Hello All,

All our current product do not require a PMCF, so we are creating a PMCF checklist to justify it.
One of the PSUR requirements is to report the main findings of the PMCF. If we don't need PMCF for our device then how do we add it's main findings?

Please advise. Thanks!
 
Elsmar Forum Sponsor

Marcelo

Inactive Registered Visitor
#3
PMCF = Post Market clinical followup
PSUR = ???
PSUR = Periodic Safety Update Report


Hello All,

All our current product do not require a PMCF, so we are creating a PMCF checklist to justify it.
One of the PSUR requirements is to report the main findings of the PMCF. If we don't need PMCF for our device then how do we add it's main findings?

Please advise. Thanks!
PMCF is an ongoing process, a continuous process that updates the clinical evaluation, and is always applicable tom all devices. You probably is saying that a PMCF study (which is one of the tools used in the PMS process) is not applicable to your device. So you will always have PMCF findings to report in the PSUR.
 

lisasolo

Starting to get Involved
#7
Section 6.1 of Annex II on Tech Documentation says to include "the PMCF plan and PMCF evaluation report referred to in Part B of Annex XIV or a justification why a PMCF is not applicable."
In Annex III on the PMS plan, it also says to include "a PMCF plan as referred to in Part B of Annex XIV, or a justification as to why a PMCF is not applicable."
These seem to imply that it is possible for PMCF to not be applicable. However, there are also many other places in the MDR that says that you shall perform PMCF, implying that it is required. Does anyone know if it is still appropriate to justify PMCF as not applicable under the MDR?
 

Marcelo

Inactive Registered Visitor
#8
There seems to be a inconsistency in the PMCF requirements, but to the this seems to stem from the confusion between PMCF and PMCF study, as I mentioned above.

So, although some parts of the MDR does mention "justification why a PMCF is not applicable", if you read the PMCF "definition"on Annex IV is says:

PART B
POST-MARKET CLINICAL FOLLOW-UP
5. PMCF shall be understood to be a continuous process that updates the clinical evaluation referred to in Article 61 and Part A of this Annex and shall be addressed in the manufacturer's post-market surveillance plan.
As clinical evaluation is required to all devices, PMS shall also be required. This whole annex is very clear on this.
 

Highground

Involved In Discussions
#9
Does that mean we have to have a procedure in place for PMCF? What if we just create a checklist that we can use to justify that we do not require PMCF study?
This is a form we are providing to justify why no PMCF study is required. Of course, there should be more information added to the form.
 

Attachments

Thread starter Similar threads Forum Replies Date
A Post-market clinical followup studies (PMCF) EU Medical Device Regulations 16
S PMCF (Post Market Clinical Followup): Rationale for Exemption - Software Medical Devices EU Medical Device Regulations 6
W Interesting Discussion PMCF (Post-Market Clinical Followup) vs PMCF studies EU Medical Device Regulations 39
C PMCF (Post Market Clinical Followup) Enrollment Sizes (Spinal Implants) EU Medical Device Regulations 5
S Understanding PMS (Post Market Surveillance) and PMCF (Vigilance and PMCF) Quality Manager and Management Related Issues 1
A Difference between general PMCF and PMS EU Medical Device Regulations 0
T PMCF Planning : Only Literature screening qualify? EU Medical Device Regulations 5
K MDCG 2020-7 PMCF Plan Similar Devices EU Medical Device Regulations 1
D Is PMCF really a continuous activity per Annex XIV,Part B? EU Medical Device Regulations 5
S PMS (Premarket Submission) and PMCF Plan EU Medical Device Regulations 5
M Clinical Evaluation Plan vs. PMCF Plan EU Medical Device Regulations 33
A CE Mark and PMCF study EU Medical Device Regulations 3
M MDR - Under what circumstances is a PMCF not required? EU Medical Device Regulations 6
R PMCF - centers/countries used CE Marking (Conformité Européene) / CB Scheme 0
R PMCF plan to justify that our product not require PMCF study EU Medical Device Regulations 11
B PMCF study plan is same to clinical investigation plan? EU Medical Device Regulations 3
S Do subjects pay for a medical implant in PMCF study? Quality Manager and Management Related Issues 3
Q Is there a way that I can combine two PMCF studies into one? EU Medical Device Regulations 4
B Inquiry about MEDDEV 2.12/2 rev2 - PMCF studies EU Medical Device Regulations 3
L Residual risks which the PMCF study is based on ISO 14971 - Medical Device Risk Management 7
A Post-market Clinical Follow-up Other Medical Device Related Standards 3
S Post Market Surveillance Report (PMSR) for Medical Devices EU Medical Device Regulations 6
A SaMD: Post Market Surveillance after Decomissioning EU Medical Device Regulations 2
MaHoDie Summative Evaluation with Post-Market Data? Human Factors and Ergonomics in Engineering 2
I Post Market Surveillance Process EU Medical Device Regulations 3
Watchcat REGULATORY WATCHCAT De Novo Post-Mortem 6 – How Does This Even Happen? Other US Medical Device Regulations 0
Watchcat REGULATORY WATCHCAT De Novo Post-Mortem 5 – What's the Rush? Other US Medical Device Regulations 0
Watchcat REGULATORY WATCHCAT De Novo Post-Mortem 4 – Design Controls Other US Medical Device Regulations 0
I Post-market surveillance and Post-market performance follow-up for in vitro diagnostic medical devices EU Medical Device Regulations 0
Watchcat REGULATORY WATCHCAT De Novo Post-Mortem 3 - Rearranging the Regs Other US Medical Device Regulations 0
Watchcat REGULATORY WATCHCAT De novo Post Mortem 2 - Disclosure and Accountability Other US Medical Device Regulations 11
Watchcat REGULATORY WATCHCAT De novo Post Mortem 1 - Regulatory Religion Other US Medical Device Regulations 0
Watchcat REGULATORY WATCHCAT De novo Post Mortem Other US Medical Device Regulations 0
A Direct vs. Post-License Investigational Test Authorization (ITA) Canada Medical Device Regulations 0
Miner Reported content: Post in thread 'Form of real time study for establish shelf time' Document Control Systems, Procedures, Forms and Templates 1
R Auditing COT's Suppliers / First Post Supplier Quality Assurance and other Supplier Issues 2
M Post Market Safety Update Report for devices that will be up-classified under MDR but are currently under MDD EU Medical Device Regulations 6
thisby_ Post Market/Production Risk Assessment ISO 14971 - Medical Device Risk Management 0
dgrainger Informational MHRA - Regulating medical devices in the UK - Post Brexit - 2021-02 UK Medical Device Regulations 0
C Production and Post Production feedback - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 4
R Impact of changing manufacturer post clinical investigation CE Marking (Conformité Européene) / CB Scheme 3
N Post Market Surveillance Package Example Wanted ISO 13485:2016 - Medical Device Quality Management Systems 1
M Differences in post market safety reporting for Combination Product Applicants Medical Device and FDA Regulations and Standards News 1
K Product (Device) Life cycle - Product Life AFTER the Post Market Phase Medical Device and FDA Regulations and Standards News 3
M Status of Medical Device Regulations in UK Post-Brexit Other Medical Device Regulations World-Wide 10
M Post Market Surveillance for Class II Medical Devices Medical Device and FDA Regulations and Standards News 3
N Post registration variation to medical device Other Medical Device Regulations World-Wide 1
F Medical Devices - Hong Kong - Post approval changes regulation (China's MOH) Other Medical Device Regulations World-Wide 3
H Post-market surveillance strategy EU Medical Device Regulations 2
B Procedures for Complaint Handling and Post Market Surveillance EU Medical Device Regulations 10

Similar threads

Top Bottom