

Can the newer data loggers/recorders be set up to perform SAT test reports
(our furnaces are done every fortnight)

My boss is interested in upgrading our old DPR3000 and some training for me and a new colleague could you recommend one or two newer data loggers/recorders so I may look them up,

The software packages I've found that automatically generate a TUS report are very expensive for what you get. The best way I've found to manage our data is through Excel. You can build your own macros and spreadsheets to manage the data and do your reporting.

I haven't encountered a data logger or software that will automatically generate a SAT report.

The electronic data loggers that we use are Yokogawa DX paperless chart recorders.
 
Thread starter Similar threads Forum Replies Date
E Performing a GRR when you get multiple readings over time per part. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
C Performing Additional Inspections IATF 16949 - Automotive Quality Systems Standard 10
Mr Roo Discovered dishonesty after performing an internal audit General Auditing Discussions 5
D ISO 13485--Question regarding performing calibrations in house. ISO 13485:2016 - Medical Device Quality Management Systems 13
M Who is performing quality control? IATF 16949 - Automotive Quality Systems Standard 2
AllTheThings ISO-17025 accredited TEST lab performing internal calibrations (Yes: Here is how) ISO 17025 related Discussions 8
Geoff Cotton Performing a Delta Audit General Auditing Discussions 12
A Does AS9100 require traceability to operators performing the work? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
I Root Cause Analysis when someone simply stops performing a task? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
H Question about implications of performing Firmware upgrade via MDDS - Medical Device Data Systems 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
N Justifications for not performing MSA (Measurement System Analysis) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
V Sequence of performing risk assessment: User_FMEA (User Errors) vs Design Inputs FMEA and Control Plans 1
Y Performing product audits in a manufacturing facility Manufacturing and Related Processes 13
armani ISO 9001:2015 - How to apply 7.1.5 to an organization performing only design ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
S Performing PQ on a Small (tabletop) Air Heat Treating Oven ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
B Thoughts on performing an ISO 9001:2015 Remote Internal Audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
drgnrider EMR performing audits of EMS Miscellaneous Environmental Standards and EMS Related Discussions 13
C Performing an ISO 9001 based Supplier Audit - Multiple Supplier Facilities Supplier Quality Assurance and other Supplier Issues 4
D Performing a Gage R and R with No Specification Range Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
C Going Rate for Performing a Supplier Audit? Consultants and Consulting 1
B Performing MSA on Straightness of a Cylinder Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 13
C Performing a MSA as it relates to Calibration Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
J Performing a MSA on a new gauge using different parts Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
K Production Operators Performing GP12 (General Motors Containment) Customer and Company Specific Requirements 2
L How is the interaction physically happening while performing ANOVA GR&R? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
E Exam for Personnel performing Calibration General Measurement Device and Calibration Topics 7
B Performing a Gage R&R Study with Minitab 13 Using Minitab Software 5
M Requirements for Laboratories performing ASTM Tests Various Other Specifications, Standards, and related Requirements 4
J 5.15 or 6 Sigma? What multiplier to use when performing a Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
A SOP for performing Annual Product Reviews for Medical Device and Pharma Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
D Severity or Occurrence? Performing an FMEA on Motorcycle Operation FMEA and Control Plans 9
K Performing PPAP Retroactively APQP and PPAP 8
H Proprietary Processes - How to Protect and still comply when performing an FAI? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
J Performing Internal Calibrations - Dial gauges, micrometers, height gauges, calipers General Measurement Device and Calibration Topics 8
D External Parties performing Internal Audits? TS 16949 Clause 8.2.2 IATF 16949 - Automotive Quality Systems Standard 15
JoCam Budding Consultant: What to charge for performing a desktop study. Career and Occupation Discussions 10
Miner Upcoming article - Within-part variation when performing an R&R Imported Legacy Blogs 6
D Performing Internal Audit of our Accounts System Internal Auditing 4
S Performing Internal Audits as per Canada MDR GD 210 Internal Auditing 7
E Minimum data for performing an Anderson Darling Test Statistical Analysis Tools, Techniques and SPC 1
A Difficulties in performing MSA on Rockwell Hardness Tester Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 20
G Performing an Internal Audit on the Cost Estimating Process AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
6 Badly Performing Packaging Machines Manufacturing and Related Processes 1
D People Management - Managing an under performing subordinate Career and Occupation Discussions 15
armani Competence for performing compliance audits - 4.5.2 Miscellaneous Environmental Standards and EMS Related Discussions 6
C New to performing internal audits - Looking for audit questions Internal Auditing 5
P For business performing PRODUCT DEVELOPMENT ONLY, what is best QMS for certification? Quality Manager and Management Related Issues 2
A Justification for not performing overseas supplier audit General Auditing Discussions 6
A Performing Ring Gage Calibration General Measurement Device and Calibration Topics 5

Similar threads

Top Bottom