Please explain MDR Rule 11 to me
Ronen E
Problem Solver
Moderator
#42
Another possible interpretation of the implementation rules (which makes more sense in my opinion) is that the part "Software, which drives a device or influences the use of a device, shall fall within the same class as the device"only applied to embedded software.
That would make sense, however the MDR's text doesn't indicate that this was the actual intention.
dgrainger
#43
I agree, it's a mess!
Implementing rule 3.3 contradicts 3.2 :( :
3.2 - If the device in question is intended to be used in combination with another device, the classification rules shall apply separately to each of the devices.​
vs
3.3 - Software, which drives a device or influences the use of a device, shall fall within the same class as the device.​
HOWEVER you have to apply 3.5 and apply all the rules and take the strictest rule. - easy!:)
kreid
#44
The same way you do with the other - analyzing the rule. If your device (in this case of rule 11, as I mentioned it has to be a SaMD) if the intended use fits into the first two paragraphs, it's class is like defined in the paragraph. If not, it's a class I.
But analysing the rule I see the following:
"Software intended to provide information which is used to take decisions with diagnosis or therapeutic purposes is classified as class IIa, except if such decisions have an impact that may cause:"
"Software intended to monitor physiological processes is classified as class IIa,"
Do not the two quotes above just about define a medical device? Hence, an software medical device has to be Class IIa or above.
I do understand dgrainger's point about accessories.
Marcelo
#45
What defines what is a medical device, is the medical device definition in article 2(1). Classification does not define if a device is a medical device or not.
Did you read document IMDRF/SaMD WG/N12 FINAL:2014? It has several examples.
Unfortunately, I don't think you can truly understand Rule 11 without having an in depth understanding of the 4 IMDRF SaMD documents (in particular, the first on definitions and this one I mentioned), even if the MDR does not follow strictly the IMDRF documents.
kreid
#46
Yes, know the IMDRF SaMD docs well, the point I was trying to make (badly) was that the two parts of Rule 11 that I quoted are very similar to that which defines a medical device in article 2, if a device is used in diagnosis, therapies or monitoring of disease, injury or disability, then it is a medical device. So how do you get to Class I and still be a medical device.
I guess this is a philosophical discussion, but in practical terms, I am telling my clients that if they have an SaMD, it is difficult to see how it can be classified as anything less than a Class IIa device, if Rule 11 is the classification rule used.
Marcelo
#47
I'm still not sure why you are claiming that software can only be Class IIA.
If the software is NOT intended to provide information which is used to take decisions with diagnosis or therapeutic purposes and is NOT intended to monitor physiological processes, but is still a SaMD, it will be class I as per the rule.
I think the confusion here is exactly you trying to link to the medical device definition.
If you look closely to what Rule 11 says, is says "provide information which is used to take decision with diagnosis or therapeutic purposes". This is not in the medical device definition (the diagnosis and therapeutic purposes are). The same for monitoring in the second part of Rule 11.
If you look at the definition of a medical device (bold emphasis is from me)
diagnosis, prevention, monitoring, prediction, prognosis, treatment or alleviation of disease,
diagnosis, monitoring, treatment, alleviation of, or compensation for, an injury or disability,
— investigation, replacement or modification of the anatomy or of a physiological or pathological process or state,
— providing information by means of in vitro examination of specimens derived from the human body, including organ, blood and tissue donations
Not all aspects of the definition of medical devices matches what is in rule 11 (basically, only the ones in bold). If the SaMD, in principle, is not related to ones in bold, it will in principle be Class I. I mean in principle because it's not automatic, you would need to check some other aspects, but an initial analysis should have this conclusion.
dgrainger
#48
The interpretation of rule 11 is that all software that is a medical device will be covered by Rule 11 - "information which is used to take decision with diagnosis or therapeutic purposes" covers everything listed in the medical device definition.
The non-bold items listed are considered to be either diagnostic or therapeutic purposes.
Marcelo
#49
Interpretation from who? You?
Ronen E Can anyone please explain this? (510k submission requirement) Other US Medical Device Regulations 14
Sidney Vianna Interesting Discussion Hello IAF. Are you listening? Please explain the inconsistency in your IAF Mandatory Documents (2019) ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 6
T From 21 CFR to IEC 60601 - Please explain IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Please Recommend books that explain how to understand IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
M Please explain the following Automotive Acronyms IATF 16949 - Automotive Quality Systems Standard 3
S Please explain Medical Device Product Classification Rationale 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
B Please Explain Clause 8.3 in ISO 9001:2008 Nonconformance and Corrective Action 4
P EN ISO vs. ISO - Please explain to me EN ISO vs. ISO standards Other ISO and International Standards and European Regulations 3
L Please Explain Inspection Levels including C=0 Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
J Please explain what formula is correct for "Mean and Range Control Chart" Statistical Analysis Tools, Techniques and SPC 1
A Calibration vs. Inspection - Please explain the difference General Measurement Device and Calibration Topics 3
G Please explain to me what .asp does Software Quality Assurance 5
A Please explain to me in laymans terms how to calculate a single T-test Six Sigma 14
P Layered Process Audit & the Document Audit - Can anyone explain to me, Please? Process Audits and Layered Process Audits 2
D Definition Technical Files - Please explain differences and what information would be included Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 23
D Amendment Class II - Please confirm if the submission is by email (Canada) - Help Canada Medical Device Regulations 0
E 13485:2016, Sections 4.1.6, 7.5.6 and 7.6 - Validation of Software - Need some Advice please ISO 13485:2016 - Medical Device Quality Management Systems 3
H MDD Article 12 Labeling for Class IIa Medical Device - Please Advise EU Medical Device Regulations 3
O Please, help with document on ISO/ TS 17758 and (IDF 87: 1979). Thank you. Manufacturing and Related Processes 1
R Please share your method used for PEMS Validation IEC 62304 - Medical Device Software Life Cycle Processes 5
J ISO 17025 - Please share a sample procedure for monitoring the validity of results ISO 17025 related Discussions 1
D Good Acceptance Activity/Inspection flowchart for reference - Wanted Please Process Maps, Process Mapping and Turtle Diagrams 2
optomist1 Share a DFMEA template please Document Control Systems, Procedures, Forms and Templates 2
alonFAI Please share a medical device company NCMR procedure Other Medical Device Related Standards 5
B Will you please help me with my academic project? Repair & calibration service for test equipment ISO 17025 related Discussions 0
J Medical Device Vigilance decision tree FORM - Please share yours EU Medical Device Regulations 15
Marc Posting Job Openings - Read Me Please - 27 January 2019 Job Openings, Consulting and Employment Opportunities 0
L Please help me with the process of getting ISO 9001 certified Quality Management System (QMS) Manuals 15
K Please help identify appropriate statistical treatment Statistical Analysis Tools, Techniques and SPC 13
L Please review my ISO 9001:2015 Internal Audit Plan Process Audits and Layered Process Audits 7
C Will anyone please share training material for ISO:13485:2016 for best practices Training - Internal, External, Online and Distance Learning 0
L ISO 9001 - 2015, is the Quality Manual needed? Please advise ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
K Please Help! what is ARL for the range and standard deviation Statistical Analysis Tools, Techniques and SPC 0
E Best GD&T training - Your suggestions and recommendations please Training - Internal, External, Online and Distance Learning 3
R Please help me with attached copy of AS 9100D internal audit check sheet AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S Supplier Classifications - Please give me a good definition of the following terms ISO 13485:2016 - Medical Device Quality Management Systems 5
L Please help my division with quality [long] Service Industry Specific Topics 1
J Please help me understand what a process is ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
D Please share your IATF 16949 Internal Audit Plan and Schedule Internal Auditing 13
M MDSAP Audit Plan - Please share an Audit Plan/Schedule/Agenda ISO 13485:2016 - Medical Device Quality Management Systems 7
S New Interaction of Processes ISO 9001:2015 - Please Review ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
M New Quality Policy for ISO 9001:2015 - Please Review ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
K Please share a Product Audit Checklist for a Wafer Fab Document Control Systems, Procedures, Forms and Templates 3
Moncia Complaint / Non-Conforming Product / CAPA Procedures - Please share your experience ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
V Revised ISO std vs ISO13485:2003 - Help Please!!! ISO 13485:2016 - Medical Device Quality Management Systems 4
S Please recommend any good training courses for Medical Device Regulations Training - Internal, External, Online and Distance Learning 8
E Supplier Evaluation Form - Example to evaluation suppliers (criteria) please Document Control Systems, Procedures, Forms and Templates 4
S Please review my Risk Analysis Table ISO 14971 - Medical Device Risk Management 13
somashekar MDSAP (Medical Device Single Audit Program) Pilot: Please share your feedback ISO 13485:2016 - Medical Device Quality Management Systems 35
L Canadian Users of ISO 14001 please contact me ISO 14001:2015 Specific Discussions 4