Performing Investigation of used returned expired devices

S

Shakstang

#1
We have a next generation device on the market. Next generation device was released in August 2007. We are still receiving back used devices, which failed during a procedure and require investigations. The devices coming back have Lot Numbers (from the old generation devices) that have already surpassed the shelf life (expired). Per FDA guideline, do we as a device manufacturer still have to perform thorough investigations on those "used" expired products or can we just create a one-page letter stating that since the shelf life has expired, it will be redundant to perform a thorough investigation. In addition, the next generation device should correct the failure mode that was seen in the expired returned/used device(s).
 
Last edited by a moderator:
Elsmar Forum Sponsor

Wes Bucey

Prophet of Profit
#2
I've never heard a problem quite like this. I'll be interested in the replies we get from Covers more familiar with dealing with devices which have failed in the field.

At first blush, it might seem important to do a thorough investigation to assure you do NOT need to issue a warning or alert to end users to withdraw devices from use at the expiration date, especially depending on whether the investigation shows some age- or use-related failure endemic to all devices (new version as well as old.)
 

CarolX

Super Moderator
Super Moderator
#3
Hi Shakstang and welcome to the Cove!

I moved your thread to the FDA board - may be able to get you some responses here.
 
M

madannc

#4
Next generation device was released in August 2007. We are still receiving back used devices, which failed during a procedure and require investigations. The devices coming back have Lot Numbers (from the old generation devices) that have already surpassed the shelf life (expired).
Why is it taking so long for you to recieve the failed devices back?, there is a clause 820.198 (a) (1) ...All complaints shall be processed in a uniform and timely manner... It appears your shelf life is short therefore the mechanism/process for returns should be investigated.

Per FDA guideline, do we as a device manufacturer still have to perform thorough investigations on those "used" expired products or can we just create a one-page letter stating that since the shelf life has expired, it will be redundant to perform a thorough investigation.
820.198 (c) ...Any complaint involving the possible failure of of a device... to meet any of its specifications shall be reviewed, evaluated and investigated, unless such investigation already been performed for similar...

You have stated that the next generation "should" correct the failure mode, I would be more assertive here it either does or does not, if it does how do you know, what I am driving at is, if you were aware of the problem and have already investigated and corrected, this may cover this clause, I say may because I do not know all the details and have not seen what work was done to produce next generation... I would get your QA and RA group to look closely at this.

In answer to your your question "thorough investigation" I have one of my own

Was this a reportable event i.e. FDA part 803?

IMHO what this hinges on is the work that was done for next generation, and what you are doing to address the lateness of customer returns
 

GStough

Staff member
Super Moderator
#5
Why is it taking so long for you to recieve the failed devices back?, there is a clause 820.198 (a) (1) ...All complaints shall be processed in a uniform and timely manner... It appears your shelf life is short therefore the mechanism/process for returns should be investigated.



820.198 (c) ...Any complaint involving the possible failure of of a device... to meet any of its specifications shall be reviewed, evaluated and investigated, unless such investigation already been performed for similar...

You have stated that the next generation "should" correct the failure mode, I would be more assertive here it either does or does not, if it does how do you know, what I am driving at is, if you were aware of the problem and have already investigated and corrected, this may cover this clause, I say may because I do not know all the details and have not seen what work was done to produce next generation... I would get your QA and RA group to look closely at this.

In answer to your your question "thorough investigation" I have one of my own

Was this a reportable event i.e. FDA part 803?

IMHO what this hinges on is the work that was done for next generation, and what you are doing to address the lateness of customer returns
I would also consider 820.198(b) "Each manufacturer shall review and evaluate all complaints to determine whether an investigation is necessary. When no investigation is made, the manufacturer shall maintain a record that includes the reason no investigation was made and the name of the individual responsible for the decision not to investigate."
 
Thread starter Similar threads Forum Replies Date
Geoff Cotton Performing a Delta Audit General Auditing Discussions 12
A Does AS9100 require traceability to operators performing the work? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
I Root Cause Analysis when someone simply stops performing a task? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
H Question about implications of performing Firmware upgrade via MDDS - Medical Device Data Systems 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
N Justifications for not performing MSA (Measurement System Analysis) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
V Sequence of performing risk assessment: User_FMEA (User Errors) vs Design Inputs FMEA and Control Plans 1
Y Performing product audits in a manufacturing facility Manufacturing and Related Processes 13
A ISO 9001:2015 - How to apply 7.1.5 to an organization performing only design ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
S Performing PQ on a Small (tabletop) Air Heat Treating Oven ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
B Thoughts on performing an ISO 9001:2015 Remote Internal Audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
H Performing a TUS on a Furnace used at One Temperature - AMS2750-E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 27
drgnrider EMR performing audits of EMS Miscellaneous Environmental Standards and EMS Related Discussions 13
C Performing an ISO 9001 based Supplier Audit - Multiple Supplier Facilities Supplier Quality Assurance and other Supplier Issues 4
D Performing a Gage R and R with No Specification Range Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
C Going Rate for Performing a Supplier Audit? Consultants and Consulting 1
B Performing MSA on Straightness of a Cylinder Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 13
C Performing a MSA as it relates to Calibration Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
J Performing a MSA on a new gauge using different parts Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
K Production Operators Performing GP12 (General Motors Containment) Customer and Company Specific Requirements 2
L How is the interaction physically happening while performing ANOVA GR&R? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
E Exam for Personnel performing Calibration General Measurement Device and Calibration Topics 7
B Performing a Gage R&R Study with Minitab 13 Using Minitab Software 5
M Requirements for Laboratories performing ASTM Tests Various Other Specifications, Standards, and related Requirements 4
J 5.15 or 6 Sigma? What multiplier to use when performing a Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
A SOP for performing Annual Product Reviews for Medical Device and Pharma Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
D Severity or Occurrence? Performing an FMEA on Motorcycle Operation FMEA and Control Plans 9
K Performing PPAP Retroactively APQP and PPAP 8
H Proprietary Processes - How to Protect and still comply when performing an FAI? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
J Performing Internal Calibrations - Dial gauges, micrometers, height gauges, calipers General Measurement Device and Calibration Topics 8
D External Parties performing Internal Audits? TS 16949 Clause 8.2.2 IATF 16949 - Automotive Quality Systems Standard 15
JoCam Budding Consultant: What to charge for performing a desktop study. Career and Occupation Discussions 10
Miner Upcoming article - Within-part variation when performing an R&R Imported Legacy Blogs 6
D Performing Internal Audit of our Accounts System Internal Auditing 4
S Performing Internal Audits as per Canada MDR GD 210 Internal Auditing 7
E Minimum data for performing an Anderson Darling Test Statistical Analysis Tools, Techniques and SPC 1
A Difficulties in performing MSA on Rockwell Hardness Tester Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 20
G Performing an Internal Audit on the Cost Estimating Process AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
6 Badly Performing Packaging Machines Manufacturing and Related Processes 1
D People Management - Managing an under performing subordinate Career and Occupation Discussions 15
A Competence for performing compliance audits - 4.5.2 Miscellaneous Environmental Standards and EMS Related Discussions 6
C New to performing internal audits - Looking for audit questions Internal Auditing 5
P For business performing PRODUCT DEVELOPMENT ONLY, what is best QMS for certification? Quality Manager and Management Related Issues 2
A Justification for not performing overseas supplier audit General Auditing Discussions 6
A Performing Ring Gage Calibration General Measurement Device and Calibration Topics 5
E Hypothesis Test - Determining if a second wash is worth performing Statistical Analysis Tools, Techniques and SPC 7
C Guidance in Performing a 13485 Gap Analysis ISO 13485:2016 - Medical Device Quality Management Systems 8
Douglas E. Purdy Personnel performing specific assigned tasks shall be qualified - Interpretation IATF 16949 - Automotive Quality Systems Standard 6
F Clinical Audits to ISO 14155 - Performing audits of clinical affairs groups ISO 13485:2016 - Medical Device Quality Management Systems 4
T Non-conformity - Not performing Supplier Development on one specific supplier IATF 16949 - Automotive Quality Systems Standard 25

Similar threads

Top Bottom